Version for device: MinION
Overview of the protocol
This end to end protocol describes our Nanopore-Only Microbial Isolate Sequencing Solution (NO-MISS): a flexible approach allowing sequencing of 4 to 24 microbial isolate genomes per MinION Flow Cell, generating a minimum coverage of 50x per genome.
The 50X coverage threshold is sufficient for downstream analyses including: accurate assembly and plasmid resolution, AMR profiling, core genome (cg) and whole genome (wg) multi-locus sequence typing (MLST), and cg/wgSNP typing. You can analyse your sequencing data using EPI2ME, which provides a user-friendly bioinformatics workflow.
We provide multiple DNA extraction approaches, depending on requirements, and starting organism (bacteria, fungi/yeast). These are key in achieving reliable flow cell output and genome coverage. The sample specific extraction methods use NEB Monarch® Spin gDNA Extraction Kit, while the universal method uses a bead-beating method and the Maxwell® RCS PureFood Pathogen Kit.
The extracted gDNA is then tagmented and sequenced using our Rapid Barcoding Kit (SQK-RBK114.24 or SQK-RBK114.96). Up to 24 samples per sequencing experiment for bacterial isolates (up to 7 Mb genomes) and up to 8 samples for fungi/yeast isolates can be processed to achieve the 50x coverage threshold. Use a minimum of four barcodes per run to maintain performance.
Detailed instructions for setting up the sequencing run on MinKNOW and downstream analysis are also included for a complete end-to-end protocol. We recommend sequencing up to 72 hours and generating at least 50x coverage per sample (approx. 0.5 Gb per barcode, assuming 5 Mb genome).
We recommend updating MinKNOW to the latest version prior to starting a sequencing run. The basecalling model v4.3 found in Dorado 0.5.0 onwards provides improved accuracy for bacterial DNA and is included in MinKNOW release v24.02 or newer.
For more information on updating MinKNOW, please refer to our MinKNOW protocol.
You will need to:
Using the relevant gDNA extraction method, you will need to lyse your cells, extract your gDNA, and quantify the DNA:
The table below is an overview of the steps required in the library preparation, including timings and stopping points.
Library preparation step | Process | Time | Stop option |
---|---|---|---|
DNA barcoding | Tagmentation of the DNA using the Rapid Barcoding Kit V14 | 15 minutes | 4°C overnight |
Sample pooling and clean-up | Pooling of barcoded libraries and AMPure XP Bead clean-up | 25 minutes | 4°C overnight |
Adapter ligation | Attach the sequencing adapters to the DNA ends | 5 minutes | We strongly recommend sequencing your library as soon as it is adapted |
Priming and loading the flow cell | Prime the flow cell and load the prepared library for sequencing | 5 minutes |
Sequencing and analysis
You will need to:
This protocol should only be used in combination with:
Input requirements per sample for the extraction methods:
Universal bead-beating gDNA extraction: 1 ml liquid overnight culture (~1 x 108 – 109 cfu/ml) or half of a loop of colonies from a plate
Bacteria gDNA extraction: 200 µl liquid overnight culture (~1 x 108 – 109 cfu/ml) or 1/8 of a loop of colonies from a plate
Hard to lyse organisms gDNA extraction: 5 – 10 mg cells from solid or liquid media
Fungi gDNA extraction: 2 ml of ~1 x 107 cfu/ml overnight culture or a full 10 µl inoculating loop from a plate
For library preparation, 200 ng in 10 µl of extracted gDNA per sample is required.
Depending on the extraction protocol used, not all third-party reagents are required.
We have validated and recommend the use of all the third-party reagents used in this protocol. Alternatives have not been tested by Oxford Nanopore Technologies.
For all third-party reagents, we recommend following the manufacturer's instructions to prepare the reagents for use.
Custom CTAB buffer can be prepared instead of purchasing. Below are the reagents and concentrations required, with suggested examples with excess.
Reagent | Stock | Final concentration | Volume for 12 samples | Volume for 24 samples |
---|---|---|---|---|
CTAB | - | 2% v/v | 60 µl | 120 µl |
EDTA | 0.5 M | 40 mM | 240 µl | 480 µl |
Sodium chloride | 5 M | 1.4 M | 833 µl | 1,666 µl |
Trizma hydrochloride solution, pH 8 | 1 M | 100 mM | 300 µl | 600 µl |
Nuclease-free water | - | - | 1567 µl | 3,134 µl |
Total | - | - | 3,000 µl | 6,000 µl |
Staphylococcal Lysis Buffer (SLB) is required for the bacterial gDNA extraction method for staphylococcal inputs.
Reagent | Stock | Final concentration | Volume for 12 samples with excess | Volume for 24 samples with excess |
---|---|---|---|---|
Trizma hydrochloride solution, pH 9 | 1 M | 100 mM | 150 ul | 300 µl |
Sodium chloride | 5 M | 10 mM | 3 ul | 6 µl |
SDS | 10% v/v | 0.1% v/v | 15 ul | 30 µl |
Nuclease-free water | - | - | 1332 ul | 2664 µl |
Total volume | - | - | 1,500 ul | 3,000 µl |
The SDS in the Staphylococcal Lysis Buffer (SLB) is essential for preventing the degradation of staphylococci DNA. Exclusion of SDS from the buffer results in a larger smear of DNA when run on a gel.
It is important that the input DNA meets the quantity and quality requirements. Using too little or too much DNA, or DNA of poor quality (e.g. highly fragmented or containing RNA or chemical contaminants) can affect your library preparation.
For instructions on how to perform quality control of your DNA sample, please read the Input DNA/RNA QC protocol.
Depending on how the DNA is extracted from the raw sample, certain chemical contaminants may remain in the purified DNA, which can affect library preparation efficiency and sequencing quality. Read more about contaminants on the Contaminants page of the Community.
We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
We are in the process of reformatting the barcodes provided in this kit into a plate format. This will reduce plastic waste and facilitates automated applications.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 1 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 2 | 1200 |
Elution Buffer | EB | Black | 1 | 500 |
Sequencing Buffer | SB | Red | 1 | 700 |
Library Beads | LIB | Pink | 1 | 600 |
Library Solution | LIS | White cap, pink label | 1 | 600 |
Flow Cell Flush | FCF | Clear cap, light blue label | 1 | 8000 |
Flow Cell Tether | FCT | Purple | 1 | 200 |
Rapid Barcode plate | RB01-24 | - | 2 plates, 3 sets of barcodes per plate | 5 µl per well |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 1 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 2 | 1,200 |
Elution Buffer | EB | Black | 1 | 500 |
Sequencing Buffer | SB | Red | 1 | 700 |
Library Beads | LIB | Pink | 1 | 600 |
Library Solution | LIS | White cap, pink label | 1 | 600 |
Flow Cell Flush | FCF | Blue | 6 | 1,170 |
Flow Cell Tether | FCT | Purple | 1 | 200 |
Rapid Barcodes | RB01-24 | Clear | 24 | 15 |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 2 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 3 | 1,200 |
Elution Buffer | EB | Black | 1 | 1,500 |
Sequencing Buffer | SB | Red | 1 | 1,700 |
Library Beads | LIB | Pink | 1 | 1,800 |
Library Solution | LIS | White cap, pink label | 1 | 1,800 |
Flow Cell Flush | FCF | Clear | 1 | 15,500 |
Flow Cell Tether | FCT | Purple | 2 | 200 |
Rapid Barcodes | RB01-96 | - | 3 plates | 8 µl per well |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
Sample extraction method selection
We have developed four optimised extraction methods to generate high quality genomic DNA from your cell cultures, allowing maximised sequencing output using this method.
What extraction method is right for me?
Sample extraction method | Sample type | Sample Input | Expected yield | Expected DNA Integrity Number (DIN) | Average sequencing read lengths | Extraction kits used |
---|---|---|---|---|---|---|
Universal bead-beating gDNA extraction | Universal applications: bacteria, fungi or yeast | 1 ml liquid overnight culture (~1 x 108 – 109 cfu/ml) or half of a loop of colonies from a plate | >200 ng/µl per sample | 7-9 | ~4-7 kb | QIAGEN PowerBead Tube and Promega Maxwell® RSC PureFood Pathogen kit |
Bacteria gDNA extraction | Bacterial | 200 µl liquid overnight culture (~1 x 108 – 109 cfu/ml) or 1/8 of a loop of colonies from a plate | ~15-20 ng/µl per sample | 9 | >7 kb - Size will vary based on sample input species | NEB Monarch Spin gDNA Extraction Kit |
Hard to lyse organisms gDNA extraction | Mycobacterium tuberculosis (or hard to extract bacterial samples) |
5 – 10 mg cells from solid or liquid media | ~15-40 ng/µl per sample | 8 | >7 kb - Size will vary based on sample input species | NEB Monarch Spin gDNA Extraction Kit |
Fungi gDNA extraction | Fungi or yeast | 2 ml of ~1 x 107 cfu/ml overnight culture or a full 10 µl inoculating loop from a plate | ~40 ng/µl per sample | N/A | >7 kb - Size will vary based on sample input species | NEB Monarch Spin gDNA Extraction Kit |
Note: The yield, DIN and sequencing read length of extracted DNA may vary depending on sample quality and species. Please ensure you are following the correct method and using high-quality sample inputs.
Follow the links in the table above for the extraction methods documentation.
Alternatively, these extraction methods can be found in the Extraction Protocols tab in the Documentation space on the Nanopore Community
This method has been developed to process 24 samples with a genome size of up to 7 Mb simultaneously.
For samples with a larger genome size ( >7 Mb), we recommend lowering the number of samples to be barcoded and sequenced simultaneously to 8 samples.
For optimal output, we currently do not recommend using fewer than 4 barcodes.
Note: This method provides a standardised process for sample throughput that we have validated in-house. These settings will be applicable to the majority of use-cases and we recommend all new users to follow the recommended method. Experienced users can adjust sample throughput (4 to 48 barcoded samples) based on sample quality, genome size, and coverage requirements.
We recommend performing a flow cell check before starting your library prep to ensure you have a flow cell with enough pores for a good sequencing run.
See the flow cell check instructions in the MinKNOW protocol for more information.
Reagent | 1. Thaw at room temperature | 2. Briefly spin down | 3. Mix well by pipetting |
---|---|---|---|
Rapid Barcodes (RB01-24 or RB01-96)) | Not frozen | ✓ | ✓ |
Rapid Adapter (RA) | Not frozen | ✓ | ✓ |
AMPure XP Beads (AXP) | ✓ | ✓ | Mix by pipetting or vortexing immediately before use |
Elution Buffer (EB) | ✓ | ✓ | ✓ |
Adapter Buffer (ADB) | ✓ | ✓ | Mix by vortexing |
Note: Use one barcode per sample.
Reagent | Volume per sample |
---|---|
Template DNA (200 ng from previous step) | 10 μl |
Rapid Barcodes (RB01-24 or RB01-96, one for each sample) | 1.5 μl |
Total | 11.5 μl |
. | Volume per sample | For 24 samples |
---|---|---|
Total volume | 11.5 µl | 276 µl |
. | Volume per sample | For 24 samples |
---|---|---|
Volume of AMPure XP Beads (AXP) added | 11.5 µl | 276 µl |
Expect ~150 ng/µl for 24 samples, assuming 70% of DNA was retained during the wash.
Reagent | Volume |
---|---|
Rapid Adapter (RA) | 1.5 μl |
Adapter Buffer (ADB) | 3.5 μl |
Total | 5 μl |
Tip: While this incubation step is taking place you can proceed to the Flow Cell priming and loading section of the protocol.
Priming and loading the MinION and GridION Flow Cell
We recommend all new users watch the 'Priming and loading your flow cell' video before your first run.
Note: We do not recommend using any other albumin type (e.g. recombinant human serum albumin).
Reagents | Volume per flow cell |
---|---|
Flow Cell Flush (FCF) | 1,170 µl |
Bovine Serum Albumin (BSA) at 50 mg/ml | 5 µl |
Flow Cell Tether (FCT) | 30 µl |
Final total volume in tube | 1,205 µl |
This step can be omitted if the flow cell has been checked previously.
See the flow cell check instructions in the MinKNOW protocol for more information.
Note: Visually check that there is continuous buffer from the priming port across the sensor array.
We recommend using the Library Beads (LIB) for most sequencing experiments. However, the Library Solution (LIS) is available for more viscous libraries.
Reagent | Volume per flow cell |
---|---|
Sequencing Buffer (SB) | 37.5 µl |
Library Beads (LIB) mixed immediately before use, or Library Solution (LIS), if using | 25.5 µl |
DNA library | 12 µl |
Total | 75 µl |
We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.
Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.
Data acquisition and basecalling
We recommend updating MinKNOW to the latest version prior to starting a sequencing run. The basecalling model v4.3 found in Dorado 0.5.0 onwards provides improved accuracy for bacterial DNA and is included in MinKNOW release v24.02 or newer.
For more information on updating MinKNOW, please refer to our MinKNOW protocol.
The sequencing device control, data acquisition and real-time basecalling are carried out by the MinKNOW software. Please ensure MinKNOW is installed on your computer or device. Further instructions for setting up a sequencing run can be found in the MinKNOW protocol.
We recommend setting up a sequencing run on a MinION or GridION device using the basecalling and barcoding recommendations outlined below. All other parameters can be left to their default settings.
Click Continue to kit selection.
Click Continue to Run Options to continue.
Click Continue to basecalling to continue.
Ensure basecalling is ON.
Next to "Models", click Edit options and choose High accuracy basecaller (HAC) from the drop-down menu.
Ensure barcoding in ON.
Click Continue to output and continue.
Ensure .POD5 is seleceted as the Raw reads output format.
Ensure .FASTQ is selected for basecalled reads.
Ensure filtering is ON.
Click Continue to final review to continue.
You will be automatically navigated to the "Sequencing Overview" page to monitor the sequencing run.
Downstream analysis
We recommend performing downstream analysis using EPI2ME which facilitates bioinformatic analyses by allowing users to run Nextflow workflows in a desktop application. EPI2ME maintains a collection of bioinformatic workflows which are curated and actively maintained by experts in long-read sequence analysis.
Further information about the available EPI2ME workflows are available here, along with the Quick Start Guide to start your first bioinformatic workflow.
For the accurate and reliable assembly or alignment of bacterial or fungal isolate genomes generated from the included protocols, we recommend using the wf-bacterial-genomes workflow. At its core, the workflow is an efficient assembly pipeline which also polishes genomes using Medaka.
Whilst running the workflow using the default parameters will produce high quality genome assemblies, using the ‘Isolates’ mode will perform additional analyses designed to increase genome quality and aid genome interrogation for common pathogens in the clinical and food safety fields. ‘Isolates’ analyses includes MLST(7-gene), species confirmation and AMR prediction in addition to sample-specific reports.
Note: You can also run this workflow through command line. However, we only recommend this option for experienced users. For more information, please visit the wf-bacterial-genomes page on GitHub.
Ensure you are using v1.3.0 or newer of the workflow. We recommend running the latest version of our workflows for the best results.
Ensure all parameter options have green ticks.
Flow cell reuse and returns
The Flow Cell Wash Kit protocol is available on the Nanopore Community.
Instructions for returning flow cells can be found here.
Issues during DNA extraction and library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
Inefficient lysis | Enzyme activity has degraded in the solution or the isolate species is hard to lyse. | - Make a fresh enzyme solution. - Follow the hard to lyse gDNA extraction method. - Increase the enzyme incubation for longer than 10 minutes. |
Low DNA concentration | Low input into the extraction method | - Check the cell input used - Add more input and perform the extraction again - Elute in less Elution Buffer - Concentrate the DNA with a 0.4X AMPure XP Bead wash. |
Low DNA integrity number (DIN) | Low quality or concentration of sample input | - Repeat the extraction with freshly made enzyme solution - Concentrate the DNA input with a 0.4X wash |
Low sequencing yield | Low sample concentration | - Concentrate the DNA with a 0.4X AMPure XP wash step to remove potential inhibitors. - Check the DNA concentration and quality. RNA presence may affect quantification of total DNA. |
Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) | The DNA extraction method does not provide the required purity | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. Consider performing an additional SPRI clean-up step. |
Observation | Possible cause | Comments and actions |
---|---|---|
Low recovery | DNA loss due to a lower than intended AMPure beads-to-sample ratio | 1. AMPure beads settle quickly, so ensure they are well resuspended before adding them to the sample. 2. When the AMPure beads-to-sample ratio is lower than 0.4:1, DNA fragments of any size will be lost during the clean-up. |
Low recovery | DNA fragments are shorter than expected | The lower the AMPure beads-to-sample ratio, the more stringent the selection against short fragments. Please always determine the input DNA length on an agarose gel (or other gel electrophoresis methods) and then calculate the appropriate amount of AMPure beads to use. |
Low recovery after end-prep | The wash step used ethanol <70% | DNA will be eluted from the beads when using ethanol <70%. Make sure to use the correct percentage. |
Issues during the sequencing run using a Rapid-based sequencing kit
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | An air bubble was introduced into the nanopore array | After the Flow Cell Check it is essential to remove any air bubbles near the priming port before priming the flow cell. If not removed, the air bubble can travel to the nanopore array and irreversibly damage the nanopores that have been exposed to air. The best practice to prevent this from happening is demonstrated in this video. |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | The flow cell is not correctly inserted into the device | Stop the sequencing run, remove the flow cell from the sequencing device and insert it again, checking that the flow cell is firmly seated in the device and that it has reached the target temperature. If applicable, try a different position on the device (GridION/PromethION). |
MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check | Contaminations in the library damaged or blocked the pores | The pore count during the Flow Cell Check is performed using the QC DNA molecules present in the flow cell storage buffer. At the start of sequencing, the library itself is used to estimate the number of active pores. Because of this, variability of about 10% in the number of pores is expected. A significantly lower pore count reported at the start of sequencing can be due to contaminants in the library that have damaged the membranes or blocked the pores. Alternative DNA/RNA extraction or purification methods may be needed to improve the purity of the input material. The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Script failed" | Restart the computer and then restart MinKNOW. If the issue persists, please collect the MinKNOW log files and contact Technical Support. If you do not have another sequencing device available, we recommend storing the flow cell and the loaded library at 4°C and contact Technical Support for further storage guidance. |
Observation | Possible cause | Comments and actions |
---|---|---|
Pore occupancy <40% | Not enough library was loaded on the flow cell | Ensure the correct concentration of good quality library is loaded on to a MinION/GridION flow cell. To check the concentration, please refer to the library preparation protocol. Please quantify the library before loading and calculate mols using tools like the Promega Biomath Calculator, choosing "dsDNA: µg to pmol" |
Pore occupancy close to 0 | The Rapid Sequencing Kit V14/Rapid Barcoding Kit V14 was used, and sequencing adapters did not attach to the DNA | Make sure to closely follow the protocol and use the correct volumes and incubation temperatures. A Lambda control library can be prepared to test the integrity of reagents. |
Pore occupancy close to 0 | No tether on the flow cell | Tethers are adding during flow cell priming (FCT tube). Make sure FCT was added to FCF before priming. |
Observation | Possible cause | Comments and actions |
---|---|---|
Shorter than expected read length | Unwanted fragmentation of DNA sample | Read length reflects input DNA fragment length. Input DNA can be fragmented during extraction and library prep. 1. Please review the Extraction Methods in the Nanopore Community for best practice for extraction. 2. Visualise the input DNA fragment length distribution on an agarose gel before proceeding to the library prep. In the image above, Sample 1 is of high molecular weight, whereas Sample 2 has been fragmented. 3. During library prep, avoid pipetting and vortexing when mixing reagents. Flicking or inverting the tube is sufficient. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of unavailable pores (shown as blue in the channels panel and pore activity plot) The pore activity plot above shows an increasing proportion of "unavailable" pores over time. |
Contaminants are present in the sample | Some contaminants can be cleared from the pores by the unblocking function built into MinKNOW. If this is successful, the pore status will change to "sequencing pore". If the portion of unavailable pores stays large or increases: 1. A nuclease flush using the Flow Cell Wash Kit (EXP-WSH004) can be performed, or 2. Run several cycles of PCR to try and dilute any contaminants that may be causing problems. |
Observation | Possible cause | Comments and actions |
---|---|---|
Large proportion of inactive/unavailable pores (shown as light blue in the channels panel and pore activity plot. Pores or membranes are irreversibly damaged) | Air bubbles have been introduced into the flow cell | Air bubbles introduced through flow cell priming and library loading can irreversibly damage the pores. Watch the Priming and loading your flow cell video for best practice |
Large proportion of inactive/unavailable pores | Certain compounds co-purified with DNA | Known compounds, include polysaccharides, typically associate with plant genomic DNA. 1. Please refer to the Plant leaf DNA extraction method. 2. Clean-up using the QIAGEN PowerClean Pro kit. 3. Perform a whole genome amplification with the original gDNA sample using the QIAGEN REPLI-g kit. |
Large proportion of inactive/unavailable pores | Contaminants are present in the sample | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. |
Observation | Possible cause | Comments and actions |
---|---|---|
Temperature fluctuation | The flow cell has lost contact with the device | Check that there is a heat pad covering the metal plate on the back of the flow cell. Re-insert the flow cell and press it down to make sure the connector pins are firmly in contact with the device. If the problem persists, please contact Technical Services. |
Observation | Possible cause | Comments and actions |
---|---|---|
MinKNOW shows "Failed to reach target temperature" | The instrument was placed in a location that is colder than normal room temperature, or a location with poor ventilation (which leads to the flow cells overheating) | MinKNOW has a default timeframe for the flow cell to reach the target temperature. Once the timeframe is exceeded, an error message will appear and the sequencing experiment will continue. However, sequencing at an incorrect temperature may lead to a decrease in throughput and lower q-scores. Please adjust the location of the sequencing device to ensure that it is placed at room temperature with good ventilation, then re-start the process in MinKNOW. Please refer to this link for more information on MinION temperature control. |
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.