Version for device: MinION
Overview of the protocol
This kit is available on the Legacy page of the store. We are in the process of gathering data to support the upgrade of this protocol to our latest chemistry. Further information regarding protocol upgrades will be provided on the Community as soon as they are available over the next few months. For further information on please see the product update page.
This protocol has been developed by Adela Alcolea-Medina, Prof. Jonathan Edgeworth and colleagues, with support from Oxford Nanopore Technologies (Novel, rapid metagenomic method to detect emerging viral pathogens applied to human monkeypox infections by Alcolea-Medina et al., 2022). It outlines a rapid method to perform metagenomic sequencing from extracted DNA and RNA from routine nasopharyngeal swab samples in viral transport media to identify viral pathogens. This method has also been used to sequence the virus currently known as mpox from skin lesion swabs (in viral transport media) of diagnosed patients, demonstrating an alternative sample site is compatible with this protocol. However, further work is required to assess and validate different sample sites and types.
While this protocol is available in the Nanopore Community, we kindly ask users to ensure they are citing the authors of the paper who have been behind the development of these methods.
This method uses random hexamers and oligo-dT primers to synthesis cDNA strands from RNA within the sample. Subsequently, the Rapid PCR Barcoding kit (SQK-RPB004) is used to amplify the DNA present from the sample alongside the cDNA strands. During development, typically three samples per patient were pooled together after the PCR and clean-up step before sequencing the sample on a single flow cell.
We recommend users follow their local health and safety recommendations when handling samples. Confirmed mpox samples were used for the development of this method and were handled in a Containment Level 3 (CL3) facility.
Prepare for your experiment
You will need to:
Library preparation
You will need to:
Sequencing and analysis
You will need to:
This protocol should only be used in combination with:
Equipment and consumables
During development of this method, primarily nasopharyngeal swabs were tested. However, swabs from skin lesions were also found to yield virus in patients diagnosed with mpox, suggesting swabs from multiple sites can be used with this method to detect for presence of a virus.
It is important that the input meets the quantity and quality requirements. Using too little or too much DNA/RNA, or DNA/RNA of poor quality (e.g. highly fragmented or containing chemical contaminants) can affect your library preparation.
For instructions on how to perform quality control of your DNA sample, please read the Input DNA/RNA QC protocol.
Depending on how the DNA is extracted from the raw sample, certain chemical contaminants may remain in the purified DNA, which can effect library preparation efficiency and sequencing quality.
For further information on using DNA as input, please read the links below:
It is important that the input RNA meets the quantity and quality requirements for highest library preparation efficiency and sequencing quality.
For further information on using RNA as input, please read the links below.
These documents can also be found in the DNA/RNA Handling page.
We have validated and recommend the use of all the third-party reagents used in this protocol. Alternatives have not been tested by Oxford Nanopore Technologies.
For all third-party reagents, we recommend following the manufacturer's instructions to prepare the reagents for use.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Fragmentation Mix | FRM | Brown | 3 | 30 |
Rapid Adapter | RAP | Green | 1 | 10 |
Sequencing Buffer | SBQ | Red | 1 | 300 |
Loading Beads | LB | Pink | 1 | 360 |
Rapid Barcode Primer 1-12 | RLB 01-12 | Clear | 12 | 10 |
Name | Acronym | Cap colour | No. of vial | Fill volume per vial (μl) |
---|---|---|---|---|
Flush Buffer | FB | Blue | 6 | 1,170 |
Flush Tether | FLT | Purple | 1 | 200 |
Component | Sequence |
---|---|
RLB01 | AAGAAAGTTGTCGGTGTCTTTGTG |
RLB02 | TCGATTCCGTTTGTAGTCGTCTGT |
RLB03 | GAGTCTTGTGTCCCAGTTACCAGG |
RLB04 | TTCGGATTCTATCGTGTTTCCCTA |
RLB05 | CTTGTCCAGGGTTTGTGTAACCTT |
RLB06 | TTCTCGCAAAGGCAGAAAGTAGTC |
RLB07 | GTGTTACCGTGGGAATGAATCCTT |
RLB08 | TTCAGGGAACAAACCAAGTTACGT |
RLB09 | AACTAGGCACAGCGAGTCTTGGTT |
RLB10 | AAGCGTTGAAACCTTTGTCCTCTC |
RLB11 | GTTTCATCTATCGGAGGGAATGGA |
RLB12A | GTTGAGTTACAAAGCACCGATCAG |
This kit contains reagents to be used with any remaining barcodes to load another six sequencing libraries.
Reagent | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RAP | Green | 1 | 10 |
Sequencing Tether | SQT | Purple | 1 | 10 |
Loading Beads | LB | Pink | 1 | 360 |
Sequencing Buffer | SQB | Red | 1 | 300 |
Computer requirements and software
Sequencing on a MinION Mk1B requires a high-spec computer or laptop to keep up with the rate of data acquisition. Read more in the MinION Mk1B IT Requirements document.
The MinION Mk1C contains fully-integrated compute and screen, removing the need for any accessories to generate and analyse nanopore data. Read more in the MinION Mk1C IT requirements document.
The MinKNOW software controls the nanopore sequencing device, collects sequencing data and basecalls in real time. You will be using MinKNOW for every sequencing experiment to sequence, basecall and demultiplex if your samples were barcoded.
For instructions on how to run the MinKNOW software, please refer to the MinKNOW protocol.
The EPI2ME cloud-based platform performs further analysis of basecalled data, for example alignment to the Lambda genome, barcoding, or taxonomic classification. You will use the EPI2ME platform only if you would like further analysis of your data post-basecalling.
For instructions on how to create an EPI2ME account and install the EPI2ME Desktop Agent, please refer to the EPI2ME Platform protocol.
We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within three months of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
RNA/DNA extraction
Dispose of the pelleted beads.
From this point, we recommend to keep the sample on ice as much as possible to prevent nucleolytic degradation.
cDNA synthesis
For further information on how to handle RNA, please see the DNA/RNA Handling tab for best practices.
Reagent | Volume |
---|---|
Extracted sample | 16 µl |
LunaScript™ RT SuperMix | 4 µl |
Total | 20 µl |
Step | Temperature | Time | Cycles |
---|---|---|---|
Primer annealing | 25°C | 2 minutes | 1 |
cDNA synthesis | 55°C | 10 minutes | 1 |
Heat inactivation | 95°C | 1 minute | 1 |
Hold | 4°C | ∞ | - |
Reagent | Volume per sample |
---|---|
5X Sequenase reaction buffer | 2 µl |
Nuclease-free water | 7.7 µl |
Sequenase Version 2.0 DNA Polymerase | 0.3 µl |
Total | 10 µl |
Reagent | Volume |
---|---|
Sequenase Dilution Buffer | 0.9 µl |
Sequenase Version 2.0 DNA Polymerase | 0.3 µl |
Total | 1.2 µl |
Dispose of the pelleted beads.
We recommend storing libraries in Eppendorf DNA LoBind tubes at -20°C for short term storage or repeated use, for example, re-loading flow cells between washes.
For single use and long term storage of more than 3 months, we recommend storing libraries at -80°C in Eppendorf DNA LoBind tubes.
Library preparation
Reagent | 1. Thaw | 2. Mix well by pipetting | 3. Briefly spin down |
---|---|---|---|
Rapid Barcode Primers (RBL 01-12A) | At room temperature | ✓ | ✓ |
Fragmentation Mix (FRM) | Not frozen | ✓ | ✓ |
Rapid Adapter | Not frozen | ✓ | ✓ |
Reagent | Volume |
---|---|
1-5 ng template DNA | 3 μl |
Fragmentation Mix (FRM) | 1 μl |
Total | 4 μl |
Reagent | Volume |
---|---|
Nuclease-free water | 20 µl |
Tagmented DNA | 4 µl |
RLB (01-12A, at 10 µM) | 1 µl |
LongAmp Taq 2X master mix |
25 µl |
Total | 50 µl |
If the amount of input material is altered, the number of PCR cycles may need to be adjusted to produce the same yield.
Cycle step | Temperature | Time | No. of cycles |
---|---|---|---|
Initial denaturation | 95°C | 3 mins | 1 |
Denaturation | 95°C | 15 secs | 30 |
Annealing | 56°C | 15 secs | 30 |
Extension | 65°C | 4 mins | 30 |
Final extension | 65°C | 4 mins | 1 |
Hold | 4°C | ∞ |
Priming and loading the SpotON Flow Cell
We recommend all new users watch the 'Priming and loading your flow cell' video before your first run.
Press down firmly on the flow cell to ensure correct thermal and electrical contact.
This step can be omitted if the flow cell has been checked previously.
See the flow cell check instructions in the MinKNOW protocol for more information.
Demo of how to use the Loading Beads.
Reagent | Volume per flow cell |
---|---|
Sequencing Buffer (SQB) | 34 µl |
Loading Beads (LB), mixed immediately before use | 25.5 µl |
Nuclease-free water | 4.5 µl |
DNA library | 11 µl |
Total | 75 µl |
We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.
Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.
Data acquisition and basecalling
For a full overview of nanopore data analysis, which includes options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.
The sequencing device control, data acquisition and real-time basecalling are carried out by the MinKNOW software. You will need to have MinKNOW already installed on your computer before starting.
For references, please refer to the MinKNOW protocol.
Instructions on how to carry out sequencing:
Downstream analysis
The wf-mpx workflow provides a basic analysis of mpox sequencing data whether targeted or metagenomics.
This workflow can be performed on commandline or through EPI2ME Labs, giving users basic QC tools, a draft consensus sequence for review, a draft assembly and SNP context.
Please note, the workflow does not perform adapter or primer trimming. This means if targeted protocols are used, there could be artefacts at these sites.
Please note that Oxford Nanopore Technologies does not currently offer a dedicated workflow for viral metagenomics analysis, meaning users will need to use or adapt third-party tools suitable for this application. However, we do offer a wf-metagenomics workflow but requires users to specify a custom database e.g. Kraken2 Metagenomic Virus Database.
For general post-basecalling data analysis options, please see the list below:
The EPI2ME platform is a cloud-based data analysis service developed by Metrichor Ltd., a subsidiary of Oxford Nanopore Technologies. The EPI2ME platform offers a range of analysis workflows, e.g. for metagenomic identification, barcoding, alignment, and structural variant calling. The FastQ WIMP (Human + Viral) workflow compares sequence reads against a comprehensive corpus of virus genome sequences. This can be used to assign sequence reads to known viruses or to provide taxonomic hints for novel virus sequences that may be detected from metagenomics samples. The analysis requires no additional equipment or compute power, and provides an easy-to-interpret report with the results. For instructions on how to run an analysis workflow in EPI2ME, please follow the instructions in the EPI2ME protocol, beginning at the "Starting data analysis" step.
For more in-depth data analysis, Oxford Nanopore Technologies offers a range of bioinformatics tutorials and workflows available in EPI2ME Labs, which are available in the EPI2ME Labs section of the Community. The platform provides a vehicle where workflows deposited in GitHub by our Research and Applications teams can be showcased with descriptive texts, functional bioinformatics code and example data.
For identifying viral sequence reads from metagenomic samples, we would recommend our wf-metagenomics workflow. This nextflow-based analysis pipeline uses the kraken2 software to assign sequence reads to an organism (or virus) of origin. The bracken software considers the kraken2 results and prepares a quantitative view of the species observed within a metagenomic sample. A few different databases suitable for this workflow are maintained by the author and are publicly available at the Kraken 2 and Bracken indexes page.
Oxford Nanopore Technologies' Research division has created a number of analysis tools, which are available in the Oxford Nanopore GitHub repository. The tools are aimed at advanced users, and contain instructions for how to install and run the software. They are provided as-is, with minimal support.
The marine phage scripts repository includes a collection of python scripts that have been used in the recovery, identification and annotation of marine virus genomes from metagenomic samples. The scripts available here could be further adapted for more general viral metagenomic requirements.
If a data analysis method for your research question is not provided in any of the resources above, please refer to the Bioinformatics section of the Resource centre. Numerous members of the Nanopore Community have developed their own tools and pipelines for analysing nanopore sequencing data, most of which are available on GitHub. Please be aware that these tools are not supported by Oxford Nanopore Technologies, and are not guaranteed to be compatible with the latest chemistry/software configuration.
Ending the experiment
Instructions for returning flow cells can be found here.
Note: All flow cells must be flushed with deionised water before returning the product.
Issues during DNA/RNA extraction and library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) | The DNA extraction method does not provide the required purity | The effects of contaminants are shown in the Contaminants document. Please try an alternative extraction method that does not result in contaminant carryover. Consider performing an additional SPRI clean-up step. |
Low RNA integrity (RNA integrity number <9.5 RIN, or the rRNA band is shown as a smear on the gel) | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. |
RNA has a shorter than expected fragment length | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. We recommend working in an RNase-free environment, and to keep your lab equipment RNase-free when working with RNA. |
Observation | Possible cause | Comments and actions |
---|---|---|
Low recovery | DNA loss due to a lower than intended AMPure beads-to-sample ratio | 1. AMPure beads settle quickly, so ensure they are well resuspended before adding them to the sample. 2. When the AMPure beads-to-sample ratio is lower than 0.4:1, DNA fragments of any size will be lost during the clean-up. |
Low recovery | DNA fragments are shorter than expected | The lower the AMPure beads-to-sample ratio, the more stringent the selection against short fragments. Please always determine the input DNA length on an agarose gel (or other gel electrophoresis methods) and then calculate the appropriate amount of AMPure beads to use. ![]() |
Low recovery after end-prep | The wash step used ethanol <70% | DNA will be eluted from the beads when using ethanol <70%. Make sure to use the correct percentage. |
Issues during the sequencing run
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | An air bubble was introduced into the nanopore array | After the Flow Cell Check it is essential to remove any air bubbles near the priming port before priming the flow cell. If not removed, the air bubble can travel to the nanopore array and irreversibly damage the nanopores that have been exposed to air. The best practice to prevent this from happening is demonstrated in this video. |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | The flow cell is not correctly inserted into the device | Stop the sequencing run, remove the flow cell from the sequencing device and insert it again, checking that the flow cell is firmly seated in the device and that it has reached the target temperature. If applicable, try a different position on the device (GridION/PromethION). |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | Contaminations in the library damaged or blocked the pores | The pore count during the Flow Cell Check is performed using the QC DNA molecules present in the flow cell storage buffer. At the start of sequencing, the library itself is used to estimate the number of active pores. Because of this, variability of about 10% in the number of pores is expected. A significantly lower pore count reported at the start of sequencing can be due to contaminants in the library that have damaged the membranes or blocked the pores. Alternative DNA/RNA extraction or purification methods may be needed to improve the purity of the input material. The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Script failed" | Restart the computer and then restart MinKNOW. If the issue persists, please collect the MinKNOW log files and contact Technical Support. If you do not have another sequencing device available, we recommend storing the flow cell and the loaded library at 4°C and contact Technical Support for further storage guidance. |
Observation | Possible cause | Comments and actions |
---|---|---|
Pore occupancy <40% | Not enough library was loaded on the flow cell | Ensure you load the recommended amount of good quality library in the relevant library prep protocol onto your flow cell. Please quantify the library before loading and calculate mols using tools like the Promega Biomath Calculator, choosing "dsDNA: µg to pmol" |
Pore occupancy close to 0 | The Ligation Sequencing Kit was used, and sequencing adapters did not ligate to the DNA | Make sure to use the NEBNext Quick Ligation Module (E6056) and Oxford Nanopore Technologies Ligation Buffer (LNB, provided in the sequencing kit) at the sequencing adapter ligation step, and use the correct amount of each reagent. A Lambda control library can be prepared to test the integrity of the third-party reagents. |
Pore occupancy close to 0 | The Ligation Sequencing Kit was used, and ethanol was used instead of LFB or SFB at the wash step after sequencing adapter ligation | Ethanol can denature the motor protein on the sequencing adapters. Make sure the LFB or SFB buffer was used after ligation of sequencing adapters. |
Pore occupancy close to 0 | No tether on the flow cell | Tethers are adding during flow cell priming (FLT/FCT tube). Make sure FLT/FCT was added to FB/FCF before priming. |
Observation | Possible cause | Comments and actions |
---|---|---|
Shorter than expected read length | Unwanted fragmentation of DNA sample | Read length reflects input DNA fragment length. Input DNA can be fragmented during extraction and library prep. 1. Please review the Extraction Methods in the Nanopore Community for best practice for extraction. 2. Visualise the input DNA fragment length distribution on an agarose gel before proceeding to the library prep. ![]() 3. During library prep, avoid pipetting and vortexing when mixing reagents. Flicking or inverting the tube is sufficient. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of unavailable pores (shown as blue in the channels panel and pore activity plot) ![]() |
Contaminants are present in the sample | Some contaminants can be cleared from the pores by the unblocking function built into MinKNOW. If this is successful, the pore status will change to "sequencing pore". If the portion of unavailable pores stays large or increases: 1. A nuclease flush using the Flow Cell Wash Kit (EXP-WSH004) can be performed, or 2. Run several cycles of PCR to try and dilute any contaminants that may be causing problems. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of inactive/unavailable pores (shown as light blue in the channels panel and pore activity plot. Pores or membranes are irreversibly damaged) | Air bubbles have been introduced into the flow cell | Air bubbles introduced through flow cell priming and library loading can irreversibly damage the pores. Watch the Priming and loading your flow cell video for best practice |
Large proportion of inactive/unavailable pores | Certain compounds co-purified with DNA | Known compounds, include polysaccharides, typically associate with plant genomic DNA. 1. Please refer to the Plant leaf DNA extraction method. 2. Clean-up using the QIAGEN PowerClean Pro kit. 3. Perform a whole genome amplification with the original gDNA sample using the QIAGEN REPLI-g kit. |
Large proportion of inactive/unavailable pores | Contaminants are present in the sample | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
Reduction in sequencing speed and q-score later into the run | For Kit 9 chemistry (e.g. SQK-LSK109), fast fuel consumption is typically seen when the flow cell is overloaded with library (please see the appropriate protocol for your DNA library to see the recommendation). | Add more fuel to the flow cell by following the instructions in the MinKNOW protocol. In future experiments, load lower amounts of library to the flow cell. |
Observation | Possible cause | Comments and actions |
---|---|---|
Temperature fluctuation | The flow cell has lost contact with the device | Check that there is a heat pad covering the metal plate on the back of the flow cell. Re-insert the flow cell and press it down to make sure the connector pins are firmly in contact with the device. If the problem persists, please contact Technical Services. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Failed to reach target temperature" | The instrument was placed in a location that is colder than normal room temperature, or a location with poor ventilation (which leads to the flow cells overheating) | MinKNOW has a default timeframe for the flow cell to reach the target temperature. Once the timeframe is exceeded, an error message will appear and the sequencing experiment will continue. However, sequencing at an incorrect temperature may lead to a decrease in throughput and lower q-scores. Please adjust the location of the sequencing device to ensure that it is placed at room temperature with good ventilation, then re-start the process in MinKNOW. Please refer to this FAQ for more information on MinION Mk 1B temperature control. |
Observation | Possible cause | Comments and actions |
---|---|---|
No input .fast5 was found or basecalled | input_path did not point to the .fast5 file location | The --input_path has to be followed by the full file path to the .fast5 files to be basecalled, and the location has to be accessible either locally or remotely through SSH. |
No input .fast5 was found or basecalled | The .fast5 files were in a subfolder at the input_path location | To allow Guppy to look into subfolders, add the --recursive flag to the command |
Observation | Possible cause | Comments and actions |
---|---|---|
No Pass or Fail folders were generated after basecalling | The --qscore_filtering flag was not included in the command | The --qscore_filtering flag enables filtering of reads into Pass and Fail folders inside the output folder, based on their strand q-score. When performing live basecalling in MinKNOW, a q-score of 7 (corresponding to a basecall accuracy of ~80%) is used to separate reads into Pass and Fail folders. |
Observation | Possible cause | Comments and actions |
---|---|---|
Unusually slow processing on a GPU computer | The --device flag wasn't included in the command | The --device flag specifies a GPU device to use for accelerate basecalling. If not included in the command, GPU will not be used. GPUs are counted from zero. An example is --device cuda:0 cuda:1, when 2 GPUs are specified to use by the Guppy command. |
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.