Version for device: MinION
Overview of the protocol
The ElysION automated Rapid Barcoding Kit 96 V14 (SQK-RBK114.96) workflow is optimised for simplicity and speed to automatically prepare and sequence between 1 and 96 purified DNA input samples, then automatically wash the used flow cell for re-use. The workflow is PCR-free, removing the PCR bias and retains information about base modifications, which can be analysed using tools developed in the Nanopore Community.
Note: This kit is able to generate high sequencing accuracies of over 99% (Q20+) and duplex data; but has been optimised for speed and simplicity. If users require high accuracy, we recommend using our Native Barcoding Kit 96 V14 kit (SQK-NBD114.96).
Due to the simple nature of the workflow and the fact that little sample manipulation is required (e.g. minimal pipetting steps and no clean-up steps) long reads can be achieved with this kit, despite the required transposase fragmentation. However, in order for long reads to be observed in sequencing, long fragments need to be present in the initial sample input.
You will need to:
The table below is an overview of the steps automated by the ElysION device.
Library preparation step | Process |
---|---|
DNA barcoding | Tagmentation of the DNA using the Rapid Barcoding Kit V14 |
Sample pooling and clean-up | Pooling of barcoded libraries and AMPure XP Bead clean-up |
Adapter ligation | Attach the sequencing adapters to the DNA ends |
Priming and loading the flow cell | Prime the flow cell and load the prepared library for sequencing |
Sequencing | Your sequencing run uses the Gourami software, which will collect raw data to basecall and demultiplex the barcoded reads. |
Data analysis | Once sequencing is complete, we have a number downstream analysis options for your data available through EPI2ME. |
This protocol should only be used in combination with:
The starting concentration of your sample(s) should be 11 ng/µl in 30 µl of volume.
Note: The output and, sequencing read length of extracted DNA may vary depending on sample quality and species. Please ensure you are using high-quality sample inputs.
Depending on the extraction protocol used, not all third-party reagents are required.
We have validated and recommend the use of all the third-party reagents used in this protocol. Alternatives have not been tested by Oxford Nanopore Technologies.
For all third-party reagents, we recommend following the manufacturer's instructions to prepare the reagents for use.
It is important that the input DNA meets the quantity and quality requirements. Using too little or too much DNA, or DNA of poor quality (e.g. highly fragmented or containing RNA or chemical contaminants) can affect your library preparation.
For instructions on how to perform quality control of your DNA sample, please read the Input DNA/RNA QC protocol.
Depending on how the DNA is extracted from the raw sample, certain chemical contaminants may remain in the purified DNA, which can affect library preparation efficiency and sequencing quality. Read more about contaminants on the Contaminants page of the Community.
The number of pores in your flow cell will be checked by the ElysION device at the start of your assay. Flow cells should be checked within 12 weeks of purchasing for MinION/GridION/PromethION.
Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions on the ElysION on-screen display.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
Name | Acronym | Cap colour | No. of vials | Fill volume per vial (µl) |
---|---|---|---|---|
Rapid Adapter | RA | Green | 2 | 15 |
Adapter Buffer | ADB | Clear | 1 | 100 |
AMPure XP Beads | AXP | Amber | 3 | 1,200 |
Elution Buffer | EB | Black | 1 | 1,500 |
Sequencing Buffer | SB | Red | 1 | 1,700 |
Library Beads | LIB | Pink | 1 | 1,800 |
Library Solution | LIS | White cap, pink label | 1 | 1,800 |
Flow Cell Flush | FCF | Clear | 1 | 15,500 |
Flow Cell Tether | FCT | Purple | 2 | 200 |
Rapid Barcodes | RB01-96 | - | 3 plates | 8 µl per well |
This Product Contains AMPure XP Reagent Manufactured by Beckman Coulter, Inc. and can be stored at -20°C with the kit without detriment to reagent stability.
ElysION RBK114 sample sheet setup
The sample sheet assigns a barcode to each sample, allowing a user to pick a range from the 96-well plate, and for a sample ID to be tracked from input to results.
For more information on setting up a sample sheet please refer to the ElysION User Guide.
Required field | User input | Definition / field info |
---|---|---|
v1 | Version field, no input required outside of field | |
assay | rbk:v1.0 | Assay ID and version |
library_id | User-defined | Identification for the DNA library |
created_by | User-defined | Identification of operator setting up sample sheet |
created_at | User-defined | Date and time of creation. Please follow the format outlined in the ElysION user guide. |
sample_count | User-defined | Number of samples processed in run (1 to 96 samples) |
well_id | A1-H12 | Positions in 96-well plate being used (1 to 96 samples). Samples must start from position A1 in the 96-well plate and run consecutively column-wise. |
barcode | barcode01-barcode96 | Rapid barcodes from SQK-RBK114.96 being used in library preparation. Up to 96 barcodes are available in the sequencing kit, these should be used secuentially. (e.g. Barcodes 01-96, Barcodes 01-24, or Barcodes 50-60) |
sample_type | User-defined | Description or characteristics of sample input |
sample_id | User-defined from LIMS system (lims-sampleid-12345) | Identification for each sample input (e.g. from LIMS system) |
Below is an example of a sample sheet CSV file for 10 samples:
Please ensure you have all the correct hardware components, software packages and workflows installed to carry out this method.
For more information on the device and the processes please refer to the ElysION User Guide.
Reagent | 1. Thaw at room temperature | 2. Briefly spin down | 3. Mix well by pipetting |
---|---|---|---|
Rapid Barcode Plate (RB01-96) | Not frozen | ✓ | ✓ |
Rapid Adapter (RA) | Not frozen | ✓ | ✓ |
AMPure XP Beads (AXP) | ✓ | ✓ | Mix by pipetting or vortexing immediately before use |
Elution Buffer (EB) | ✓ | ✓ | ✓ |
Adapter Buffer (ADB) | ✓ | ✓ | Mix by vortexing |
Flow Cell Flush (FCF) | ✓ | ✓ | ✓ |
Flow Cell Tether (FCT) | ✓ | ✓ | ✓ |
Sequencing Buffer (SB) | ✓ | ✓ | ✓ |
Library Beads (LIB) | ✓ | ✓ | Mix by pipetting or vortexing immediately before use |
Bovine Serum Albumin (BSA) | ✓ | ✓ | ✓ |
Wash Mix (WMX) | ✓ | ✓ | ✓ |
Wash Diluent (DIL) | ✓ | ✓ | Mix by vortexing |
Storage Buffer (S) | ✓ | ✓ | Mix by vortexing |
If the application is not running, double click the "Background Services" desktop application on your ElysION device.
Failure to ensure the deck is clear can lead to errors in checks or damage to the equipment.
Note: The ElysION device will perform automated checks.
Allow the initialisation checks to complete before proceeding with the library preparation method.
Click next to proceed.
Instructions for inserting a flow cell onto the ElysION device can be found in the ElysION User Guide.
Once flow cell check begins, click next to proceed.
Once completed click "next" to proceed.
Reagent | Volume |
---|---|
Rapid Adapter (RA) | 1.5 μl |
Adapter Buffer (ADB) | 3.5 μl |
Total | 5 μl |
Once completed click "next" to proceed.
Note: Reagent volumes and guidance for flow cell wash or flush will be given on the deck loading page. If you would like to perform a flow cell wash for re-use it is recommended to thaw the wash reagents at the same time as the library preparation reagents.
For more information on the run conditions please refer to the ElysION User Guide.
Please consider the following when preparing and loading your reagents and consumables into the ElysION device to mitigate risk of workflow failure and equipment damage:
Note: The position of the tip boxes and reagents will change depending on the sample count, and run setup. Please ensure you are following the instructions on the on-screen display correctly for your run.
Issues during the automated library preparation
For commonly encountered issues please refer to the ElysION User Guide.
For in-depth troubleshooting please refer to the Set-up and operating manual: Early access ElysION provided with your ElysION device.
For additional customer support contact the nanoporetech support channels.
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.