- Materials
-
- <1 µg of each DNA sample to be barcoded in 45 µl
- Consumables
-
- NEBNext Ultra II End repair/dA-tailing Module (NEB, E7546)
- Freshly prepared 70% ethanol in nuclease-free water
- Nuclease-free water (e.g. ThermoFisher, AM9937)
- Agencourt AMPure XP beads (Beckman Coulter, A63881)
- 0.2 ml thin-walled PCR tubes
- Equipment
-
- Thermal cycler
- Ice bucket with ice
- Microfuge
- Magnetic rack, suitable for 1.5 ml Eppendorf tubes
- Optional equipment
-
- Qubit fluorometer (or equivalent for QC check)
-
Prepare the DNA in nuclease-free water.
- Transfer <1 μg DNA of each sample into a fresh 0.2 ml PCR tube or plate
- Adjust the volume to 45 μl with nuclease-free water
- Mix thoroughly by flicking the tube to avoid unwanted shearing
- Spin down briefly in a microfuge
-
Using a separate 0.2 ml thin-walled PCR tube for each DNA sample to be barcoded, set up the end-repair / dA-tailing reactions as follows:
Between each addition, pipette mix 10-20 times.
Reagent Volume <1 µg DNA 45 µl Ultra II End-prep reaction buffer 7 µl Ultra II End-prep enzyme mix 3 µl Nuclease-free water 5 µl Total 60 µl -
Mix by pipetting.
-
Using a thermal cycler, incubate at 20°C for 5 minutes and 65°C for 5 minutes.
-
Transfer each sample to a separate 1.5 ml Eppendorf DNA LoBind tube.
-
Resuspend the AMPure XP beads by vortexing.
-
Add 60 µl of resuspended AMPure XP beads to the end-prep reaction and mix by pipetting.
-
Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
-
Prepare sufficient fresh 70% ethanol in nuclease-free water.
-
Place on a magnetic rack, allow beads to pellet and pipette off supernatant.
-
Keep the tube on the magnet and wash the beads with 180 µl of freshly-prepared 70% ethanol without disturbing the pellet. Remove the ethanol using a pipette and discard.
-
Repeat the previous step.
-
Spin down and place the tubes back on the magnet. Pipette off any residual ethanol. Allow to dry for ~30 seconds, but do not dry the pellet to the point of cracking.
-
Remove the tubes from the magnetic rack and resuspend the pellet in 31 µl nuclease-free water. Incubate for 2 minutes at room temperature.
-
Pellet the beads on a magnet until the eluate is clear and colourless.
-
Remove eluate once it is clear and colourless. Transfer each eluted sample to a new 0.2 ml thin-walled PCR tube.
-
Quantify 1 µl of end-prepped DNA using a Qubit fluorometer - recovery aim >700 ng.