Version for device: Flongle
Other available device versions: MinION
This protocol describes how to carry out rapid barcoding of 16S amplicons using the 16S Barcoding Kit 24 V14 (SQK-16S114.24). Due to the presence of both highly conserved (adequate for universal primers and phylogenetic signal) and highly variant regions (different across species), the 16S rRNA gene is often used for sequence-based bacterial identification.
The 16S Barcoding Kit 24 V14 enables access to rapid 16S sequencing for organism identification. By narrowing down to a specific region of interest, you can see all the organisms in the sample without sequencing unnecessary regions of the genome, making the test quicker and more economical. There are 24 unique barcodes, allowing you to pool up to 24 different samples in one sequencing experiment.
After sequencing, you can perform downstream analysis using the EPI2ME 16S workflow (wf-16s) to classify 16S amplicons from your samples.
You will need to:
The table below is an overview of the steps required in the library preparation, including timings and stopping points.
Library preparation step | Process | Time | Stop option |
---|---|---|---|
16S barcoded PCR amplification | Amplify the 16S gene using barcodes supplied in the kit | 10 minutes + PCR | 4°C overnight |
Barcoded sample pooling and bead clean-up | Quantify and pool the barcoded samples and perform a library clean-up using beads | 15 minutes | 4°C short-term storage or for repeated use, such as re-loading your flow cell. -80°C for single-use long-term storage. |
Adapter ligation | Attach the rapid sequencing adapters to the to the DNA ends. | 5 minutes | We strongly recommend sequencing your library as soon as it is adapted. |
Priming and loading the flow cell | Prime the flow cell and load the prepared DNA library for sequencing | 5 minutes |
You will need to:
This protocol should only be used in combination with:
There are three buffers that come into direct contact with a flow cell at point of loading (SB: Sequencing Buffer, FCF: Flow Cell Flush and LIB: Library Beads or LIS: Library Solution). When looking at these buffers, we found that there are a very low level of contaminants seeping out of the plastic vials that impacts the robustness of the Flongle flow cell system (MinION and PromethION are not impacted by this).
We have found that when storing these buffers in glass vials instead of plastic, incidence of deterioration is reduced.
To rapidly deploy this to Flongle users, we have produced a Flongle Sequencing Expansion (EXP-FSE002) with these three components in glass vials, which can perform 12 Flongle flow cell loads in total.
To load a library onto your Flongle flow cell, you will need to use the following components:
Flongle Sequencing Expansion (EXP-FSE002) components
- Sequencing Buffer (SB)
- Flow Cell Flush (FCF)
- Library Beads (LIB) or Library Solution (LIS)
Sequencing Kit components
- Flow Cell Tether (FCT)
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
For optimal output, we currently do not recommend using fewer than 4 barcodes. If you wish to multiplex less than 4 samples, please ensure you split your sample(s) across multiple barcodes so at least 4 barcodes are run (e.g. for 2 samples, use 16S Barcode Primers 01-02 for sample A, and 16S Barcode Primers 03-04 for sample B). Please note that the required sample input for each barcode is 10 ng gDNA.
We have validated and recommend the use of all the third-party reagents used in this protocol. Alternatives have not been tested by Oxford Nanopore Technologies.
For all third-party reagents, we recommend following the manufacturer's instructions to prepare the reagents for use.
We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (μl) |
---|---|---|---|---|
16S Barcode Primers 01-24 | 1-24 | - | 2 plates, 3 sets of barcodes per plate | 15 μl per well |
Rapid Adapter | RA | Green | 1 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Clear cap, light teal label | 1 | 6,000 |
Elution buffer | EB | Black | 1 | 1,500 |
EDTA | EDTA | Blue | 1 | 700 |
Sequencing Buffer | SB | Red | 1 | 700 |
Library Beads | LIB | Pink | 1 | 600 |
Library Solution | LIS | White cap, pink label | 1 | 600 |
Flow Cell Flush | FCF | Clear cap, light blue label | 1 | 8,000 |
Flow Cell Tether | FCT | Purple | 1 | 200 |
Note: This product contains AMPure XP Reagent manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Name | Acronym | Cap colour | Number of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Sequencing Buffer | SB | Blue | 1 | 250 |
Library Beads | LIB | Blue | 1 | 200 |
Library Solution | LIS | Blue | 1 | 200 |
Flow Cell Flush | FCF | Blue | 1 | 1,600 |
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
For optimal output, we currently do not recommend using fewer than 4 barcodes. If you wish to multiplex less than 4 samples, please ensure you split your sample(s) across barcodes so a minimum of 4 barcodes are run:
Please note that the required sample input for each barcode is 10 ng gDNA.
We recommend performing a flow cell check before starting your library prep to ensure you have a flow cell with enough pores for a good sequencing run.
See the flow cell check instructions in the MinKNOW protocol for more information.
16S barcode plate layout:
Reagent | Volume |
---|---|
10 ng input DNA (from previous step) | 15 μl |
LongAmp Hot Start Taq 2X Master Mix | 25 μl |
Total | 40 μl |
Note: If the amount of input material is altered, the number of PCR cycles may need to be adjusted to produce the same yield.
Note: Mix gently to minimise introducing air bubbles to the reactions.
Cycle step | Temperature | Time | No. of cycles |
---|---|---|---|
Initial denaturation | 95 °C | 1 min | 1 |
Denaturation | 95 °C | 20 secs | 25 |
Annealing | 55 °C | 30 secs | 25 |
Extension | 65 °C | 2 mins | 25 |
Final extension | 65 °C | 5 mins | 1 |
Hold | 4 °C | ∞ |
Reagent | 1. Thaw at room temperature | 2. Briefly spin down | 3. Mix well by pipetting or vortexing |
---|---|---|---|
Rapid Adapter (RA) | Not frozen | ✓ | Pipette |
Adapter Buffer (ADB) | ✓ | ✓ | Vortex or Pipette |
AMPure XP Beads (AXP) | ✓ | ✓ | Mix by vortexing immediately before use |
Elution Buffer (EB) | ✓ | ✓ | Vortex or Pipette |
EDTA (EDTA) | ✓ | ✓ | Vortex or Pipette |
Note: Once thawed, keep all reagents on ice.
Note: Please ensure you have quantified your samples prior to this step and take forward an equimolar concentration of each of the samples for optimal barcode balancing.
Samples may vary in concentration following the barcoded PCR, therefore the volume of each barcoded sample added to the pool will be different.
Volume of barcoded sample pool | 37.5 μl | 75 μl | 150 μl | 300 μl | 600 μl |
---|---|---|---|---|---|
Volume of AMPure XP Beads (AXP) | 22.5 μl | 45 μl | 90 μl | 180 μl | 360 μl |
Note: Table contains example volumes for reference. Please adjust the volume of AMPure XP Beads (AXP) added for the volume of your barcoded sample pool to ensure a 0.6X volume ratio.
Reagent | Volume |
---|---|
Rapid Adapter (RA) | 1.5 μl |
Adapter Buffer (ADB) | 3.5 μl |
Total | 5 μl |
Loading the Flongle Flow Cell
There are three buffers that come into direct contact with a flow cell at point of loading (SB: Sequencing Buffer, FCF: Flow Cell Flush and LIB: Library Beads or LIS: Library Solution). When looking at these buffers, we found that there are a very low level of contaminants seeping out of the plastic vials that impacts the robustness of the Flongle flow cell system (MinION and PromethION are not impacted by this).
We have found that when storing these buffers in glass vials instead of plastic, incidence of deterioration is reduced.
To rapidly deploy this to Flongle users, we have produced a Flongle Sequencing Expansion (EXP-FSE002) with these three components in glass vials, which can perform 12 Flongle flow cell loads in total.
To load a library onto your Flongle flow cell, you will need to use the following components:
Flongle Sequencing Expansion (EXP-FSE002) components
- Sequencing Buffer (SB)
- Flow Cell Flush (FCF)
- Library Beads (LIB) or Library Solution (LIS)
Sequencing Kit components
- Flow Cell Tether (FCT)
Oxford Nanopore Technologies deem the useful life of the Flow Cell Expansion to be 6 months from receipt by the customer.
The seal tab, air vent, waste channel, drain port and sample port are visible here. The sample port, drain port and air vent only become accessible once the seal tab is peeled back.
The adapter should sit evenly and flat on the MinION Mk1B or GridION platform. This ensures the flow cell assembly is flat during the next stage.
The flow cell should sit evenly and flat inside the adapter, to avoid any bubbles forming inside the fluidic compartments.
Lift up the seal tab:
Pull the seal tab to open access to the sample port:
Hold the seal tab open by using adhesive on the tab to stick to the MinION Mk 1B lid:
We recommend using the Library Beads (LIB) for most sequencing experiments. However, the Library Solution (LIS) is available for more viscous libraries.
Reagents | Volume |
---|---|
Sequencing Buffer (SB) | 15 µl |
Library Beads (LIB) mixed immediately before use, or Library Solution (LIS), if using. | 10 µl |
DNA library | 5 µl |
Total | 30 µl |
Stick the transparent adhesive tape to the sample port.
Replace the top (Wheel icon section) of the seal tab to its original position.
Data acquisition and basecalling
Once you have loaded your flow cell, the sequencing run can be started on MinKNOW, our sequencing software that controls the device, data acquisition and real-time basecalling. For more detailed information on setting up and using MinKNOW, please see the MinKNOW protocol.
MinKNOW can be used and set up to sequence in multiple ways:
For more information on using MinKNOW on a sequencing device, please see the device user manuals:
To start a sequencing run on MinKNOW:
1. Navigate to the start page and click Start sequencing.
2. Fill in your experiment details, such as name and flow cell position and sample ID.
3. Select the sequencing kit used in the library preparation on the Kit page.
4. Configure the sequencing parameters for your sequencing run or keep to the default settings on the Run options and Analysis tabs.
Note: If basecalling was turned off when a sequencing run was set up, basecalling can be performed post-run on MinKNOW. For more information, please see the MinKNOW protocol.
5. On the Output page, set up the output parameters or keep to the default settings.
6. Click Start on the Review page to start the sequencing run.
After sequencing and basecalling, the data can be analysed. For further information about options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.
In the Downstream analysis section, we outline further options for analysing your data.
Downstream analysis
We recommend performing downstream analysis using EPI2ME which facilitates bioinformatic analyses by allowing users to run Nextflow workflows in a desktop application. EPI2ME maintains a collection of bioinformatic workflows which are curated and actively maintained by experts in long-read sequence analysis.
Further information about the available EPI2ME workflows are available here, along with the Quick Start Guide to start your first bioinformatic workflow.
The 16S workflow (wf-16s) is a Nextflow workflow leveraging the power of wf-metagenomics for identification of the origin of reads from targeted amplicon sequencing. The workflow has two modes of operation, it can use either kraken2 or minimap2 to determine the origin of reads.
More information on the EPI2ME 16S workflow (wf-16s) can be found here.
For installation instructions please click here.
1. Research analysis tools
Oxford Nanopore Technologies' Research division has created a number of analysis tools, that are available in the Oxford Nanopore GitHub repository. The tools are aimed at advanced users, and contain instructions for how to install and run the software. They are provided as-is, with minimal support.
2. Community-developed analysis tools
If a data analysis method for your research question is not provided in any of the resources above, please refer to the Bioinformatics section of the Resource centre. Numerous members of the Nanopore Community have developed their own tools and pipelines for analysing nanopore sequencing data, most of which are available on GitHub. Please be aware that these tools are not supported by Oxford Nanopore Technologies, and are not guaranteed to be compatible with the latest chemistry/software configuration.
Issues during DNA/RNA extraction and library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) | The DNA extraction method does not provide the required purity | The effects of contaminants are shown in the Contaminants document. Please try an alternative extraction method that does not result in contaminant carryover. Consider performing an additional SPRI clean-up step. |
Low RNA integrity (RNA integrity number <9.5 RIN, or the rRNA band is shown as a smear on the gel) | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. |
RNA has a shorter than expected fragment length | The RNA degraded during extraction | Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page. We recommend working in an RNase-free environment, and to keep your lab equipment RNase-free when working with RNA. |
Observation | Possible cause | Comments and actions |
---|---|---|
Low recovery | DNA loss due to a lower than intended AMPure beads-to-sample ratio | 1. AMPure beads settle quickly, so ensure they are well resuspended before adding them to the sample. 2. When the AMPure beads-to-sample ratio is lower than 0.4:1, DNA fragments of any size will be lost during the clean-up. |
Low recovery | DNA fragments are shorter than expected | The lower the AMPure beads-to-sample ratio, the more stringent the selection against short fragments. Please always determine the input DNA length on an agarose gel (or other gel electrophoresis methods) and then calculate the appropriate amount of AMPure beads to use. |
Low recovery after end-prep | The wash step used ethanol <80% | DNA will be eluted from the beads when using ethanol <80%. Make sure to use the correct percentage. |
Issues during the sequencing run using a Rapid-based sequencing kit
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | An air bubble was introduced into the nanopore array | After the Flow Cell Check it is essential to remove any air bubbles near the priming port before priming the flow cell. If not removed, the air bubble can travel to the nanopore array and irreversibly damage the nanopores that have been exposed to air. The best practice to prevent this from happening is demonstrated in this video. |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | The flow cell is not correctly inserted into the device | Stop the sequencing run, remove the flow cell from the sequencing device and insert it again, checking that the flow cell is firmly seated in the device and that it has reached the target temperature. If applicable, try a different position on the device (GridION/PromethION). |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | Contaminations in the library damaged or blocked the pores | The pore count during the Flow Cell Check is performed using the QC DNA molecules present in the flow cell storage buffer. At the start of sequencing, the library itself is used to estimate the number of active pores. Because of this, variability of about 10% in the number of pores is expected. A significantly lower pore count reported at the start of sequencing can be due to contaminants in the library that have damaged the membranes or blocked the pores. Alternative DNA/RNA extraction or purification methods may be needed to improve the purity of the input material. The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Script failed" | Restart the computer and then restart MinKNOW. If the issue persists, please collect the MinKNOW log files and contact Technical Support. If you do not have another sequencing device available, we recommend storing the flow cell and the loaded library at 4°C and contact Technical Support for further storage guidance. |
Observation | Possible cause | Comments and actions |
---|---|---|
Pore occupancy <40% | Not enough library was loaded on the flow cell | 10–50 fmol of good quality library can be loaded on to a MinION Mk1B/GridION flow cell. Please quantify the library before loading and calculate mols using tools like the Promega Biomath Calculator, choosing "dsDNA: µg to pmol" |
Pore occupancy close to 0 | The Rapid PCR Barcoding Kit V14 was used, and sequencing adapters did not attach to the DNA | Make sure to closely follow the protocol and use the correct volumes and incubation temperatures. A Lambda control library can be prepared to test the integrity of reagents. |
Pore occupancy close to 0 | No tether on the flow cell | Tethers are added during flow cell priming (FCT tube). Make sure FCT was added to FCF before priming. |
Observation | Possible cause | Comments and actions |
---|---|---|
Shorter than expected read length | Unwanted fragmentation of DNA sample | Read length reflects input DNA fragment length. Input DNA can be fragmented during extraction and library prep. 1. Please review the Extraction Methods in the Nanopore Community for best practice for extraction. 2. Visualise the input DNA fragment length distribution on an agarose gel before proceeding to the library prep. In the image above, Sample 1 is of high molecular weight, whereas Sample 2 has been fragmented. 3. During library prep, avoid pipetting and vortexing when mixing reagents. Flicking or inverting the tube is sufficient. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of unavailable pores (shown as blue in the channels panel and pore activity plot) The pore activity plot above shows an increasing proportion of "unavailable" pores over time. |
Contaminants are present in the sample | Some contaminants can be cleared from the pores by the unblocking function built into MinKNOW. If this is successful, the pore status will change to "sequencing pore". If the portion of unavailable pores stays large or increases: 1. A nuclease flush using the Flow Cell Wash Kit (EXP-WSH004) can be performed, or 2. Run several cycles of PCR to try and dilute any contaminants that may be causing problems. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of inactive/unavailable pores (shown as light blue in the channels panel and pore activity plot. Pores or membranes are irreversibly damaged) | Air bubbles have been introduced into the flow cell | Air bubbles introduced through flow cell priming and library loading can irreversibly damage the pores. Watch the Priming and loading your flow cell video for best practice |
Large proportion of inactive/unavailable pores | Certain compounds co-purified with DNA | Known compounds, include polysaccharides, typically associate with plant genomic DNA. 1. Please refer to the Plant leaf DNA extraction method. 2. Clean-up using the QIAGEN PowerClean Pro kit. 3. Perform a whole genome amplification with the original gDNA sample using the QIAGEN REPLI-g kit. |
Large proportion of inactive/unavailable pores | Contaminants are present in the sample | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
Temperature fluctuation | The flow cell has lost contact with the device | Check that there is a heat pad covering the metal plate on the back of the flow cell. Re-insert the flow cell and press it down to make sure the connector pins are firmly in contact with the device. If the problem persists, please contact Technical Services. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Failed to reach target temperature" | The instrument was placed in a location that is colder than normal room temperature, or a location with poor ventilation (which leads to the flow cells overheating) | MinKNOW has a default timeframe for the flow cell to reach the target temperature. Once the timeframe is exceeded, an error message will appear and the sequencing experiment will continue. However, sequencing at an incorrect temperature may lead to a decrease in throughput and lower q-scores. Please adjust the location of the sequencing device to ensure that it is placed at room temperature with good ventilation, then re-start the process in MinKNOW. Please refer to this FAQ for more information on MinION temperature control. |
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.