Version for device: MinION
Overview of the protocol
For more information about our Early Access programmes, please see this article on product release phases.
Please ensure you always use the most recent version of the protocol.
This kit is recommended for users who:
This protocol describes the complete workflow from extracting gDNA from frozen tissue or purified cells from whole blood to the sequencing of ultra-high molecular weight (uHMW) gDNA using the Ultra-Long DNA Sequencing Kit (SQK-ULK114). We have also included the procedure to isolate white blood cells (WBCs) from whole blood and how to quantify gDNA developed by Paul A ‘Giron’ Koetsier & Eric J Cantor, 2021.
We have used the NEB Monarch® HMW DNA Extraction Kit for Tissue (cat # T3060) to extract the uHMW gDNA for both input types when developing this protocol. Alternative kits are available from NEB which are specifically designed for the extraction from blood and cells. However, they have not been validated by Oxford Nanopore Technologies.
Per reaction, there is enough library generated for multiple consecutive loads onto a flow cell to increase output. To load a library six times on a MinION/GridION flow cell, a flow cell wash is required to recover channels.
Steps in the sequencing workflow:
Prepare for your experiment
You will need to:
Library preparation
You will need to:
Sequencing and analysis
You will need to:
The Ultra-Long DNA Sequencing Kit (SQK-ULK114) protocol generates viscous DNA which can affect flow cell loading. We have modified the flow cell loading steps to take account for this. Please take care and follow the steps carefully to avoid damaging the flow cell.
To increase output, we recommend loading an ultra-long library three times per flow cell. A flow cell wash using the Flow Cell Wash Kit (EXP-WSH004) is required between each subsequent library load to recover channels. To run a second library straight away, please follow the modified method in this protocol: To run another library of ultra-long DNA on a MinION/GridION flow cell straight away.
When mixing, we recommend using wide-bore pipette tips to mix the full volume of a sample to ensure thorough mixing whilst minimising mechanical shearing of long fragments.
To preserve longer DNA, mix slower and more gently. Vortexing on low speeds may also be used at the expense of very long fragments.
While precautions should be taken to ensure that DNA fragment lengths are preserved, there should be no compromise to ensuring that reagents are thoroughly mixed with DNA. Insufficient mixing will lead to reduced read length and output.
For further information, please refer to the troubleshooting section.
This protocol should only be used in combination with:
Equipment and consumables
This protocol has been developed using the NEB Monarch® HMW DNA Extraction Kit for Tissue (cat # T3060). Alternative kits are available from NEB which are specifically designed for the extraction from blood and cells. However, they have not been validated by Oxford Nanopore Technologies.
This method has been validated for use on the following inputs:
Name | Acronym | Cap colour | Number of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 1 | 40 |
Fragmentation Mix | FRA | Amber | 1 | 50 |
FRA Dilution Buffer | FDB | Clear | 1 | 1,600 |
Elution Buffer | EB | Black | 2 | 1,500 |
Extraction EB | EEB | Orange | 3 | 1,700 |
Sequencing Buffer UL | SBU | Red | 2 | 1,000 |
Loading Solution UL | LSU | White cap, pink label | 1 | 200 |
Flush Tether UL | FTU | Purple | 1 | 600 |
Flow Cell Flush | FCF | Blue | 2 | 15,500 |
Precipitation Buffer | PTB | Blue | 2 | 1,700 |
Precipitation Star | PS | Yellow | 6 | 1 star |
Contents | Volume (µl) | No. of tubes | No. of uses |
---|---|---|---|
Wash Mix (WMX) | 15 | 1 | 6 |
Wash Diluent (DIL) | 1,300 | 2 | 6 |
Storage Buffer (S) | 1,600 | 2 | 6 |
These expansions provide extra library preparation and flow cell priming reagents to allow users to maximise the use out of their Ultra-Long DNA Sequencing Kit V14.
The EEB Expansion (EXP-EEB001) contains enough reagents for at least 6 standard extraction elution steps.
The Ultra-Long Auxiliary Vials (EXP-ULA001) provides enough reagents to carry out twelve additional flow cell loads on MinION or PromethION flow cells.
EEB Expansion (EXP-EEB001) contents:
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (μl) |
---|---|---|---|---|
Extraction EB | EEB | White | 1 | 6,000 |
Ultra-Long Auxiliary Vials (EXP-ULA001) contents:
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (μl) |
---|---|---|---|---|
Elution Buffer | EB | Black | 1 | 1,500 |
Sequencing Buffer UL | SBU | Red | 2 | 1,000 |
Loading Solution UL | LSU | White cap, pink label | 1 | 200 |
Flush Tether UL | FTU | Purple | 1 | 600 |
Flow Cell Flush | FCF | Clear cap, light blue label | 1 | 15,500 |
Computer requirements and software
Sequencing on a MinION Mk1B requires a high-spec computer or laptop to keep up with the rate of data acquisition. Read more in the MinION Mk1B IT Requirements document.
The MinION Mk1C contains fully-integrated compute and screen, removing the need for any accessories to generate and analyse nanopore data. Read more in the MinION Mk1C IT requirements document.
The MinKNOW software controls the nanopore sequencing device, collects sequencing data and basecalls in real time. You will be using MinKNOW for every sequencing experiment to sequence, basecall and demultiplex if your samples were barcoded.
For instructions on how to run the MinKNOW software, please refer to the MinKNOW protocol.
The EPI2ME cloud-based platform performs further analysis of basecalled data, for example alignment to the Lambda genome, barcoding, or taxonomic classification. You will use the EPI2ME platform only if you would like further analysis of your data post-basecalling.
For instructions on how to create an EPI2ME account and install the EPI2ME Desktop Agent, please refer to the EPI2ME Platform protocol.
We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within three months of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
Isolation of white blood cells (WBCs) from whole blood
Approximately 6 million isolated white blood cells must be prepared from 1.6 ml of whole blood to use as input in the Ultra-long DNA experiment.
Users may isolate white blood cells by any means they feel are most appropriate for the whole blood sample to be used. If 6 million cells have been isolated, users can start from the uHMW gDNA extraction step.
Preparation of tissue samples for gDNA extraction
Reagent | Stock | Final conc. | Volume |
---|---|---|---|
Tris.HCl, pH 8 | 1 M | 0.05 M | 50 ml |
EDTA | 0.5 M | 0.1 M | 200 ml |
Sucrose | 2.5 M | 0.35 M | 140 ml |
Nuclease-free water | - | - | 610 ml |
Total | - | - | 1000 ml |
The homogenate can be gently agitated, and a small amount of negative pressure can be applied with the syringe to help pass the homogenate through the strainer.
Transfer any intact tissue caught by the 200 µm pluriStrainer® into a fresh 50 ml Falcon tube by inverting the strainer and tapping out the intact tissue.
Tip: A spatula can be used to help remove the intact tissue from the strainer.
Add 10 ml of the Cell Suspension Buffer (CSB) into the 50 ml Falcon tube.
Assemble the pluriStrainer apparatus with a 100 µm strainer, connector ring, funnel and 50 ml Falcon tube according to the manufacturer’s instructions.
Pass the full volume of the 200 µm strained homogenate through the 100 µm PluriStrainer®.
Tip: The homogenate can be gently agitated, and a small amount of negative pressure can be applied with the syringe to help pass the homogenate through the strainer.
Disassemble the pluriStrainer® and retain the 100 µm strained homogenate in the 50 ml Falcon tube.
Assemble the pluriStrainer apparatus with a 50 µm strainer, connector ring, funnel and 50 ml Falcon tube according to the manufacturer’s instructions.
Pass the full volume of the 100 µm strained homogenate through the 50 µm PluriStrainer®.
Tip: The homogenate can be gently agitated, and a small amount of negative pressure can be applied with the syringe to help pass the homogenate through the strainer.
Disassemble the pluriStrainer® and retain the 50 µm strained homogenate in the 50 ml Falcon tube.
Assemble the pluriStrainer apparatus with a 30 µm strainer, connector ring, funnel and 50 ml Falcon tube according to the manufacturer’s instructions.
Pass the full volume of the 50 µm strained homogenate through the 30 µm PluriStrainer®.
Tip: The homogenate can be gently agitated, and a small amount of negative pressure can be applied with the syringe to help pass the homogenate through the strainer.
Disassemble the pluriStrainer® and retain the 30 µm strained homogenate in the 50 ml Falcon tube.
Note: You may need to flick quite hard and thoroughly to ensure the pellet breaks up and no clumps remain.
uHMW gDNA extraction
Reagent | Volume |
---|---|
Monarch HMW gDNA Tissue Lysis Buffer | 1,800 µl |
Proteinase K | 60 µl |
Total | 1860 µl |
Note: When using cell lines, we have found that this step can be omitted.
DNA will be present in the upper phase, whereas protein and other contaminants will be in the lower phase.
The DNA in the upper phase should be extremely viscous and should only be possible to aspirate using a wide-bore pipette tip.
Note: The first bead is a sacrificial bead and will remain at the bottom of the tube throughout the remainder of the process.
Note: if ~100 µl of supernatant is remaining in the tube, perfomance will not be affected.
Ensure ethanol is added to the Monarch gDNA Wash Buffer as per kit guidance.
Thorough but gentle resuspension of DNA is required to prevent heterogeneity in the sample.
(Optional) gDNA quantification
The method to quantify uHMW gDNA was developed by Paul A ‘Giron’ Koetsier & Eric J Cantor, 2021, which recommends the use of a regular P200 pipette and tip.
This optional uHMW gDNA quantification step has also been included in the protocol for user QC. However, this step can be omitted and 750 µl of DNA in Extraction EB (EEB) can be taken straight into the tagmentation step of the protocol.
Tagmentation
When mixing, we recommend using wide-bore pipette tips to mix the full volume of a sample to ensure thorough mixing whilst minimising mechanical shearing of long fragments.
To preserve longer DNA, mix slower and more gently. Vortexing on low speeds may also be used at the expense of very long fragments.
While precautions should be taken to ensure that DNA fragment lengths are preserved, there should be no compromise to ensuring that reagents are thoroughly mixed with DNA. Insufficient mixing will lead to reduced read length and output.
For further information, please refer to the troubleshooting section.
Once thawed, keep all the kit components on ice.
Reagent | Thaw at room temperature | Briefly spin down | Mix well by pipetting |
---|---|---|---|
Fragmentation Mix (FRA) | Not frozen | ✓ | ✓ |
FRA dilution buffer (FDB) | Not frozen | ✓ | ✓ |
Rapid Adapter (RA) | Not frozen | ✓ | ✓ |
Reagent | Volume |
---|---|
Fragmentation Mix (FRA) | 6 µl |
FRA dilution buffer (FDB) | 244 µl |
Total | 250 µl |
Visually check the reagents are thoroughly mixed. It is important to immediately mix the diluted Fragmentation Mix (FRA) with the DNA thoroughly.
Temperature | Time |
---|---|
Room temperature | 10 minutes |
75°C | 10 minutes |
On ice | Cool on ice for a minimum of 10 minutes |
Note: the reaction must be cooled on ice before adding Rapid Adapter (RA) to prevent denaturing the enzyme.
Note: visually check to ensure the reaction is thoroughly mixed.
Clean-up
The method has been temporarily updated to omit the use of the Precipitation Star (PS) in the Clean-up step of this protocol. This is due to observations that using the Precipitation Star (PS) can lead to, in some cases, immediate pore loss after loading the ultra-long DNA library.
While we continue to investigate, this solution will prevent potential issues and has no negative impact on flow cell output, read length or pore occupancy.
For more information please visit this post in the Nanopore Community.
Reagent | Thaw at room temperature | Briefly spin down | Mix well by pipetting |
---|---|---|---|
Precipitation buffer (PTB) | ✓ | ✓ | ✓ |
Elution Buffer (EB) | ✓ | ✓ | ✓ |
Once thawed, keep all the kit components on ice.
Visually inspect to check the DNA has precipitated, forming a glassy white mass.
Thorough but gentle resuspension of DNA is required to prevent heterogeneity in the sample.
We recommend storing libraries in Eppendorf DNA LoBind tubes at 4°C for short-term storage or repeated use, for example, re-loading flow cells between washes.
For single use and long-term storage of more than 3 months, we recommend storing libraries at -80°C in Eppendorf DNA LoBind tubes.
Priming and loading the SpotON flow cell
Only use the reagents provided with the SQK-ULK114 kit for priming and loading the flow cell.
Reagents from other kits are not compatible with this protocol.
We recommend all new users watch the 'Priming and loading your flow cell' video before your first run.
To ensure optimal interactions and access to the flow cell ports, please ensure the MinION Flow Cell Light Shield is not installed on your flow cell until after loading your DNA library.
If the MinION Flow Cell Light Shield has already been installed, remove it from the flow cell and store until required later in the protocol.
Reagent | Volume per flow cell |
---|---|
Sequencing Buffer UL (SBU) | 37.5 µl |
Loading Solution UL (LSU) | 3.7 µl |
DNA library | 33.8 µl |
Total | 75 µl |
Note: Ensure the Sequencing Buffer UL (SBU) and Loading Solution UL (LSU) are thoroughly mixed by pipetting before the addition of the DNA library.
This step can be omitted if the flow cell has been checked previously.
See the flow cell check instructions in the MinKNOW protocol for more information.
For optimal sequencing performance and improved output on R10.4.1 flow cells (FLO-MIN114), we recommend adding Bovine Serum Albumin (BSA) to the flow cell priming mix at a final concentration of 0.2 mg/ml. We do not recommend using recombinant BSA.
Reagent | Volume |
---|---|
Bovine Serum Albumin (BSA) at 50 mg/ml | 5 µl |
Flush Tether UL (FTU) | 30 µl |
Flow Cell Flush (FCF) | 1170 µl |
Total | 1205 µl |
If the MinION Flow Cell Light Shield has already been installed, remove it from the flow cell and store until required later in the protocol.
Take care not to place the pipette tip directly onto/into the SpotON port as this could damage the array.
Allow the DNA library to flow through the SpotON port by waiting up to two minutes.
If the DNA library does not enter the SpotON port, apply negative pressure in the flow cell as explained further below.
Note: Take care to not apply too much negative pressure too quickly to avoid bringing air bubbles into the flow cell. Air bubbles will cause irreversible damage to the flow cell.
We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.
Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.
A nuclease wash using the Flow Cell Wash Kit (EXP-WSH004) is required between each subsequent library load to recover channels and maximise sequencing output.
For MinION/GridION flow cells, there is enough library generated for six consecutive loads per reaction, using 33.8 µl of fresh library combined with 37.5 µl of Sequencing Buffer (SBU) and 3.7 µl of Loading Solution (LSU) before re-loading for further sequencing.
Please follow Flushing a MinION/GridION Flow Cell in the Flow Cell Wash Kit protocol for the nuclease wash instructions. To run another library straight away, follow the modified method: Reloading ultra-long DNA library on a MinION/GridION flow cell.
Reloading ultra-long DNA library on a MinION/GridION flow cell
Follow the instructions in the Flow Cell Wash Kit (EXP-WSH004) for MinION/GridION protocol.
If the flow cell is to be washed and stored, the light shield can be removed.
Reagent | Volume per flow cell |
---|---|
Sequencing Buffer UL (SBU) | 37.5 µl |
Loading Solution UL (LSU) | 3.7 µl |
DNA library | 33.8 µl |
Total | 75 µl |
Note: Ensure the Sequencing Buffer UL (SBU) and Loading Solution UL (LSU) are thoroughly mixed by pipetting before the addition of the DNA library.
Note: We do not recommend using any other albumin type (e.g. recombinant human serum albumin).
Reagent | Volume |
---|---|
Bovine Serum Albumin (BSA) at 50 mg/ml | 5 µl |
Flush Tether UL (FTU) | 30 µl |
Flow Cell Flush (FCF) | 1170 µl |
Total | 1205 µl |
Note: As both the priming port and SpotON sample port are closed, no fluid should leave the sensor array area.
Take care not to place the pipette tip directly onto/into the SpotON port as this could damage the array.
Allow the DNA library to flow through the SpotON port by waiting up to two minutes.
If the DNA library does not enter the SpotON port, apply negative pressure in the flow cell as explained further below.
Note: Take care to not apply too much negative pressure too quickly to avoid bringing air bubbles into the flow cell. Air bubbles will cause irreversible damage to the flow cell.
We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.
Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.
To resume sequencing run, navigate to the Experiments page, click 'Resume' and select flow cell position.
To manually trigger a channel scan, click 'Start pore scan' and select flow cell position.
For further information, please see the MinKNOW protocol.
Resume run:
Pore scan:
Data acquisition and basecalling
Once you have loaded your flow cell, the sequencing run can be started on MinKNOW, our sequencing software that controls the device, data acquisition and real-time basecalling. For more detailed information on setting up and using MinKNOW, please see the MinKNOW protocol.
MinKNOW can be used and set up to sequence in multiple ways:
For more information on using MinKNOW on a sequencing device, please see the device user manuals:
To start a sequencing run on MinKNOW:
1. Navigate to the start page and click Start sequencing.
2. Fill in your experiment details, such as name and flow cell position and sample ID.
3. Select the sequencing kit used in the library preparation on the Kit page.
4. Configure the sequencing parameters for your sequencing run or keep to the default settings on the Run options and Analysis tabs.
Note: If basecalling was turned off when a sequencing run was set up, basecalling can be performed post-run on MinKNOW. For more information, please see the MinKNOW protocol.
5. On the Output page, set up the output parameters or keep to the default settings.
6. Click Start on the Review page to start the sequencing run.
After sequencing has completed on MinKNOW, the flow cell can be reused or returned, as outlined in the Flow cell reuse and returns section.
After sequencing and basecalling, the data can be analysed. For further information about options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.
In the Downstream analysis section, we outline further options for analysing your data.
Flow cell reuse and returns
The Flow Cell Wash Kit protocol is available on the Nanopore Community.
Instructions for returning flow cells can be found here.
Note: All flow cells must be flushed with deionised water before returning the product.
Downstream analysis
There are several options for further analysing your basecalled data:
For in-depth data analysis, Oxford Nanopore Technologies offers a range of bioinformatics tutorials and workflows available in EPI2ME. The platform provides a vehicle where workflows deposited in GitHub by our Research and Applications teams can be showcased with descriptive texts, functional bioinformatics code and example data.
Oxford Nanopore Technologies' Research division has created a number of analysis tools, which are available in the Oxford Nanopore GitHub repository. The tools are aimed at advanced users, and contain instructions for how to install and run the software. They are provided as-is, with minimal support.
If a data analysis method for your research question is not provided in any of the resources above, please refer to the Bioinformatics section of the Resource centre. Numerous members of the Nanopore Community have developed their own tools and pipelines for analysing nanopore sequencing data, most of which are available on GitHub. Please be aware that these tools are not supported by Oxford Nanopore Technologies, and are not guaranteed to be compatible with the latest chemistry/software configuration.
Issues during library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Comments and actions |
---|---|
Low throughput | 1. Vortex gently after adding the diluted Fragmentation Mix (FRA) to break up the largest fragments. 2. Ensure the diluted Fragmentation Mix (FRA) is thoroughly mixed with the gDNA. 3. Use less input material if the DNA library was too viscous to load onto the flow cell. |
DNA is too viscous and will not load onto a flow cell | 1. Lower the input material to reduce the amount of gDNA going into the library preparation and reduce viscosity. 2. If DNA library will not load using the method outlined in this protocol, slowly pipette mix 5 times with a standard P200 pipette set to the full volume of the library and reload the flow cell. |
Read lengths are not long enough | 1. Increase input material. Note: Library viscosity increases with more gDNA. 2. Reduce volume of Fragmentation Mix (FRA) added to FRA Dilution Buffer (FDB) to avoid over-fragmentation of gDNA. Note: We do not recommend diluting less than 2 µl Fragmentation Mix (FRA). 3. We recommend using PFGE to check the extracted gDNA is of ultra-high molecular weight (uHMW), thus capable of generating ultra-long read lengths. |
No sequencing output | 1. Check gDNA has been recovered in library preparation using a NanoDrop spectrophotometer. 2. Check viscosity of the sample. The library should be viscous if it contains uHMW gDNA in this protocol. |
Aspirating supernatant when the DNA has precipitated | Take care to not aspirate the DNA. Remove smaller volumes of supernatant incrementally to reduce the risk of aspirating the DNA. |
Mixing | Mix slowly and carefully to prevent DNA shearing. Low vortexing can be used to mix at the expense of ultra-long reads. With vortexing, long read lengths of ~90 kb N50 can still be generated with improved outputs. |
No DNA recovered from the library preparation clean-up | If the DNA is no longer viscous or the NanoDrop reading is low, DNA may have been lost during the clean-up step of the library preparation. 1. Ensure uHMW DNA is used or users risk DNA loss. 2. Take care to not aspirate the precipitated DNA during the clean-up step. To avoid this, remove smaller volumes of supernatant incrementally. Ensure as much supernatant is removed as possible. |
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.