- Materials
-
- Flow Cell Wash Kit (EXP-WSH004) or Flow Cell Wash Kit XL (EXP-WSH004-XL)
- Equipment
-
- P20 pipette and tips
- P1000 pipette and tips
-
Storage Buffer (S) can be used to flush flow cells for storage for later use or to check number of available nanopores before loading another library.
-
Thaw one tube of Storage Buffer (S) at room temperature.
-
Mix contents thoroughly by pipetting and spin down briefly.
-
Slide the flow cell priming port cover clockwise to open.
-
After opening the priming port, check for a small air bubble under the cover. Draw back a small volume to remove any bubbles:
- Set a P1000 pipette to 200 µl.
- Insert the tip into the flow cell priming port.
- Turn the wheel until the dial shows 220-230 µl, or until you can see a small volume of buffer/liquid entering the pipette tip.
- Visually check that there is continuous buffer from the flow cell priming port across the sensor array.
-
Slowly add 500 μl of Storage Buffer (S) through the flow cell priming port.
-
Close the priming port.
-
Remove all fluid from the waste channel through waste port 1 using a P1000 pipette.
As both the flow cell priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
-
The flow cell can now be stored at 4-8°C.