Version for device: MinION
Overview of the protocol
The Flow Cell Wash Kit allows sequential runs of multiple sequencing libraries on the same flow cell. It works by flushing out and digesting the first library, refreshing the system for a subsequent library to be loaded. This procedure provides the opportunity to utilise the same flow cell a number of times, maximising the available run time, particularly for cases where less data per library is required. Following the wash step, Storage Buffer can be introduced into the flow cell, allowing storage of the flow cell at 4-8°C. The Flow Cell Wash Kit is compatible with Q-Line, R9.4.1 and R10.4.1 flow cells.
The nuclease digests DNA library and for effective removal from the flow cell before loading a new library, we recommend removing the waste buffer from the waste ports after each flush step. This is to ensure there is no nuclease in the waste channel that may diffuse through the flow cell during sequencing. To further improve flow cell output after a wash, we recommend pipetting slowly during flushing steps by either twisting the pipette wheel down slowly or pressing the plunger down slowly.
Please note, although the wash procedure should remove 99.9% of the library, some residual DNA may remain on the flow cell. For this reason, we recommend to barcode your libraries when used in conjunction with the Flow Cell Wash Kit, to ensure that reads from different libraries can be separated from each other.
Successful deconvolution of DNA reads has been demonstrated in Oxford Nanopore's internal development:
Figure 1. A sample with four barcodes was sequenced and washed using EXP-WSH004 before a sample with four different barcodes was loaded. This was repeated for a third sequencing run.
For users who wish to use barcoding to run multiple libraries at one time rather than washing the flow cells, please see the barcoding kits we have available in the chemistry technical document.
The Flow Cell Wash Kit is can also be used with our RNA flow cells to flush out RNA libraries, but will not remove RNA-related blocking and restore nanopores.
The Flow Cell Wash Kit contains DNase I, that is used to digest any remaining DNA library on a flow cell. Once the library is removed, the flow cell can be re-used immediately or stored for later use.
During sequencing, an accumulation of pores in the “unavailable" state (Figure 2) may be observed, causing the rate of data acquisition to decline as fewer pores are available to accept and sequence strands. We have demonstrated that in these circumstances, pores can be reverted to the “active pore” state by pausing sequencing and washing the flow cell with the DNase I in the Flow Cell Wash Kit. In Figure 2, the astrisks indicate where sequencing has been paused and the flow cell washed. Note: If the sequencing run is paused in MinKNOW for the flow cell wash, you will only see the restoration of sequencing pores after a new pore scan has been performed.
The wash step is recommended where sequencing channels are lost to the “unavailable” state (Figure 2). In circumstances where channels have been lost by other means, for example “saturated”, the wash step is not effective at reverting channels to the “active pore” state. If your sample is not known to block and cause saturation of the flow cell, we recommend loading a fresh library on a new flow cell. If you do not have excess library to load on a new flow cell, you can recover the library and reload on a new flow cell following method 1 of the Library recovery from flow cells protocol.
Figure 2. Pore states observed on a flow cell before and after wash steps are performed. A flow cell has been loaded with a sequencing library that has resulted in an accumulation of pores in the “unavailable” state, leading to a decrease in the rate of data acquisition. The red asterisks indicate when a wash step has been performed. A significant number of the pores that had been lost to the “unavailable” state have reverted to the “Pore available” state and are available for sequencing once again.
In experiments where output is limited by the increase in pores in the “unavailable” state, we have shown that output can be improved by performing several wash steps over the lifetime of a flow cell. Figure 3 shows the output obtained from a PromethION Flow Cell loaded with a library of DNA extracted from chicken - Gallus gallus, and a MinION Flow Cell loaded with a library of DNA extracted from a type of Japense ricefish - Oryzias latipes, where unavailable pores increased over the course of the experiment, and so flow cell washes were performed to unblock the pores (Figure 3). In each case, the use of multiple washes allowed for an improvement of the output from the flow cell, without any compromise in observed read length (Figure 4).
Figure 3. Throughput observed from Gallus gallus and Oryzias latipes libraries run on a PromethION Flow Cell and a MinION Flow Cell, respectively. The arrows indicate the timing of each wash step: wash steps were performed at the point where the rate of data acquisition started to slow due to the accumulation of “recovering” pores. In each case, output is more than doubled from the point of the first wash.
Figure 4. Effective inactivation of the DNase I prevents read length deterioration in the experiments after a nuclease wash is performed. In this example, the read length of the Gallus gallus library was recorded before the first wash (load #1) and then again after the first and second washes (load #2 and #3, respectively). No decrease in read length is observed.
Equipment and consumables
Additional buffers are required for reloading a library following the washing of a flow cell. These can be found in the one of the following expansion kits:
Sequencing Auxiliary Vials V14 (EXP-AUX003). This expansion contains vials of Sequencing Buffer (SB), Elution Buffer (EB), Library Solution (LIS), and Library Beads (LIB) with Kit 14 flow cell priming reagents: Flow Cell Flush (FCF) and Flow Cell Tether (FCT). This expansion is only compatible with our Kit 14 chemistry e.g. SQK-LSK114.
Flow Cell Priming Kit (EXP-FLP004). This expansion contains both Kit 14 flow cell priming reagents required: Flow Cell Flush (FCF) and Flow Cell Tether (FCT). This expansion is only compatible with Kit 14 chemistry.
For our previous chemistries:
Sequencing Auxiliary Vials expansion (EXP-AUX001). This expansion contains vials of Elution Buffer (EB), Sequencing Buffer (SQB), and Loading Beads (LB), additional to those found in DNA sequencing kits for our Kit 9 chemistry.
Sequencing Auxiliary Vials expansion (EXP-AUX002). This expansion contains vials of Sequencing Buffer II (SBII), Elution Buffer (EB), Loading Solution (LS), and Loading Beads II (LBII), additional to those found in Kit 10 and 11 chemistry, such as:
Contents | Volume (µl) | No. of tubes | No. of uses |
---|---|---|---|
Wash Mix (WMX) | 15 | 1 | 6 |
Wash Diluent (DIL) | 1,300 | 2 | 6 |
Storage Buffer (S) | 1,600 | 2 | 6 |
Name | Acronym | Fill volume per vial (µl) | Cap colour | No. of vials | No. of uses |
---|---|---|---|---|---|
Wash Mix | WMX | 150 | Brown | 1 | 48 |
Wash Diluent | DIL | 20,000 | White cap, brown stripe | 1 | 48 |
Storage Buffer | S | 25,000 | White cap, yellow stripe | 1 | 48 |
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (μl) |
---|---|---|---|---|
Flow Cell Flush | FCF | 6 | Clear cap, light blue lable | 8,000 |
Flow Cell Tether | FCT | 1 | Purple | 200 |
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (μl) |
---|---|---|---|---|
Elution Buffer | EB | Black | 2 | 500 |
Sequencing Buffer | SB | Red | 2 | 700 |
Library Solution | LIS | White cap, pink label | 2 | 600 |
Library Beads | LIB | Pink | 2 | 600 |
Flow Cell Flush | FCF | Light blue label | 2 | 8,000 |
Flow Cell Tether | FCT | Purple | 2 | 200 |
Flushing a MinION/GridION Flow Cell
If the flow cell is to be washed and stored, the light shield can be removed.
Reagent | Volume per flow cell |
---|---|
Wash Mix (WMX) | 2 μl |
Wash Diluent (DIL) | 398 μl |
Total | 400 μl |
Note: As both the priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
Note: As both the priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
To run a second library on a MinION/GridION Flow Cell straight away
If using a previous version of our chemistry or a kit with specific sequencing reagents, please ensure you are using the correct sequencing reagents for your flow cell.
Note: We do not recommend using any other albumin type (e.g. recombinant human serum albumin).
Reagent | Volume per flow cell |
---|---|
Flow Cell Flush (FCF) | 1,170 µl |
Bovine Serum Albumin (BSA) at 50 mg/ml | 5 µl |
Flow Cell Tether (FCT) | 30 µl |
Total volume | 1,205 µl |
During this time, prepare the library for loading by following the steps below.
We recommend using the Library Beads (LIB) for most sequencing experiments. However, the Library Solution (LIS) is available for more viscous libraries.
For Kit 14 chemistry:
Reagent | Volume per flow cell |
---|---|
Sequencing Buffer (SB) | 37.5 µl |
Library Beads (LIB) mixed immediately before use, or Library Solution (LIS), if using | 25.5 µl |
Recovered DNA library | 12 µl |
Total | 75 µl |
Note: As both the priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
Note: As both the priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.
Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.
We recommend storing libraries in Eppendorf DNA LoBind tubes at 4°C for short term storage or repeated use, for example, reloading flow cells between washes.
For single use and long-term storage of more than 3 months, we recommend storing libraries at -80°C in Eppendorf DNA LoBind tubes.
For further information, please refer to the DNA library stability Know-How document.
As both the flow cell priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
Load your flow cell into the device with Storage Buffer (S) and start a Flow cell check to detect the number of active pores. For more information, please visit the Flow cell check section of our MinKNOW protocol.
After the Flow cell check, prime your flow cell and load the library before starting a new sequencing run.
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.