Version for device: MinION
Overview of the protocol
This kit is recommended for users who:
This protocol describes how to carry out rapid barcoding of plasmid DNA using the Rapid Barcoding Kit 24 or 96 V14 (SQK-RBK114.24 or SQK-RBK114.96) to sequence up to 96 plasmid samples. This method can be utilised for routine verification of plasmid constructs in molecular biology research, quality control of plasmid DNA samples in biotechnology applications, and analysis of engineered plasmids for gene therapy development. During library preparation, the plasmid DNA is tagmented with the Rapid Barcodes before the samples are pooled and cleaned up. Rapid sequencing adapters are attached to the DNA ends before sequencing on a flow cell.
We recommend new users to sequence for 12 hours, although a shorter run-time may be sufficient. After sequencing, perform downstream analysis using the EPI2ME Labs Clone Validation (wf-clone-validation) workflow. A report is generated with a consensus sequence from each plasmid. Detailed instructions for setting up MinKNOW and the EPI2ME Labs workflow are included.
You will need to:
You will need to:
You will need to:
This protocol should only be used in combination with:
It is important that the input DNA meets the quantity and quality requirements. Using too little or too much DNA, or DNA of poor quality (e.g. highly fragmented or containing RNA or chemical contaminants) can affect your library preparation.
For instructions on how to perform quality control of your DNA sample, please read the Input DNA/RNA QC protocol.
Depending on how the DNA is extracted from the raw sample, certain chemical contaminants may remain in the purified DNA, which can affect library preparation efficiency and sequencing quality. Read more about contaminants on the Contaminants page of the Community.
We are in the process of reformatting the barcodes provided in this kit into a plate format. This will reduce plastic waste and facilitates automated applications.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 1 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 2 | 1200 |
Elution Buffer | EB | Black | 1 | 500 |
Sequencing Buffer | SB | Red | 1 | 700 |
Library Beads | LIB | Pink | 1 | 600 |
Library Solution | LIS | White cap, pink label | 1 | 600 |
Flow Cell Flush | FCF | Clear cap, light blue label | 1 | 8000 |
Flow Cell Tether | FCT | Purple | 1 | 200 |
Rapid Barcode plate | RB01-24 | - | 2 plates, 3 sets of barcodes per plate | 5 µl per well |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 1 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 2 | 1,200 |
Elution Buffer | EB | Black | 1 | 500 |
Sequencing Buffer | SB | Red | 1 | 700 |
Library Beads | LIB | Pink | 1 | 600 |
Library Solution | LIS | White cap, pink label | 1 | 600 |
Flow Cell Flush | FCF | Blue | 6 | 1,170 |
Flow Cell Tether | FCT | Purple | 1 | 200 |
Rapid Barcodes | RB01-24 | Clear | 24 | 15 |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 2 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 3 | 1,200 |
Elution Buffer | EB | Black | 1 | 1,500 |
Sequencing Buffer | SB | Red | 1 | 1,700 |
Library Beads | LIB | Pink | 1 | 1,800 |
Library Solution | LIS | White cap, pink label | 1 | 1,800 |
Flow Cell Flush | FCF | Clear | 1 | 15,500 |
Flow Cell Tether | FCT | Purple | 2 | 200 |
Rapid Barcodes | RB01-96 | - | 3 plates | 8 µl per well |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Component | Sequence |
---|---|
RB01 | AAGAAAGTTGTCGGTGTCTTTGTG |
RB02 | TCGATTCCGTTTGTAGTCGTCTGT |
RB03 | GAGTCTTGTGTCCCAGTTACCAGG |
RB04 | TTCGGATTCTATCGTGTTTCCCTA |
RB05 | CTTGTCCAGGGTTTGTGTAACCTT |
RB06 | TTCTCGCAAAGGCAGAAAGTAGTC |
RB07 | GTGTTACCGTGGGAATGAATCCTT |
RB08 | TTCAGGGAACAAACCAAGTTACGT |
RB09 | AACTAGGCACAGCGAGTCTTGGTT |
RB10 | AAGCGTTGAAACCTTTGTCCTCTC |
RB11 | GTTTCATCTATCGGAGGGAATGGA |
RB12 | CAGGTAGAAAGAAGCAGAATCGGA |
RB13 | AGAACGACTTCCATACTCGTGTGA |
RB14 | AACGAGTCTCTTGGGACCCATAGA |
RB15 | AGGTCTACCTCGCTAACACCACTG |
RB16 | CGTCAACTGACAGTGGTTCGTACT |
RB17 | ACCCTCCAGGAAAGTACCTCTGAT |
RB18 | CCAAACCCAACAACCTAGATAGGC |
RB19 | GTTCCTCGTGCAGTGTCAAGAGAT |
RB20 | TTGCGTCCTGTTACGAGAACTCAT |
RB21 | GAGCCTCTCATTGTCCGTTCTCTA |
RB22 | ACCACTGCCATGTATCAAAGTACG |
RB23 | CTTACTACCCAGTGAACCTCCTCG |
RB24 | GCATAGTTCTGCATGATGGGTTAG |
RB25 | GTAAGTTGGGTATGCAACGCAATG |
RB26 | CATACAGCGACTACGCATTCTCAT |
RB27 | CGACGGTTAGATTCACCTCTTACA |
RB28 | TGAAACCTAAGAAGGCACCGTATC |
RB29 | CTAGACACCTTGGGTTGACAGACC |
RB30 | TCAGTGAGGATCTACTTCGACCCA |
RB31 | TGCGTACAGCAATCAGTTACATTG |
RB32 | CCAGTAGAAGTCCGACAACGTCAT |
RB33 | CAGACTTGGTACGGTTGGGTAACT |
RB34 | GGACGAAGAACTCAAGTCAAAGGC |
RB35 | CTACTTACGAAGCTGAGGGACTGC |
RB36 | ATGTCCCAGTTAGAGGAGGAAACA |
RB37 | GCTTGCGATTGATGCTTAGTATCA |
RB38 | ACCACAGGAGGACGATACAGAGAA |
RB39 | CCACAGTGTCAACTAGAGCCTCTC |
RB40 | TAGTTTGGATGACCAAGGATAGCC |
RB41 | GGAGTTCGTCCAGAGAAGTACACG |
RB42 | CTACGTGTAAGGCATACCTGCCAG |
RB43 | CTTTCGTTGTTGACTCGACGGTAG |
RB44 | AGTAGAAAGGGTTCCTTCCCACTC |
RB45 | GATCCAACAGAGATGCCTTCAGTG |
RB46 | GCTGTGTTCCACTTCATTCTCCTG |
RB47 | GTGCAACTTTCCCACAGGTAGTTC |
RB48 | CATCTGGAACGTGGTACACCTGTA |
RB49 | ACTGGTGCAGCTTTGAACATCTAG |
RB50 | ATGGACTTTGGTAACTTCCTGCGT |
RB51 | GTTGAATGAGCCTACTGGGTCCTC |
RB52 | TGAGAGACAAGATTGTTCGTGGAC |
RB53 | AGATTCAGACCGTCTCATGCAAAG |
RB54 | CAAGAGCTTTGACTAAGGAGCATG |
RB55 | TGGAAGATGAGACCCTGATCTACG |
RB56 | TCACTACTCAACAGGTGGCATGAA |
RB57 | GCTAGGTCAATCTCCTTCGGAAGT |
RB58 | CAGGTTACTCCTCCGTGAGTCTGA |
RB59 | TCAATCAAGAAGGGAAAGCAAGGT |
RB60 | CATGTTCAACCAAGGCTTCTATGG |
RB61 | AGAGGGTACTATGTGCCTCAGCAC |
RB62 | CACCCACACTTACTTCAGGACGTA |
RB63 | TTCTGAAGTTCCTGGGTCTTGAAC |
RB64 | GACAGACACCGTTCATCGACTTTC |
RB65 | TTCTCAGTCTTCCTCCAGACAAGG |
RB66 | CCGATCCTTGTGGCTTCTAACTTC |
RB67 | GTTTGTCATACTCGTGTGCTCACC |
RB68 | GAATCTAAGCAAACACGAAGGTGG |
RB69 | TACAGTCCGAGCCTCATGTGATCT |
RB70 | ACCGAGATCCTACGAATGGAGTGT |
RB71 | CCTGGGAGCATCAGGTAGTAACAG |
RB72 | TAGCTGACTGTCTTCCATACCGAC |
RB73 | AAGAAACAGGATGACAGAACCCTC |
RB74 | TACAAGCATCCCAACACTTCCACT |
RB75 | GACCATTGTGATGAACCCTGTTGT |
RB76 | ATGCTTGTTACATCAACCCTGGAC |
RB77 | CGACCTGTTTCTCAGGGATACAAC |
RB78 | AACAACCGAACCTTTGAATCAGAA |
RB79 | TCTCGGAGATAGTTCTCACTGCTG |
RB80 | CGGATGAACATAGGATAGCGATTC |
RB81 | CCTCATCTTGTGAAGTTGTTTCGG |
RB82 | ACGGTATGTCGAGTTCCAGGACTA |
RB83 | TGGCTTGATCTAGGTAAGGTCGAA |
RB84 | GTAGTGGACCTAGAACCTGTGCCA |
RB85 | AACGGAGGAGTTAGTTGGATGATC |
RB86 | AGGTGATCCCAACAAGCGTAAGTA |
RB87 | TACATGCTCCTGTTGTTAGGGAGG |
RB88 | TCTTCTACTACCGATCCGAAGCAG |
RB89 | ACAGCATCAATGTTTGGCTAGTTG |
RB90 | GATGTAGAGGGTACGGTTTGAGGC |
RB91 | GGCTCCATAGGAACTCACGCTACT |
RB92 | TTGTGAGTGGAAAGATACAGGACC |
RB93 | AGTTTCCATCACTTCAGACTTGGG |
RB94 | GATTGTCCTCAAACTGCCACCTAC |
RB95 | CCTGTCTGGAAGAAGAATGGACTT |
RB96 | CTGAACGGTCATAGAGTCCACCAT |
Computer requirements and software
Sequencing on a MinION Mk1B requires a high-spec computer or laptop to keep up with the rate of data acquisition. For more information, refer to the MinION Mk1B IT requirements document.
The MinION Mk1C contains fully-integrated compute and screen, removing the need for any accessories to generate and analyse nanopore data. For more information refer to the MinION Mk1C IT requirements document.
Sequencing on a MinION Mk1D requires a high-spec computer or laptop to keep up with the rate of data acquisition. For more information, refer to the MinION Mk1D IT requirements document.
The MinKNOW software controls the nanopore sequencing device, collects sequencing data and basecalls in real time. You will be using MinKNOW for every sequencing experiment to sequence, basecall and demultiplex if your samples were barcoded.
For instructions on how to run the MinKNOW software, please refer to the MinKNOW protocol.
The EPI2ME cloud-based platform performs further analysis of basecalled data, for example alignment to the Lambda genome, barcoding, or taxonomic classification. You will use the EPI2ME platform only if you would like further analysis of your data post-basecalling.
For instructions on how to create an EPI2ME account and install the EPI2ME Desktop Agent, please refer to this link.
We highly recommend that you check the number of pores in your MinION Flow Cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing the flow cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.
The minimum number of active pores in a MinION Flow Cell that is covered by warranty is 800 pores.
Reagent | 1. Thaw at room temperature | 2. Briefly spin down | 3. Mix well by pipetting |
---|---|---|---|
Rapid Barcodes (RB01-24 or RB01-96)) | Not frozen | ✓ | ✓ |
Rapid Adapter (RA) | Not frozen | ✓ | ✓ |
AMPure XP Beads (AXP) | ✓ | ✓ | Mix by pipetting or vortexing immediately before use |
Elution Buffer (EB) | ✓ | ✓ | ✓ |
Adapter Buffer (ADB) | ✓ | ✓ | Mix by vortexing |
Starting Conc. | Volume of DNA | Volume of nuclease-free water | Total volume |
---|---|---|---|
100 ng/µl | 2 µl | 34 µl | 36 µl |
90 ng/µl | 2 µl | 31 µl | 33 µl |
80 ng/µl | 2 µl | 27 µl | 29 µl |
70 ng/µl | 3 µl | 35 µl | 38 µl |
60 ng/µl | 2 µl | 20 µl | 22 µl |
50 ng/µl | 2 µl | 16 µl | 18 µl |
40 ng/µl | 5 µl | 31 µl | 36 µl |
30 ng/µl | 5 µl | 22 µl | 27 µl |
20 ng/µl | 5 µl | 13 µl | 18 µl |
10 ng/µl | 10 µl | 8 µl | 18 µl |
<5.56 ng/µl | 9 µl | 0 µl | 9 µl |
Please note: Only use one barcode per sample.
Reagent | Volume |
---|---|
50 ng template DNA | 9 μl |
Rapid Barcodes (RB01-96, one for each sample) | 1 μl |
Total | 10 μl |
. | Volume per sample | For 12 samples | For 24 samples | For 48 samples | For 96 samples |
---|---|---|---|---|---|
Total volume | 10 μl | 120 μl | 240 μl | 480 μl | 960 μl |
. | Volume per sample | For 12 samples | For 24 samples | For 48 samples | For 96 samples |
---|---|---|---|---|---|
Volume of AXP | 10 μl | 120 μl | 240 μl | 480 μl | 960 μl |
Note: We recommend transfering a maximum of 800 ng of the DNA library.
If necessary, take forward only the necessary volume for 800 ng of DNA library and make up the rest of the volume to 11 µl using Elution Buffer (EB).
Reagent | Volume |
---|---|
Rapid Adapter (RA) | 1.5 μl |
Adapter Buffer (ADB) | 3.5 μl |
Total | 5 μl |
Priming and loading the SpotON flow cell
We recommend all new users watch the 'Priming and loading your flow cell' video before your first run.
For most sequencing experiments, use the Library Beads (LIB) for loading your library onto the flow cell. However, for viscous libraries it may be difficult to load with the beads and may be appropriate to load using the Library Solution (LIS).
Note: We do not recommend using any other albumin type (e.g. recombinant human serum albumin).
Note: The vials of Flow Cell Flush (FCF) in kit SQK-RBK114.24 and SQK-RBK114.96 have different formats. Please ensure you are using the correct volume when preparing your flow cell priming mix.
Reagents | Volume per flow cell |
---|---|
Flow Cell Flush (FCF) | 1,170 µl |
Bovine Serum Albumin (BSA) at 50 mg/ml | 5 µl |
Flow Cell Tether (FCT) | 30 µl |
Total volume | 1,205 µl |
This step can be omitted if the flow cell has been checked previously.
See the flow cell check instructions in the MinKNOW protocol for more information.
Note: Visually check that there is continuous buffer from the priming port across the sensor array.
We recommend using the Library Beads (LIB) for most sequencing experiments. However, the Library Solution (LIS) is available for more viscous libraries.
Reagent | Volume per flow cell |
---|---|
Sequencing Buffer (SB) | 37.5 µl |
Library Beads (LIB) mixed immediately before use, or Library Solution (LIS), if using | 25.5 µl |
DNA library | 12 µl |
Total | 75 µl |
We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.
Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.
Data acquisition and basecalling
The sequencing device control, data acquisition and real-time basecalling are carried out by the MinKNOW software. Please ensure MinKNOW is installed on your computer or device. Further instructions for setting up a sequencing run can be found in the MinKNOW protocol.
We recommend setting up a sequencing run on a MinION or GridION device using the basecalling and barcoding recommendations outlined below. All other parameters can be left to their default settings.
Click Continue to kit selection.
Click Continue to Run Options to continue.
Click Continue to basecalling to continue.
Ensure basecalling is ON.
Next to "Models", click Edit options and choose High accuracy basecaller (HAC) from the drop-down menu.
Ensure barcoding in ON.
All other options can be kept to their default settings.
Click Continue to output and continue.
Select either .POD5 or .FAST5 (legacy) as the output format.
Ensure .FASTQ is selected for basecalled reads.
Ensure filtering is ON and read splitting is enabled. Other parameters can be kept to their default settings.
Click Continue to final review to continue.
You will be automatically navigated to the "Sequencing Overview" page to monitor the sequencing run.
Downstream analysis using EPI2ME Labs
We recommend performing downstream analysis using EPI2ME Labs which facilitates bioinformatic analyses by allowing users to run Nextflow workflows in a desktop application. EPI2ME Labs maintains a collection of bioinformatic workflows which are curated and actively maintained by experts in long-read sequence analysis.
Further information about the available EPI2ME Labs workflows are available here, along with the Quick Start Guide to start your first bioinformatic workflow.
For the assembly of small plasmid sequences, we recommend using the wf-clone-validation workflow which requires Nextflow and Docker to be installed before running the workflow.
Ensure all parameter options have green ticks.
A report is produced containing the results of the assembled plasmid sequences. The primary outputs of the workflow include:
A sample report can be viewed here.
The summary graph shows the number of reads per barcode and the table shows the length of the consensus sequence for each barcode. These data may be used to identify the samples that may have dropped out of the sequence analysis due to insufficient sequence reads.
For each barcode, read length statistics and a pLannotate plot is presented to illustrate the polished plasmid consensus sequence. In the sample report, barcode01 has been assembled into a 5385 bp consensus sequence. The unfilled features on the plot are incomplete features.
A feature table is also provided for each barcode to give descriptions of the annotated sequence which can be used to identify the precise location of the annotated features.
Ending the experiment
The Flow Cell Wash Kit protocol is available on the Nanopore Community.
Instructions for returning flow cells can be found here.
Issues during DNA/RNA extraction and library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) | The DNA extraction method does not provide the required purity | The effects of contaminants are shown in the Contaminants document. Please try an alternative extraction method that does not result in contaminant carryover. Consider performing an additional SPRI clean-up step. |
Low RNA integrity (RNA integrity number <9.5 RIN, or the rRNA band is shown as a smear on the gel) | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. |
RNA has a shorter than expected fragment length | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. We recommend working in an RNase-free environment, and to keep your lab equipment RNase-free when working with RNA. |
Observation | Possible cause | Comments and actions |
---|---|---|
Low recovery | DNA loss due to a lower than intended AMPure beads-to-sample ratio | 1. AMPure beads settle quickly, so ensure they are well resuspended before adding them to the sample. 2. When the AMPure beads-to-sample ratio is lower than 0.4:1, DNA fragments of any size will be lost during the clean-up. |
Low recovery | DNA fragments are shorter than expected | The lower the AMPure beads-to-sample ratio, the more stringent the selection against short fragments. Please always determine the input DNA length on an agarose gel (or other gel electrophoresis methods) and then calculate the appropriate amount of AMPure beads to use. |
Low recovery after end-prep | The wash step used ethanol <70% | DNA will be eluted from the beads when using ethanol <70%. Make sure to use the correct percentage. |
Observation | Possible cause | Comments and actions |
---|---|---|
The green light was switched off or An adapter was used to connect the VolTRAX USB-C cable to the computer |
Insufficient power supply to the VolTRAX | The green LED signals that 3 A are being supplied to the device. This is the requirement for the full capabilities of the VolTRAX V2 device. Please use computers that meet the requirements listed on the VolTRAX V2 protocol. |
Observation | Possible cause | Comments and actions |
---|---|---|
The VolTRAX software shows an inaccurate amount of reagents loaded | Pipette tips do not fit the VolTRAX cartridge ports | Rainin 20 μl or 30 μl and Gilson 10 μl, 20 μl or 30 μl pipette tips are compatible with loading reagents into the VolTRAX cartridge. Rainin 20 μl is the most suitable. |
The VolTRAX software shows an inaccurate amount of reagents loaded | The angle at which reagents are pipetted into the cartridge is incorrect | The pipetting angle should be slightly greater than the cartridge inlet angle. Please watch the demo video included in the VolTRAX software before loading. |
Issues during the sequencing run
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | An air bubble was introduced into the nanopore array | After the Flow Cell Check it is essential to remove any air bubbles near the priming port before priming the flow cell. If not removed, the air bubble can travel to the nanopore array and irreversibly damage the nanopores that have been exposed to air. The best practice to prevent this from happening is demonstrated in this video. |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | The flow cell is not correctly inserted into the device | Stop the sequencing run, remove the flow cell from the sequencing device and insert it again, checking that the flow cell is firmly seated in the device and that it has reached the target temperature. If applicable, try a different position on the device (GridION/PromethION). |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | Contaminations in the library damaged or blocked the pores | The pore count during the Flow Cell Check is performed using the QC DNA molecules present in the flow cell storage buffer. At the start of sequencing, the library itself is used to estimate the number of active pores. Because of this, variability of about 10% in the number of pores is expected. A significantly lower pore count reported at the start of sequencing can be due to contaminants in the library that have damaged the membranes or blocked the pores. Alternative DNA/RNA extraction or purification methods may be needed to improve the purity of the input material. The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Script failed" | Restart the computer and then restart MinKNOW. If the issue persists, please collect the MinKNOW log files and contact Technical Support. If you do not have another sequencing device available, we recommend storing the flow cell and the loaded library at 4°C and contact Technical Support for further storage guidance. |
Observation | Possible cause | Comments and actions |
---|---|---|
Pore occupancy <40% | Not enough library was loaded on the flow cell | Ensure you load the recommended amount of good quality library in the relevant library prep protocol onto your flow cell. Please quantify the library before loading and calculate mols using tools like the Promega Biomath Calculator, choosing "dsDNA: µg to pmol" |
Pore occupancy close to 0 | The Ligation Sequencing Kit was used, and sequencing adapters did not ligate to the DNA | Make sure to use the NEBNext Quick Ligation Module (E6056) and Oxford Nanopore Technologies Ligation Buffer (LNB, provided in the sequencing kit) at the sequencing adapter ligation step, and use the correct amount of each reagent. A Lambda control library can be prepared to test the integrity of the third-party reagents. |
Pore occupancy close to 0 | The Ligation Sequencing Kit was used, and ethanol was used instead of LFB or SFB at the wash step after sequencing adapter ligation | Ethanol can denature the motor protein on the sequencing adapters. Make sure the LFB or SFB buffer was used after ligation of sequencing adapters. |
Pore occupancy close to 0 | No tether on the flow cell | Tethers are adding during flow cell priming (FLT/FCT tube). Make sure FLT/FCT was added to FB/FCF before priming. |
Observation | Possible cause | Comments and actions |
---|---|---|
Shorter than expected read length | Unwanted fragmentation of DNA sample | Read length reflects input DNA fragment length. Input DNA can be fragmented during extraction and library prep. 1. Please review the Extraction Methods in the Nanopore Community for best practice for extraction. 2. Visualise the input DNA fragment length distribution on an agarose gel before proceeding to the library prep. In the image above, Sample 1 is of high molecular weight, whereas Sample 2 has been fragmented. 3. During library prep, avoid pipetting and vortexing when mixing reagents. Flicking or inverting the tube is sufficient. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of unavailable pores (shown as blue in the channels panel and pore activity plot) The pore activity plot above shows an increasing proportion of "unavailable" pores over time. |
Contaminants are present in the sample | Some contaminants can be cleared from the pores by the unblocking function built into MinKNOW. If this is successful, the pore status will change to "sequencing pore". If the portion of unavailable pores stays large or increases: 1. A nuclease flush using the Flow Cell Wash Kit (EXP-WSH004) can be performed, or 2. Run several cycles of PCR to try and dilute any contaminants that may be causing problems. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of inactive/unavailable pores (shown as light blue in the channels panel and pore activity plot. Pores or membranes are irreversibly damaged) | Air bubbles have been introduced into the flow cell | Air bubbles introduced through flow cell priming and library loading can irreversibly damage the pores. Watch the Priming and loading your flow cell video for best practice |
Large proportion of inactive/unavailable pores | Certain compounds co-purified with DNA | Known compounds, include polysaccharides, typically associate with plant genomic DNA. 1. Please refer to the Plant leaf DNA extraction method. 2. Clean-up using the QIAGEN PowerClean Pro kit. 3. Perform a whole genome amplification with the original gDNA sample using the QIAGEN REPLI-g kit. |
Large proportion of inactive/unavailable pores | Contaminants are present in the sample | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
Reduction in sequencing speed and q-score later into the run | For Kit 9 chemistry (e.g. SQK-LSK109), fast fuel consumption is typically seen when the flow cell is overloaded with library (please see the appropriate protocol for your DNA library to see the recommendation). | Add more fuel to the flow cell by following the instructions in the MinKNOW protocol. In future experiments, load lower amounts of library to the flow cell. |
Observation | Possible cause | Comments and actions |
---|---|---|
Temperature fluctuation | The flow cell has lost contact with the device | Check that there is a heat pad covering the metal plate on the back of the flow cell. Re-insert the flow cell and press it down to make sure the connector pins are firmly in contact with the device. If the problem persists, please contact Technical Services. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Failed to reach target temperature" | The instrument was placed in a location that is colder than normal room temperature, or a location with poor ventilation (which leads to the flow cells overheating) | MinKNOW has a default timeframe for the flow cell to reach the target temperature. Once the timeframe is exceeded, an error message will appear and the sequencing experiment will continue. However, sequencing at an incorrect temperature may lead to a decrease in throughput and lower q-scores. Please adjust the location of the sequencing device to ensure that it is placed at room temperature with good ventilation, then re-start the process in MinKNOW. Please refer to this link for more information on MinION temperature control. |
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.