Version for device: Flongle
Other available device versions: MinION
Overview of the protocol
This protocol describes how to carry out sequencing of a Lambda control DNA sample using the Ligation Sequencing Kit V14 (SQK-LSK114) and the Control Expansion (EXP-CTL001).
You will need to:
The table below is an overview of the steps required in the library preparation, including timings and stopping points.
Library preparation step | Process | Time | Stop option |
---|---|---|---|
DNA repair and end-prep | Repair the DNA, prepare the DNA ends for adapter attachment, and perform a bead clean-up. | 35 minutes | 4°C overview |
Adapter ligation and clean-up | Attach the sequencing adapters to the DNA ends and perform a bead clean-up | 20 minutes | 4°C short-term shortage or for repeated use, such as re-loading your flow cell. -80°C for single-use long-term storage. We strongly recommend sequencing your library as soon as it is adapted. |
Priming and loading the flow cell | Prime the flow cell and load the prepared library for sequencing | 5 minutes | - |
Sequencing and analysis
You will need to:
This protocol should only be used in combination with:
At Oxford Nanopore we look to continuously improve our production processes to deliver a more robust product. In the case of Flongle, we are seeing the stability of the flow cells we ship improve. However for a small number of flow cells, upon loading, the flow cell rapidly deteriorates. This can be seen as saturation in the MinKNOW GUI. We are working hard to resolve this, however in the meantime we suggest the following loading recommendations and to use the buffers from the Flongle Sequencing Expansion (EXP-FSE002) shipped with your Flongle flow cells. If you do see rapid deteriorate/saturation on your flow cell, please contact support@nanoporetech.com for assistance.
Loading recommendations
Following standard input recommendations, the protocol should produce enough final library (adapted DNA in EB) to load at least two Flongle flow cells. We recommend reserving enough library to load a second Flongle flow cell should you need to generate more data from a second flongle flow cell.
There are three buffers that come into direct contact with a flow cell at point of loading (SB: Sequencing Buffer, FCF: Flow Cell Flush and LIB: Library Beads or LIS: Library Solution). When looking at these buffers, we found that there are a very low level of contaminants seeping out of the plastic vials that impacts the robustness of the Flongle flow cell system (MinION and PromethION are not impacted by this).
We have found that when storing these buffers in glass vials instead of plastic, incidence of deterioration is reduced.
To rapidly deploy this to Flongle users, we have produced a Flongle Sequencing Expansion (EXP-FSE002) with these three components in glass vials, which can perform 12 Flongle flow cell loads in total.
To load a library onto your Flongle flow cell, you will need to use the following components:
Flongle Sequencing Expansion (EXP-FSE002) components
- Sequencing Buffer (SB)
- Flow Cell Flush (FCF)
- Library Beads (LIB) or Library Solution (LIS)
Sequencing Kit components
- Flow Cell Tether (FCT)
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
We recommend buying the NEBNext® Companion Module v2 for Oxford Nanopore Technologies® Ligation Sequencing (catalogue number E7672S or E7672L), which contains all the NEB reagents needed for use with the Ligation Sequencing Kit.
The previous version, NEBNext® Companion Module for Oxford Nanopore Technologies® Ligation Sequencing (NEB, E7180S or E7180L) is compatible, but the recommended v2 module offers more efficient dA-tailing and ligation, a result of the FFPEv2 DNA Repair Buffer and Salt-T4 DNA Ligase, respectively. A marked cost saving per sample preparation is also realised when using the v2 module.
Note: for our amplicon protocols, NEBNext FFPE DNA Repair Mix is not required and purchasing the required reagents separately is more cost effective.
We have validated and recommend the use of all the third-party reagents used in this protocol. Alternatives have not been tested by Oxford Nanopore Technologies.
For all third-party reagents, we recommend following the manufacturer's instructions to prepare the reagents for use.
We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
Note: We are in the process of reformatting our kits with single-use tubes into a bottle format.
Single-use tubes format:
Bottle format:
Note: This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Note: The DNA Control Sample (DCS) is a 3.6 kb standard amplicon mapping the 3' end of the Lambda genome.
Name | Acronym | Cap colour | Number of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Sequencing Buffer | SB | Blue | 1 | 250 |
Library Beads | LIB | Blue | 1 | 200 |
Library Solution | LIS | Blue | 1 | 200 |
Flow Cell Flush | FCF | Blue | 1 | 1,600 |
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
Note: The DNA Control Sample (DCS) is a 3.6 kb standard amplicon mapping the 3' end of the Lambda genome. Lambda DNA (LMD) fragment length is 48.5 kb.
DNA repair and end-prep
The previous version, NEBNext® Companion Module for Oxford Nanopore Technologies® Ligation Sequencing (NEB, E7180S or E7180L) is also compatible, but the recommended v2 module offers more efficient dA-tailing and ligation.
We recommend performing a flow cell check before starting your library prep to ensure you have a flow cell with enough pores for a good sequencing run.
See the flow cell check instructions in the MinKNOW protocol for more information.
At Oxford Nanopore we look to continuously improve our production processes to deliver a more robust product. In the case of Flongle, we are seeing the stability of the flow cells we ship improve. However for a small number of flow cells, upon loading, the flow cell rapidly deteriorates. This can be seen as saturation in the MinKNOW GUI. We are working hard to resolve this, however in the meantime we suggest the following loading recommendations and to use the buffers from the Flongle Sequencing Expansion (EXP-FSE002) shipped with your Flongle flow cells. If you do see rapid deteriorate/saturation on your flow cell, please contact support@nanoporetech.com for assistance.
Loading recommendations
Following standard input recommendations, the protocol should produce enough final library (adapted DNA in EB) to load at least two Flongle flow cells. We recommend reserving enough library to load a second Flongle flow cell should you need to generate more data from a second flongle flow cell.
There are three buffers that come into direct contact with a flow cell at point of loading (SB: Sequencing Buffer, FCF: Flow Cell Flush and LIB: Library Beads or LIS: Library Solution). When looking at these buffers, we found that there are a very low level of contaminants seeping out of the plastic vials that impacts the robustness of the Flongle flow cell system (MinION and PromethION are not impacted by this).
We have found that when storing these buffers in glass vials instead of plastic, incidence of deterioration is reduced.
To rapidly deploy this to Flongle users, we have produced a Flongle Sequencing Expansion (EXP-FSE002) with these three components in glass vials, which can perform 12 Flongle flow cell loads in total.
To load a library onto your Flongle flow cell, you will need to use the following components:
Flongle Sequencing Expansion (EXP-FSE002) components
- Sequencing Buffer (SB)
- Flow Cell Flush (FCF)
- Library Beads (LIB) or Library Solution (LIS)
Sequencing Kit components
- Flow Cell Tether (FCT)
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
By default, the protocol contains no DNA fragmentation step, however in some cases it may be advantageous to fragment your sample. For example, when working with lower amounts of input gDNA (25 ng–125 ng), fragmentation will increase the number of DNA molecules and therefore increase throughput. Instructions are available in the DNA Fragmentation section of Extraction methods.
Additionally, we offer several options for size-selecting your DNA sample to enrich for long fragments - instructions are available in the Size Selection section of Extraction methods.
For optimal performance, NEB recommend the following:
Thaw all reagents on ice.
Flick and/or invert the reagent tubes to ensure they are well mixed.
Note: Do not vortex the FFPE DNA Repair Mix or Ultra II End Prep Enzyme Mix.
Always spin down tubes before opening for the first time each day.
Vortex the FFPE DNA Repair Buffer v2, or the NEBNext FFPE DNA Repair Buffer and Ultra II End Prep Reaction Buffer to ensure they are well mixed.
Note: These buffers may contain a white precipitate. If this occurs, allow the mixture(s) to come to room temperature and pipette the buffer several times to break up the precipitate, followed by a quick vortex to mix.
The FFPE DNA Repair Buffer may have a yellow tinge and is fine to use if yellow.
Between each addition, pipette mix 10-20 times.
Reagent | Volume |
---|---|
Nuclease-free water | 13.5 µl |
Lambda DNA (50 µg/ml) | 10 µl |
DNA CS (optional) | 0.5 µl |
NEBNext FFPE DNA Repair Buffer v2 | 3.5 µl |
NEBNext FFPE DNA Repair Mix | 1 µl |
Ultra II End-prep Enzyme Mix | 1.5 µl |
Total | 30 µl |
If using the previous version of the NEBNext® Companion Module for Oxford Nanopore Technologies® Ligation Sequencing (NEB, E7180S or E7180L):
Between each addition, pipette mix 10-20 times
Reagent | Volume |
---|---|
Nuclease-free water | 13.5 µl |
Lambda DNA (50 µg/ml) | 10 µl |
DNA Control Sample (DCS) | 0.5 µl |
NEBNext FFPE DNA Repair Buffer | 1.75 µl |
NEBNext FFPE DNA Repair Mix | 1 µl |
Ultra II End-Prep Reaction Buffer | 1.75 µl |
Ultra II Endp-prep Enzyme Mix | 1.5 µl |
Total | 30 µl |
Adapter ligation and clean-up
Salt-T4® DNA Ligase (NEB, M0467) can be bought separately or is provided in the NEBNext® Companion Module v2 for Oxford Nanopore Technologies® Ligation Sequencing (catalogue number E7672S or E7672L).
The Quick T4 DNA Ligase (NEB, E6057) available in the previous version NEBNext® Companion Module for Oxford Nanopore Technologies® Ligation Sequencing (NEB, E7180S or E7180L) is also compatible, but the new recommended reagent offers more efficient and ligation.
Between each addition, pipette mix 10-20 times.
Reagent | Volume |
---|---|
DNA sample from the previous step | 30 µl |
Ligation Adapter (LA) | 2.5 µl |
Ligation Buffer (LNB) | 12.5 µl |
Salt-T4® DNA Ligase | 5 µl |
Total | 50 µl |
Dispose of the pelleted beads.
If required, we recommend using a mass to mol calculator such as the NEB calculator.
Following standard input recommendations, the protocol should produce enough final library (adapter DNA in EB) to load at least two Flongle flow cells. We recommend reserving enough library to load onto a second flow cell. Loading more than 10 fmol can have a detrimental effect on output. Dilute the library in EB or nuclease-free water to a final volume of 5 μl.
We recommend storing libraries in Eppendorf DNA LoBind tubes at 4°C for short-term storage or repeated use, for example, re-loading flow cells between washes.
For single use and long-term storage of more than 3 months, we recommend storing libraries at -80°C in Eppendorf DNA LoBind tubes.
Flongle Flow Cell loading
There are three buffers that come into direct contact with a flow cell at point of loading (SB: Sequencing Buffer, FCF: Flow Cell Flush and LIB: Library Beads or LIS: Library Solution). When looking at these buffers, we found that there are a very low level of contaminants seeping out of the plastic vials that impacts the robustness of the Flongle flow cell system (MinION and PromethION are not impacted by this).
We have found that when storing these buffers in glass vials instead of plastic, incidence of deterioration is reduced.
To rapidly deploy this to Flongle users, we have produced a Flongle Sequencing Expansion (EXP-FSE002) with these three components in glass vials, which can perform 12 Flongle flow cell loads in total.
To load a library onto your Flongle flow cell, you will need to use the following components:
Flongle Sequencing Expansion (EXP-FSE002) components
- Sequencing Buffer (SB)
- Flow Cell Flush (FCF)
- Library Beads (LIB) or Library Solution (LIS)
Sequencing Kit components
- Flow Cell Tether (FCT)
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
The seal tab, air vent, waste channel, drain port and sample port are visible here. The sample port, drain port and air vent only become accessible once the seal tab is peeled back.
The adapter should sit evenly and flat on the MinION Mk1B or GridION platform. This ensures the flow cell assembly is flat during the next stage.
The flow cell should sit evenly and flat inside the adapter, to avoid any bubbles forming inside the fluidic compartments.
Lift up the seal tab:
Pull the seal tab to open access to the sample port:
Hold the seal tab open by using adhesive on the tab to stick to the MinION Mk 1B lid:
We recommend using the Library Beads (LIB) for most sequencing experiments. However, the Library Solution (LIS) is available for more viscous libraries.
Reagents | Volume |
---|---|
Sequencing Buffer (SB) | 15 µl |
Library Beads (LIB) mixed immediately before use, or Library Solution (LIS), if using. | 10 µl |
DNA library | 5 µl |
Total | 30 µl |
Stick the transparent adhesive tape to the sample port.
Replace the top (Wheel icon section) of the seal tab to its original position.
Data acquisition and basecalling
Once you have loaded your flow cell, the sequencing run can be started on MinKNOW, our sequencing software that controls the device, data acquisition and real-time basecalling. For more detailed information on setting up and using MinKNOW, please see the MinKNOW protocol.
MinKNOW can be used and set up to sequence in multiple ways:
For more information on using MinKNOW on a sequencing device, please see the device user manuals:
To start a sequencing run on MinKNOW:
1. Navigate to the start page and click Start sequencing.
2. Fill in your experiment details, such as name and flow cell position and sample ID.
3. Select the sequencing kit used in the library preparation on the Kit page.
4. Configure the sequencing parameters for your sequencing run or keep to the default settings on the Run options and Analysis tabs.
Note: If basecalling was turned off when a sequencing run was set up, basecalling can be performed post-run on MinKNOW. For more information, please see the MinKNOW protocol.
5. On the Output page, set up the output parameters or keep to the default settings.
6. Click Start on the Review page to start the sequencing run.
Kit 14 chemistry has improved duplex basecalling which requires rebasecalling on Dorado after simplex basecalling has been performed on MinKNOW.
For detailed information on setting up your sequencing run, for both simplex and duplex basecalling, please see the Kit 14 sequencing and duplex basecalling info sheet.
Note: When running Dorado, we recommend stopping other basecalling for the best performance by maximising memory available to Dorado. This can be stopped and restarted when Dorado has finished via the GUI on MinKNOW.
After sequencing and basecalling, the data can be analysed. For further information about options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.
In the Downstream analysis section, we outline further options for analysing your data.
Issues during DNA/RNA extraction and library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) | The DNA extraction method does not provide the required purity | The effects of contaminants are shown in the Contaminants document. Please try an alternative extraction method that does not result in contaminant carryover. Consider performing an additional SPRI clean-up step. |
Low RNA integrity (RNA integrity number <9.5 RIN, or the rRNA band is shown as a smear on the gel) | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. |
RNA has a shorter than expected fragment length | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. We recommend working in an RNase-free environment, and to keep your lab equipment RNase-free when working with RNA. |
Observation | Possible cause | Comments and actions |
---|---|---|
Low recovery | DNA loss due to a lower than intended AMPure beads-to-sample ratio | 1. AMPure beads settle quickly, so ensure they are well resuspended before adding them to the sample. 2. When the AMPure beads-to-sample ratio is lower than 0.4:1, DNA fragments of any size will be lost during the clean-up. |
Low recovery | DNA fragments are shorter than expected | The lower the AMPure beads-to-sample ratio, the more stringent the selection against short fragments. Please always determine the input DNA length on an agarose gel (or other gel electrophoresis methods) and then calculate the appropriate amount of AMPure beads to use. ![]() |
Low recovery after end-prep | The wash step used ethanol <70% | DNA will be eluted from the beads when using ethanol <70%. Make sure to use the correct percentage. |
Issues during the sequencing run
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | An air bubble was introduced into the nanopore array | After the Flow Cell Check it is essential to remove any air bubbles near the priming port before priming the flow cell. If not removed, the air bubble can travel to the nanopore array and irreversibly damage the nanopores that have been exposed to air. The best practice to prevent this from happening is demonstrated in this video. |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | The flow cell is not correctly inserted into the device | Stop the sequencing run, remove the flow cell from the sequencing device and insert it again, checking that the flow cell is firmly seated in the device and that it has reached the target temperature. If applicable, try a different position on the device (GridION/PromethION). |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | Contaminations in the library damaged or blocked the pores | The pore count during the Flow Cell Check is performed using the QC DNA molecules present in the flow cell storage buffer. At the start of sequencing, the library itself is used to estimate the number of active pores. Because of this, variability of about 10% in the number of pores is expected. A significantly lower pore count reported at the start of sequencing can be due to contaminants in the library that have damaged the membranes or blocked the pores. Alternative DNA/RNA extraction or purification methods may be needed to improve the purity of the input material. The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Script failed" | Restart the computer and then restart MinKNOW. If the issue persists, please collect the MinKNOW log files and contact Technical Support. If you do not have another sequencing device available, we recommend storing the flow cell and the loaded library at 4°C and contact Technical Support for further storage guidance. |
Observation | Possible cause | Comments and actions |
---|---|---|
Pore occupancy <40% | Not enough library was loaded on the flow cell | Ensure the correct volume and concentration as stated on the appropriate protocol for your sequencing library is loaded onto the flow cell. Please quantify the library before loading and calculate fmols using tools like the Promega Biomath Calculator, choosing "dsDNA: µg to fmol" |
Pore occupancy close to 0 | The Ligation Sequencing Kit was used, and sequencing adapters did not ligate to the DNA | Make sure to use the NEBNext Quick Ligation Module (E6056) and Oxford Nanopore Technologies Ligation Buffer (LNB, provided in the sequencing kit) at the sequencing adapter ligation step, and use the correct amount of each reagent. A Lambda control library can be prepared to test the integrity of the third-party reagents. |
Pore occupancy close to 0 | The Ligation Sequencing Kit was used, and ethanol was used instead of LFB or SFB at the wash step after sequencing adapter ligation | Ethanol can denature the motor protein on the sequencing adapters. Make sure the LFB or SFB buffer was used after ligation of sequencing adapters. |
Pore occupancy close to 0 | No tether on the flow cell | Tethers are adding during flow cell priming (FLT tube for Kit 9, 10, 11, FCT for Kit 14, and FTU for ultra-long DNA kits). Make sure FLT/FCT/FTU was added to the buffer (FB for Kit 9, 10, 11, and FCF for Kit 14) before priming. |
Observation | Possible cause | Comments and actions |
---|---|---|
Shorter than expected read length | Unwanted fragmentation of DNA sample | Read length reflects input DNA fragment length. Input DNA can be fragmented during extraction and library prep. 1. Please review the Extraction Methods in the Nanopore Community for best practice for extraction. 2. Visualise the input DNA fragment length distribution on an agarose gel before proceeding to the library prep. ![]() 3. During library prep, avoid pipetting and vortexing when mixing reagents. Flicking or inverting the tube is sufficient. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of unavailable pores (shown as blue in the channels panel and pore activity plot) ![]() |
Contaminants are present in the sample | Some contaminants can be cleared from the pores by the unblocking function built into MinKNOW. If this is successful, the pore status will change to "sequencing pore". If the portion of unavailable pores stays large or increases: 1. A nuclease flush using the Flow Cell Wash Kit (EXP-WSH004) can be performed, or 2. Run several cycles of PCR to try and dilute any contaminants that may be causing problems. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of inactive/unavailable pores (shown as light blue in the channels panel and pore activity plot. Pores or membranes are irreversibly damaged) | Air bubbles have been introduced into the flow cell | Air bubbles introduced through flow cell priming and library loading can irreversibly damage the pores. Watch the Priming and loading your flow cell video for best practice |
Large proportion of inactive/unavailable pores | Certain compounds co-purified with DNA | Known compounds, include polysaccharides, typically associate with plant genomic DNA. 1. Please refer to the Plant leaf DNA extraction method. 2. Clean-up using the QIAGEN PowerClean Pro kit. 3. Perform a whole genome amplification with the original gDNA sample using the QIAGEN REPLI-g kit. |
Large proportion of inactive/unavailable pores | Contaminants are present in the sample | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
Temperature fluctuation | The flow cell has lost contact with the device | Check that there is a heat pad covering the metal plate on the back of the flow cell. Re-insert the flow cell and press it down to make sure the connector pins are firmly in contact with the device. If the problem persists, please contact Technical Services. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Failed to reach target temperature" | The instrument was placed in a location that is colder than normal room temperature, or a location with poor ventilation (which leads to the flow cells overheating) | MinKNOW has a default timeframe for the flow cell to reach the target temperature. Once the timeframe is exceeded, an error message will appear and the sequencing experiment will continue. However, sequencing at an incorrect temperature may lead to a decrease in throughput and lower q-scores. Please adjust the location of the sequencing device to ensure that it is placed at room temperature with good ventilation, then re-start the process in MinKNOW. Please refer to this link for more information on MinION temperature control. |
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.