Version for device: MinION
Overview of the protocol
This end to end protocol describes our automated Nanopore-Only Microbial Isolate Sequencing Solution (NO-MISS): a flexible approach allowing sequencing of up to 24 microbial isolate genomes per MinION Flow Cell, generating a coverage of 50x per genome.
The 50X coverage threshold is sufficient for downstream analysis including: accurate assembly and plasmid resolution, AMR profiling, core genome (cg) and whole genome (wg) multi-locus sequence typing (MLST), and cg/wgSNP typing. Your sequencing data will be analysed by the device using the wf-bacterial-genomes workflow, which produces a user-friendly report of the results.
We provide multiple DNA extraction approaches, depending on requirements, and starting organism (bacteria, fungi/yeast). These are key in achieving reliable flow cell output and genome coverage.
The extracted gDNA is then tagmented and sequenced using our Rapid Barcoding Kit (SQK-RBK114.96). Up to 24 samples per sequencing experiment for bacterial isolates (up to 7 Mb genomes) and up to 8 samples for fungi/yeast isolates can be processed. Bacteria with a genome size of ≤4Mb can expect to generate a minimum coverage of 50x per genome. Please note, coverage may vary depending on larger genome size and input sample quality.
We recommend sequencing up to 72 hours and generating at least 50x coverage per sample (approx. 0.5 Gb per barcode, assuming 5 Mb genome).
You will need to:
We have developed several optimised extraction methods to generate high quality genomic DNA from your cell cultures, allowing maximised sequencing output using the Automated Nanopore-only Microbial Isolate Sequencing Solution (NO-MISS) end-to-end protocol.
What extraction method is right for me?
Using the relevant gDNA extraction method, you will need to lyse your cells and extract your gDNA:
Universal bead-beating method:
Bacteria gDNA extraction:
Hard to lyse organisms gDNA extraction:
The table below is an overview of the steps automated by the ElysION device.
Library preparation step | Process |
---|---|
Sample extraction | |
Bead binding | Addition of the DNA binding bead mix from the MagMAX™ DNA Multi-Sample Ultra 2.0 Kit to your lysed sample(s) |
Clean-up | gDNA purification and clean-up |
Elution | Elution of your extracted and cleaned sample gDNA |
Library preparation | |
DNA barcoding | Tagmentation of the DNA using the Rapid Barcoding Kit V14 |
Sample pooling | Pooling of barcoded samples |
Bead binding | Addition of AMPure XP Beads |
Flow cell priming | Prime the flow cell for sequencing |
Clean-up | Clean-up your samples by performing an Ethanol wash |
Elution | Elute the pooled and cleaned DNA, and attach the sequencing adapters to the DNA ends |
Flow cell loading | Prime the flow cell and load the prepared library for sequencing |
Sequencing | Your sequencing run uses the Gourami software, which will collect raw data to basecall and demultiplex the barcoded reads. |
Data analysis | Once sequencing is complete, the device will perform downstream analysis of the data uing the isolate mode of the wf-bacterial-genomes workflow. |
Flow cell wash or flush | Flow cell is washed using the EXP-WSH004 for reuse or flushed for return |
This protocol should only be used in combination with:
Sample extraction method | Sample type | Sample input | Expected yield | Expected DNA Integrity Number (DIN) | Average sequencing read lengths |
---|---|---|---|---|---|
Universal bead-beating gDNA extraction | Universal applications: bacteria, fungi or yeast |
1 ml liquid overnight culture (~1 x 108 – 109 cfu/ml) or half of a loop of colonies from a plate |
>200 ng/µl per sample | 7-9 | ~4-7 kb |
Bacteria gDNA extraction | Bacterial | 200 µl liquid overnight culture (~1 x 108 – 109 cfu/ml) or 1/8 of a loop of colonies from a plate |
15-20 ng/µl per sample | 9 | >7 kb - Size will vary based on sample input species |
Hard to lyse organisms gDNA extraction | Hard to extract bacterial samples (e.g. Mycobacterium tuberculosis) | For hard to lyse bacterial samples: 200 µl liquid overnight culture (~1 x 108 – 109 cfu/ml) or 1/8 of a loop of colonies from a plate For Mycobacterium tuberculosis: 5 – 10 mg cells from solid or liquid media |
15-40 ng/µl per sample | 8 | >7 kb - Size will vary based on sample input species |
Note: The output and, sequencing read length of extracted DNA may vary depending on sample quality and species. Please ensure you are following the correct method and using high-quality sample inputs.
Staphylococcal Lysis Buffer (SLB) is required for the bacterial gDNA extraction method for staphylococcal inputs.
Reagent | Stock | Final concentration | Volume for 12 samples with excess | Volume for 24 samples with excess |
---|---|---|---|---|
Trizma hydrochloride solution, pH 9 | 1 M | 100 mM | 150 ul | 300 µl |
Sodium chloride | 5 M | 10 mM | 3 ul | 6 µl |
SDS | 10% v/v | 0.1% v/v | 15 ul | 30 µl |
Nuclease-free water | - | - | 1332 ul | 2664 µl |
Total volume | - | - | 1,500 ul | 3,000 µl |
The SDS in the Staphylococcal Lysis Buffer (SLB) is essential for preventing the degradation of staphylococci DNA. Exclusion of SDS from the buffer results in a larger smear of DNA when run on a gel.
Depending on the extraction protocol used, not all third-party reagents are required.
We have validated and recommend the use of all the third-party reagents used in this protocol. Alternatives have not been tested by Oxford Nanopore Technologies.
For all third-party reagents, we recommend following the manufacturer's instructions to prepare the reagents for use.
The number of pores in your flow cell will be checked by the ElysION device at the start of your assay. Flow cells should be checked within 12 weeks of purchasing for MinION/GridION/PromethION.
Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions on the ElysION on-screen display.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 2 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 3 | 1,200 |
Elution Buffer | EB | Black | 1 | 1,500 |
Sequencing Buffer | SB | Red | 1 | 1,700 |
Library Beads | LIB | Pink | 1 | 1,800 |
Library Solution | LIS | White cap, pink label | 1 | 1,800 |
Flow Cell Flush | FCF | Clear | 1 | 15,500 |
Flow Cell Tether | FCT | Purple | 2 | 200 |
Rapid Barcodes | RB01-96 | - | 3 plates | 8 µl per well |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
ElysION NO-MISS sample sheet setup
The sample sheet assigns a barcode to each sample, allowing a user to pick a range from the 96-well plate, and for a sample ID to be tracked from input to results.
For more information on setting up a sample sheet please refer to the ElysION User Guide.
Required field | User input | Definition / field info |
---|---|---|
v1 | Version field, no input required outside of field | |
assay | isoseq-1-24:v1.1 | Assay ID and version |
library_id | User-defined | Identification for the DNA library |
created_by | User-defined | Identification of operator setting up sample sheet |
created_at | User-defined | Date and time of creation. Please follow the format outlined in the ElysION user guide |
sample_count | User-defined | Number of samples processed in run (up to 24 samples) |
well_id | A1-H3 | Positions in 96-well plate being used (up to 24 samples). Samples must start from position A1 in the 96-well plate and run consecutively column-wise. |
barcode | barcode01-barcode96 | Rapid barcodes from SQK-RBK114.96 being used in library preparation. Up to 96 barcodes are available in the sequencing kit, these should be used in sub-sets of up to 24 barcodes, used sequentially. (e.g. Barcodes 01-24, or Barcodes 50-60) |
sample_type | User-defined | Description or characteristics of sample input |
sample_id | User-defined from LIMS system (lims-sampleid-12345) | Identification for each sample input (e.g. from LIMS system) |
Below is an example of a sample sheet CSV file for 10 samples:
Option 1: Universal bead-beating sample preparation for ElysION
This extraction method is a universal gDNA extraction from bacteria and fungi/yeast. It uses vortex bead-beating followed by a proteinase K and RNase treatment.
Note: If your desired yield is not achieved or you are looking to maximise the contiguity of your assemblies by generating longer read lengths, try one of our sample-tailored extraction methods:
Sample extraction method | Sample type | Sample input | Expected yield | Expected DNA Integrity Number (DIN) | Average sequencing read lengths |
---|---|---|---|---|---|
Universal bead-beating gDNA extraction | Universal applications: bacteria, fungi or yeast |
1 ml liquid overnight culture (~1 x 108 – 109 cfu/ml) or half of a loop of colonies from a plate |
>200 ng/µl per sample | 7-9 | ~4-7 kb |
Note: The output and, sequencing read length of extracted DNA may vary depending on sample quality and species. Please ensure you are following the correct method and using high-quality sample inputs.
If using liquid media for your culture: |
---|
1. Centrifuge 1 ml liquid overnight culture (~108 – 109 cfu/ml) at 12,000 xg for 1 minute |
2. Remove the supernatant and resuspend the pellet in 1 ml Phosphate Buffered Saline (PBS) Note: Washing the sample by removing the supernatant and resuspending the pellet in clean Phosphate Buffered Saline (PBS) removes potential inhibitors from the nutrient broth and removes free DNA that may be degraded |
3. Centrifuge at 12,000 x g 1 minute |
4. Remove the supernatant, without disturbing the pellet |
5. Resuspend the pellet in 350 µl bashing bead buffer (Zymo, D6001-3-40) |
6. Add the 350 µl of resuspended sample to a PowerBead Pro tube (Qiagen, 19301) |
7. Take forward the PowerBead Pro tube containing 350 µl of resuspended sample into the next step |
If using solid media for your culture: |
---|
1. Add 350 µl BashingBead Buffer (Zymo, D6001-3-40) to a PowerBead Pro tube (Qiagen, 19301) |
2. Using a sterile 10 µl inoculating loop, pick half a loop worth of colonies, avoiding scratching the agar |
3. Place the inoculating loop with the colonies in the PowerBead Pro tube containing the BashingBead Buffer, and agitate to relieve the cells into the tube |
4. Discard the used inoculating loop |
5. Take forward the PowerBead Pro tube containing 350 µl of resuspended sample into the next step |
Note: Expect ~200–250 µl of supernatant.
Reagent | Volume |
---|---|
Proteinase K | 10 µl |
RNase A | 3 µl |
Reagent | Volume |
---|---|
Proteinase K | 30 µl |
RNase A | 3 µl |
Note: Depending on the bacterial input, the solution may become fully transparent or may remain hazy.
Note: Ensure you keep your samples separate, with each sample being transferred into a separate well of the Nunc™ 96 Well 2 ml DeepWell Plate.
Option 2: Bacterial sample preparation for ElysION
This extraction protocol is recommended for use with general bacterial samples. We have validated this method with Escherichia coli, Klebsiella pneumoniae, Pseudomonas aeruginosa, Serratia marcescens, Enterococcus faecalis, and Bacillus subtilis or staphylococci such as Staphylococcus aureus, and Staphylococcus epidermidis.
If the desired yield is not achieved from your extraction, try one of the following recommendations:
Sample extraction method | Sample type | Sample input | Expected yield | Expected DNA Integrity Number (DIN) | Average sequencing read lengths |
---|---|---|---|---|---|
Bacteria gDNA extraction | Bacterial | 200 µl liquid overnight culture (~1 x 108 – 109 cfu/ml) or 1/8 of a loop of colonies from a plate |
15-20 ng/µl per sample | 9 | >7 kb - Size will vary based on sample input species |
Note: The output and, sequencing read length of extracted DNA may vary depending on sample quality and species. Please ensure you are following the correct method and using high-quality sample inputs.
For general bacterial inputs: |
---|
1. Reconstitute the bacterial lysozyme in TE buffer to 10 mg/ml. Note: The Lysozyme can be stored at 100 mg/ml in a stabilized solution for long-term use as follows: 25 mM Sodium acetate, 50% glycerol 100 mg/ml lysozyme – stored at -20C. |
For staphylococcal inputs: |
---|
1. Prepare the Staphylococcal Lysis Buffer (SLB) (100 mM Tris pH9, 10 mM NaCl, 0.1% SDS). Information for preparing the Staphylococcal Lysis Buffer (SLB) can be found in the Equipment and consumables section of this protocol. |
2. Reconstitute the achromopeptidase in nuclease-free water to 10 mg/ml. |
Tip: We recommend preparing enzymes freshly on the day as enzymatic activity reduces over time once reconstituted.
If using liquid media for your culture: |
---|
1. Centrifuge 200 µl liquid overnight culture (~108 – 109 cfu/ml) at 12,000 x g for 1 minute |
2. Remove the supernatant and resuspend the pellet in 1 ml Phosphate Buffered Saline (PBS) Note: Washing the sample by removing the supernatant and resuspending the pellet in clean Phosphate Buffered Saline (PBS) removes potential inhibitors from the nutrient broth and removes free DNA that may be degraded |
3. Centrifuge at 12,000 x g 1 minute |
4. Remove the supernatant, without disturbing the pellet Note: It is important to remove as much PBS as possible without disturbing the pellet as the buffer has an inhibitory effect on enzyme activity. |
5. Resuspend the pellet depending on your input type: For bacterial samples: resuspend the pellet in 100 µl TE buffer with 10 µl lysozyme (10 mg/ml) For staphylococci samples: Resuspend the pellet in 100 µl SLB with 10 µl achromopeptidase (10 mg/ml) |
If using solid media for your culture: |
---|
1. Add 1 ml of PBS to a fresh 1.5 ml Eppendorf DNA LoBind tube. |
2. Using a sterile 10 µl inoculating loop, pick 1/8 a loop worth of colonies, avoiding scratching the agar |
3. Place the inoculating loop with the colonies in the tube containing the PBS, and agitate to relieve the cells into the tube |
4. Discard the used inoculating loop |
5. Centrifuge at 12,000 x g 1 minute |
6. Remove the supernatant, without disturbing the pellet Note: It is important to remove as much PBS as possible without disturbing the pellet as the buffer has an inhibitory effect on enzyme activity. |
7. Resuspend the pellet depending on your input type: For bacterial samples: resuspend the pellet in 100 µl TE buffer with 10 µl lysozyme (10 mg/ml) For staphylococci samples: Resuspend the pellet in 100 µl SLB with 10 µl achromopeptidase (10 mg/ml). |
If a thermomixer is unavailable, the reaction can be incubated stationary and agitated by gently flicking for 10 seconds every two minutes.
Reagent | Volume |
---|---|
Proteinase K | 10 µl |
RNase A | 3 µl |
Total (including your sample) | 113 µl |
Note: Depending on the bacterial input, the solution may become fully transparent or may remain hazy.
Note: Ensure you keep your samples separate, with each sample being transferred into a separate well of the Nunc™ 96 Well 2 ml DeepWell Plate.
Option 3: Hard to lyse organisms sample preparation for ElysION
This extraction method is recommended for hard to lyse organisms, such as Mycobacterium tuberculosis, or for bacteria that was not successfully extracted using the Bacteria gDNA extraction method.
Sample extraction method | Sample type | Sample input | Expected yield | Expected DNA Integrity Number (DIN) | Average sequencing read lengths |
---|---|---|---|---|---|
Hard to lyse organisms gDNA extraction | Hard to extract bacterial samples (e.g. Mycobacterium tuberculosis) | For hard to lyse bacterial samples: 200 µl liquid overnight culture (~1 x 108 – 109 cfu/ml) or 1/8 of a loop of colonies from a plate For Mycobacterium tuberculosis: 5 – 10 mg cells from solid or liquid media |
15-40 ng/µl per sample | 8 | >7 kb - Size will vary based on sample input species |
Note: The output and, sequencing read length of extracted DNA may vary depending on sample quality and species. Please ensure you are following the correct method and using high-quality sample inputs.
If using liquid media for your culture: |
---|
1. Centrifuge 200 µl liquid overnight culture (~108 – 109 cfu/ml) at 12,000 x g for 1 minute. |
2. Remove the supernatant without disturbing the pellet. |
If using solid media for your culture: |
---|
1. Using a sterile 10 µl inoculating loop, collect 1/8 a loop worth of colonies into a fresh 1.5 ml Eppendorf tube, avoiding scratching the agar |
IMPORTANT: For Mycobacterium tuberculosis, we recommend avoiding overgrowing the cultures:
We also strongly advise harvesting the recommended amount of input material to yield the correct amount of gDNA for the library preparation step.
Less than 5 mg will not yield enough gDNA for library preparation (less than 10 ng/µl), negatively impacting sequencing output.
More than 10 mg may result in carryover of sequencing inhibitors onto the flow cell, also negatively impacting sequencing output.
For Mycobacterium tuberculosis solid media: |
---|
1. Harvest 5-10 mg of cells from your Mycobacterium tuberculosis solid media and transfer to a 1.5 ml Eppendorf DNA LoBind tube. |
For Mycobacterium tuberculosis liquid media: |
---|
1. Add ~500 µl of Mycobacterium tuberculosis liquid media to a fresh 1.5 ml Eppendorf DNA LoBind tube. |
2. Centrifuge at 12,000 x g 1 minute |
3. Remove the supernatant, without disturbing the pellet. |
4. Harvest 5-10 mg of cells from the pellet and transfer to a 1.5 ml Eppendorf DNA LoBind tube. |
If the pipette tip becomes clogged, gently press the pipette tip down on the bottom of the tube to break up clumps while pipetting up and down.
We do not recommend increasing the temperature of the incubation as this may negatively impact DNA quality and sequencing data.
This heated incubation is important for efficient cell lysis.
Vortexing with beads helps homogenise the sample to improve efficiency of later steps. Using 4 mm beads avoids overshearing of the DNA.
Light agitation prevents aggregation and precipitation in the reaction and allows the enzyme to access the cell walls.
Constant shaking is essential to achieve sufficient lysis within 15 minutes and occasional agitation may not prevent the cells from clumping.
Note: If a thermomixer is not available, briefly vortex at regular intervals of 2 minutes.
Reagent | Volume |
---|---|
Proteinase K | 10 µl |
RNase A | 3 µl |
Total (including your sample) | 123 µl |
Note: Depending on the bacterial input, the solution may become fully transparent or may remain hazy.
The high salt step reduces yield but is essential to remove any inhibitors of the library preparation.
Note: Ensure you keep your samples separate, with each sample being transferred into a separate well of the Nunc™ 96 Well 2 ml DeepWell Plate.
Please ensure you have all the correct hardware components, software packages and workflows installed to carry out this method.
For more information on the device and the processes please refer to the ElysION User Guide.
Reagent | 1. Thaw at room temperature | 2. Briefly spin down | 3. Mix well by pipetting |
---|---|---|---|
Rapid Barcode Plate (RB01-96) | Not frozen | ✓ | ✓ |
Rapid Adapter (RA) | Not frozen | ✓ | ✓ |
AMPure XP Beads (AXP) | ✓ | ✓ | Mix by pipetting or vortexing immediately before use |
Elution Buffer (EB) | ✓ | ✓ | ✓ |
Adapter Buffer (ADB) | ✓ | ✓ | Mix by vortexing |
Flow Cell Flush (FCF) | ✓ | ✓ | ✓ |
Flow Cell Tether (FCT) | ✓ | ✓ | ✓ |
Sequencing Buffer (SB) | ✓ | ✓ | ✓ |
Library Beads (LIB) | ✓ | ✓ | Mix by pipetting or vortexing immediately before use |
Wash Mix (WMX) | ✓ | ✓ | ✓ |
Wash Diluent (DIL) | ✓ | ✓ | Mix by vortexing |
Storage Buffer (S) | ✓ | ✓ | Mix by vortexing |
Bovine Serum Albumin (BSA) | ✓ | ✓ | ✓ |
Note: The volumes will change depending on the number of samples being processed.
For runs with 1-9 samples
Use the Sarstedt Screw Cap Micro tube 2 ml, with 1980 µl DNA Binding Bead Mix in each tube per 3 samples.
Reagent | Volume for 1-3 samples (1x Sarstedt tubes) | For 4-6 samples (2x Sarstedt tubes) | For 7-9 samples (3x Sarstedt tubes) |
---|---|---|---|
Lysis/Binding Solution | 1.8 ml | 3.6 ml | 5.4 ml |
Binding Beads | 180 µl | 360 µl | 540 µl |
Total volume | 1980 µl | 3960 µl | 5940 µl |
For runs with 10-24 samples
Use the reservoir plate, with 6 ml of DNA Binding Bead Mix + 600 µl additional volume per sample being processed.
Reagent | For 10 samples (Reservoir Plate) | For 12 samples (Reservoir Plate) | For 18 samples (Reservoir Plate) | For 24 samples (Reservoir Plate) |
---|---|---|---|---|
Lysis/Binding Solution | 5,455 µl + 5,455 µl = 10,910 µl |
5,455 µl + 6,545 µl = 12,000 µl |
5,455 µl + 9,818 µl = 15,273 µl |
5,455 µl + 13,090 µl = 18,545 µl |
Binding Beads | 545 µl + 545 µl = 1,090 µl |
545 µl + 655 µl = 1,200 µl |
545 µl + 982 µl = 1,527 µl |
545 µl + 1,310 µl = 1,855 µl |
Total volume | 6,000 µl + 6,000 µl = 12,000 µl |
6,000 µl + 7,200 µl = 13,200 µl |
6,000 µl + 10,800 µl = 16,800 µl |
6,000 µl + 14,400 µl = 20,400 µl |
If the application is not running, double click the "Background Services" desktop application on your ElysION device.
Failure to ensure the deck is clear can lead to errors in checks or damage to the equipment.
Note: The ElysION device will perform automated checks.
Allow the initialisation checks to complete before proceeding with the library preparation method.
Click "next" to proceed.
Instructions for inserting a flow cell onto the ElysION device can be found in the ElysION User Guide.
Once flow cell check begins, click next to proceed.
Once completed click "next" to proceed.
Reagent | Volume |
---|---|
Rapid Adapter (RA) | 1.5 μl |
Adapter Buffer (ADB) | 3.5 μl |
Total | 5 μl |
Once completed click "next" to proceed.
Note: Reagent volumes and guidance for flow cell wash or flush will be given on the deck loading page. If you would like to perform a flow cell wash for re-use it is recommended to thaw the wash reagents at the same time as the library preparation reagents.
For more information on the run conditions please refer to the ElysION User Guide.
Please consider the following when preparing and loading your reagents and consumables into the ElysION device to mitigate risk of workflow failure and equipment damage:
Note: The position of the tip boxes and reagents will change depending on the sample count, and run setup. Please ensure you are following the instructions on the on-screen display correctly for your run.
The analysis pipeline used is the wf-bacterial-genomes.
For more information on how to access your sequencing data and the analysis outputs please consult the ElysION User Guide.
Issues during DNA extraction and automated library preparation
We also have an FAQ section available on the Nanopore Community Support section.
If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
Observation | Possible cause | Comments and actions |
---|---|---|
Inefficient lysis | Enzyme activity has degraded in the solution or the isolate species is hard to lyse. | - Make a fresh enzyme solution. - Follow the hard to lyse gDNA extraction option. - Increase the enzyme incubation for longer than 10 minutes. |
Low DNA concentration | Low input into the extraction method | - Check the cell input used - Add more input and perform the extraction again |
Low DNA integrity number (DIN) | Low quality or concentration of sample input | - Repeat the extraction with freshly made enzyme solution |
Low sequencing yield | Low sample concentration | - Check the DNA concentration and quality. RNA presence may affect quantification of total DNA. |
Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) | The DNA extraction method does not provide the required purity | The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover. Consider performing an additional SPRI clean-up step. |
For commonly encountered issues please refer to the ElysION User Guide.
For in-depth troubleshooting please refer to the Set-up and operating manual: Early access ElysION provided with your ElysION device.
For additional customer support contact the nanoporetech support channels.
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.