Device information
The MinION Mk1C combines the real-time, rapid, portable sequencing of a MinION Mk1B with a Graphical Processing Unit (GPU) and a high-resolution screen.
The device applies the technology of the MinION Mk1B with the basecalling power of the MinIT and the additional features of smartphone, such as a cellular modem and touch screen.
Component | Specification |
---|---|
Model Number | MIN-101C |
Size and weight | H 32 mm x L 142 mm x W 118 mm; 440 g |
Power | • Supplied with a 6.3-19.6 VDC power supply • Max rated current 10 A • Max rated power 60 W |
Compute spec | 1 TB SSD Storage, 8 GB RAM, GPU embedded analysis accelerator (ARM processor 6 cores, 256 core GPU) |
Connections | • 1x USB 2.0 • 1x eSATA (currently non-functional) • 1x microSD card • 1x 1 GB Ethernet • Wi-Fi connectivity |
Pre-loaded software | Linux OS, MinKNOW and Dorado |
Security | Kensington Nano Security Slot |
Environmental conditions | • Designed to sequence at +10°C to +30°C* • Do not cover vents on the sides of the device • Indoor use • Altitude up to 2,000 metres • Maximum relative humidity 80% for temperatures up to 31°C, decreasing linearly to 50% relative humidity at 40°C • Applicable pollution degree 2 in the intended environment |
* Functional range of electronics +5°C to +40°C
MinION Mk1C IT requirements
MinION Mk1C technical specification
Safety and regulatory information
Our MinION Mk1C FAQs are located here.
What's in the box
The MinION Mk1C is shipped with:
- Configuration Test Cell (CTC)
- Power adapter
- Plugs compatible with UK, US, EU and Australian sockets
- Quick Start Guide
Touch screen
The touch screen allows the user to interact with the MinION Mk1C software directly on the device.
LED lights
There are five LED lights to the right of the touch screen. All five LEDs illuminate white when the device starts up and then cycle as the system software loads. The cycling indicates that the device is working. When a USB is inserted into the MinION Mk1C, all LEDs illuminate and flash three times before the normal cyclical pattern resumes.
Lid
The lid can be flipped open to access the flow cell housing.
Clip
The clip holds the flow cell or Configuration Test Cell securely in place.
Configuration Test Cell (CTC)
The CTC is used during the hardware check to ensure that the communication between the device and the flow cell is working correctly.
Powering the device
The power requirements for the MinION Mk1C using the AC/DC adapter supplied with the device are detailed below.
Power supply | Power input | Power output |
---|---|---|
AC/DC adapter | Standard Mains input: 100-240 VAC 50/60 Hz 1.5 A |
Output: 6.3-19.6 VDC Max rated current 10 A Max rated power 60 W Estimated power dissipation: 20-25 W at room temperature |
Oxford Nanopore Technologies does not supply a portable power supply for this purpose and as such it is not officially supported. If you wish to make use of a portable power supply, then it must be compatible with the below specification.
DC power pack requirements
Power supply | Power input | Power output |
---|---|---|
DC power pack | Standard Mains input for charging: 100-240 VAC 50/60 Hz |
Estimated power dissipation: 20-25 W at room temperature |
If you experience login issues, contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.
To log in, you must be connected to the internet.
For more information see the Onboarding process section in this document.
Onboarding process
The onboarding process is available when a MinION Mk1C is first switched on to guide users through setting up a network connection, logging into MinKNOW, enabling remote access and running a hardware check.
Please visit our Oxford Nanopore FAQ and Help Centre if you are having difficulties updating MinKNOW on the MinION Mk1C.
We strongly recommend users to connect their device to the internet to allow them to log into their MinKNOW account.
There are two options:
Connect using WiFi: click WiFi and select the network to join and enter password, if prompted.
Connect using Ethernet: plug-in the Ethernet cable and click Ethernet.
Note: If your network configuration requires you to use a static IP address, click Offline and configure the network through the main Network Settings page.
These settings can be altered again using the Host settings after the onboarding flow has ended.
A hardware check only needs to be completed once. However, it can be skipped for users who will not be using their device for sequencing.
General connection information
There are several options to connect to your MinION Mk1C. These are:
General Options
Advanced users
MinION Mk1C sequences and basecalls data via the integrated sequencing device and Oxford Nanopore Technologies basecaller - Dorado - in MinKNOW. POD5, FASTQ and BAM files are output and stored locally on the MinION Mk1C in the Data folder.
This folder and the output files can be accessed through File Manager in MinKNOW. For more information, please see the Data management section of this document.
Changing the default password and hostname
<new_hostname>
portion with your desired hostname (omitting the angle bracket characters <>
):sudo -S hostnamectl set-hostname <new_hostname>
sudo shutdown -h now
Network settings and connection
Both options are supported, however if Ethernet is available, we recommend using Ethernet because the connection will likely be quicker and more reliable.
To connect using Ethernet:
1. Connect a Ethernet cable to the device
2. Navigate to network settings in host settings to confirm connection. IP configuration information will appear when connected.
To connect using Wi-Fi, navigate to network settings and toggle Wi-Fi on. Select a network from the Available Networks panel to connect to the internet.
Select Advanced to open Advanced Wi-Fi Configurations. From here, all available networks can be viewed, connected or removed.
The static IP address can be configured through MinKNOW UI for both WiFI and Ethernet. Below is an example for Ethernet.
This can be used to connect to the MinION Mk1C in places where there is no network infrastructure, e.g. in the field. To enable hotspot, select the toggle to ON. When active, the device hotspot name will be displayed.
Hotspot configs can be edited to select a channel if required for congested networks where there may be competition for channels.
1. Navigate to the network settings
2. For hotspot, select the toggle to ON
3. Click 'Configure hotspot'
4. Select channel using the dropdown menu
5. Enter new password
6. Click 'Confirm'
The hotspot should be confirmed as specified in the hotspot configs.
Login with the Security Key:
WarmButterflyWings98
Identify the MC-XXXXXX ID in the Network Name under Network Panel (the final option under the Wi-Fi icon), which is show in the window below (left)‡. Selecting the MinION Mk1C ID and clicking Apply will present the Wi-Fi login pop-up. Use the password credentials shown below.
Login with the Password:
WarmButterflyWings98
MC-XXXXXX (XXXXXX is the MinION Mk1C ID) will be displayed as a server option. Click the MC-XXXXXX icon, click Connect as under the Search bar and enter the network credentials.
‡ When connecting to your device by SSH, it can be addressed in three different ways:
In most organisations, central DNS servers are used. If your device is using a 'static' IP configuration, your IT department will have to create a DNS entry for it manually. If your device is using a dynamic (DHCP) IP configuration, the DNS entry will often be created automatically. To SSH into your device using DNS, set the hostname in your client to the device's hostname, e.g. MC-112345.
If DNS is not provided by your organisation or your broadband router, try using mDNS. When your device is connected to a network, it announces its presence to all other nearby devices such as your laptop or PC. If your laptop or PC is configured to recognise these announcements, you can SSH into your device using mDNS by setting the hostname in your client to the device's hostname followed by ".local", e.g. MC-112345.local.
If the above methods are not successful, use the devices IP address directly. The MinKNOW UI shows the IP addresses for each interface in the Settings - Network page. If your device has LAN and Wi-Fi interfaces, each will have a different IP address. To SSH into your device using its IP address, set the hostname in your client to the IP address, e.g. 10.20.30.41.
Identify and click on the MC-XXXXXX ID under the Wi-Fi network options icon (shown above). Enter the Wi-Fi password (shown below) in the pop-up Wi-Fi Network Authentication window:
Login with the Password:
WarmButterflyWings98
The following window will be presented when you have connected to the MinION Mk1C Wi-Fi router.
Open the File Browser (this can be found using the Search tool). Select Connect to server on the bottom of the list in the left panel. You will need to enter:
smb://mc-110013
Mount network drive
The MinION Mk1C contains 1 TB SSD storage, which provides an average capacity of roughly 50 Gbases, stored in .POD5, .fast5 and/or FASTQ format. Once the SSD fills up, you will need to move your sequencing data off the device to a networked storage location. Network drives for data storage can be mounted from either the Device Settings or the File Manager in the Network tab:
Device Settings:
File manager:
Use SMB for connecting to a shared drive on a Windows server, and NFS for Linux.
If your shared drive is on a macOS machine or NAS (Network-Attached Storage), you can use either SMB or NFS.
SMB:
NFS:
/data
./data/network/mt-111111-data
for example.Providing the settings were correct, the mounted drive will be added to the 'Disk Management' overview. The user will be able to:
/data
./data/network/mt-111111-data
for example.Providing the settings were correct, the mounted drive will be added to the 'Disk Management' overview. You will be able to:
(Advanced) Connecting by the command-line
The MinION Mk1C has two interfaces that we support for data transfer. These are:
Networking with MinION Mk1C. Blue lines indicate a connection via the internet. The black line indicates a wired connection. The computer on the LAN (B.) can be used to configure the device using parameters that are not available through the Admin interface. You can also configure the MinION Mk1C using the Wi-Fi connected device (A.), however network compatibility issues may inhibit the update.
Internet connection:
This configuration requires an ethernet connection to the Local Area Network (LAN) from the MinION Mk1C. The LAN is connected to the internet, which enables updating of MinKNOW on the device.
Interface connection:
An interface, such as a laptop/mobile device/computer, can be wirelessly connected to the MinION Mk1C. Alternatively, the MinION Mk1C may be accessed and controlled by a computer on the LAN.
Other configurations are also possible, however they are not supported by Oxford Nanopore Technologies:
Connecting by the command line via SSH or the terminal enables several options:
/data
If SSH is enabled in your Windows build, you can use the built-in Windows Command Prompt (CMD). Open a CMD prompt, then enter ssh minit@hostname
(e.g. ssh minit@mc-111414
):
If it is your first time connecting, you will be asked to confirm the authenticity of the host. Enter yes
to continue:
Enter the password to establish an SSH connection. The default password is minit
. Please note that characters will not appear on-screen as you type in the password.
If the SSH connection is successfully established, the shell prompt will now read minit@hostname:~$
A third-party SSH client such as Bitvise (example images shown below) or Putty may be used. The following default options can be entered:
Host – device ID or IP address (e.g. MC-111414)
Port – 22
Username – minit
Initial method – password
Password – minit
If a connection is successfully established, Bitvise will automatically open a terminal window and an SFTP browser. Alternatively, you can click on New terminal console or New SFTP window on the left-hand side:
Open a Terminal, then type ssh minit@hostname
(e.g. ssh minit@mc-111414
):
If it is your first time connecting, you will be asked to confirm the authenticity of the host. Enter yes
to continue:
Enter the password to establish an SSH connection. The default password is minit
. Please note that characters will not appear on-screen as you type in the password.
If the SSH connection is successfully established, the shell prompt will now read minit@hostname:~$
(Advanced) Network configuration - static vs dynamic IP
The LAN interface on the MinION Mk1C can operate in either DHCP (dynamic) mode or static mode. When in DHCP mode, the settings are acquired from the network’s DHCP server. Some networks do not have a DHCP server or you may wish to override the settings it provides. In such circumstances, the static mode can be used.
The static mode comes pre-configured with the following settings:
IP Address/Subnet Mask: 192.168.0.1/24
Default Gateway: 192.168.0.1
DNS Servers: 9.9.9.9, 8.8.8.8
Here is an example of possible static settings that may be provided by your IT department:
IP Address: 10.20.30.40
Subnet Mask: 255.255.255.128
Default Gateway: 10.20.30.1
DNS Servers: 10.0.0.53, 10.0.1.53
To change the static profile to match these settings:
xxx.xxx.xxx.xxx
, convert this to CIDR format, which is a number between 1 and 32 (common values are 16, 24 and 25). You can use the following website to convert from xxx.xxx.xxx.xxx
to a CIDR value: https://docs.netgate.com/pfsense/en/latest/network/cidr.htmlxxx.xxx.xxx.xxx
value in the first column and find the corresponding CIDR value in the second column. In the example above, 255.255.255.128
corresponds to a CIDR value of /25
.sudo nmcli con down dhcp
sudo nmcli con down static
sudo nmcli con mod static ipv4.method manual ipv4.address 10.20.30.40/25 ipv4.gateway 10.20.30.1 ipv4.dns 10.0.0.53,10.0.1.53 ipv4.routes "" connection.autoconnect yes
sudo nmcli con up static
In some instances, these commands may give an error message, however this is not a problem and the error should be ignored.
The addresses specified in the third command line are the values that need to be changed to match particular settings. Note that there is no space between the DNS server addresses.
Use the following command to verify that your settings are correctly applied:
sudo nmcli dev show eth0
Open a terminal window and enter the following commands:
sudo nmcli con down static
sudo nmcli con up dhcp
Hardware check
A hardware check must be performed on all new devices or when software has been updated. This uses the Configuration Test Cell(s) (CTC), that come pre-inserted into your device in place of flow cells.
Note: If using a Flongle Flow Cell, we recommend regularly checking the Flongle adapter by inserting the empty adapter and completing a hardware check. For checking the device or flow cell position, use a MinION CTC, even if a Flongle Flow Cell will be used for sequencing.
The hardware check page will open.
Note: To check the Flongle adapter, insert the EMPTY adapter. To check device or position, insert only a MinION CTC.
A loading bar will be displayed under the flow cell during the check.
Hardware check pass is indicated by a green check icon.
An orange check icon is a fail.
If the hardware check fails, remove and reinsert the CTC, and run a hardware check again. If the check fails for a second time, please contact Technical Support via email (support@nanoporetech.com).
Flow cell check
We strongly recommend performing a flow cell check before loading a DNA or RNA library to assess the number of pores available.
Oxford Nanopore Technologies will replace any flow cell that falls below the warranty number of active pores within three months of purchase, provided that you report the results within two days of performing the flow cell check and you have followed the storage recommendations.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell (FLO-FLG001) | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
A loading bar will be displayed under the flow cell during the check.
The quality of the flow cell will be shown as one of the three outcomes:
Yellow exclamation mark (Flongle flow cell): The number of sequencing pores is below warranty. Take the flow cell out of the device, re-insert it and run a flow cell check again. If the flow cell is still below warranty, contact support@nanoporetech.com
Green tick (MinION flow cell): The number of sequencing pores is above warranty.
Question mark (PromethION flow cell): A flow cell check has not been run on the flow cell during this MinKNOW session.
Note: The indicator of quality will only remain visible during a MinKNOW session. Once the MinKNOW session has ended, the status of the flow cell will be erased.
Sequencing and monitoring the experiment
To run a sequencing experiment, you will need your DNA or RNA sample purified and a sequencing library prepared. For more information about sample preparation, refer to the Prepare section of our Documentation.
You can access the MinKNOW protocol by following this link.
Data management
Data can be managed and transferred from the file manager tab on the host settings.
Navigate through the tabs to view the data stored:
Data will now be able to be shared and accessible on different networks.
Note: The image above is an example of sharing on a GridION. On MinION devices, 'minit' will appear and PromethION devices will show 'prom' when active.
Note: The password will be removed and must be reset when sharing is turned back on.
Note: a USB drive or SD card plugged in will appear under the Devices folder.
Click the icon in the bottom right corner to open Create New Directory and type in a directory name. Click Create.
Navigate to Device settings to view availabe disk space on the device.
Any peripherals plugged in, including USB, can be ejected using the Eject option.
A network drive can also be connected to the device by selecting Mount Network Drive and filing in the required credentials for connecting using a Samba (SMB) server or a Network File System (NFS).
Compatible disk formats:
SD and USB on MinION Mk1C supports DOS FAT, Windows 95 FAT, exFAT, and Linux ext(2|3|4).
Please refer to the 'Mount Network Drive' section of this protocol for more information.
Data transfer via SFTP
For instructions, see the (Advanced) connecting by the command-line section of this user manual.
You can use software like Filezilla, which enables data transfer for mounted hard drives on Linux, Windows, and macOS. It may also transfer using the SFTP protocol which is available whenever an SSH connection can be established.
Filezilla can be downloaded for various OS to enable data transfer from your MinION Mk1C.
The following default options can be entered:
Host – device ID or IP address (e.g. MC-111414)
Username – minit
Password – minit
Port – 22
If it is your first time connecting, you will be asked to confirm the authenticity of the host. Check the box to Always trust this host, add this key to the cache and click OK to continue:
If the SFTP connection is successfully established, you will see the Mk1C file directory shown on the right-hand side of the UI. Please note that Filezilla defaults to the home directory (/home/minit
):
You can navigate to the /data directory by entering /data
into the Remote site field and clicking Enter:
Rebooting and shutting down
You will need to reconnect the MinION Mk1C to any device that was connected before the reboot using the procedure described in the Network settings and connection section of this document.
You will initially see some scrolling code. After ~10 seconds, the screen will go blank and the lights at the side of the screen will power down. It is safe to remove the power cable from the MinION Mk1C at this point, if shutting down.
Updating the software via the UI
MinKNOW is the software used on nanopore sequencing devices. The software is described in more detail in the MinKNOW Technical Document. The MinKNOW Protocol provides instructions on how to use the software.
Oxford Nanopore Technologies will post a release note when new software or updates go out via the Nanopore Community.
We urge users to update as soon as reasonably possible after the release has been made available.
Note: Please check the information for firewall access to the necessary IP addresses and ranges in this section.
Users will not be able to update their device if connected remotely or if a run is in progress.
The Oxford Nanopore Technologies software will require access to the AWS IP ranges currently listed here:
http://docs.aws.amazon.com/general/latest/gr/aws-ip-ranges.html
Access to the following IP addresses is also needed:
178.79.175.200
96.126.99.215
Select Get Update to update the device software automatically.
Updates may be skipped. However, we recommend to update the device as soon as updates are available. Some updates will be mandatory to use the device and are unable to be skipped.
Note: For MinION Mk1B, clicking Get Update will open the Software Downloads page on the connected computer for you to download the updated MinION software.
An update button will only appear when a new version of the software is available to download.
Note: For MinION Mk1B devices, clicking Get Update will open the Software Downloads page on the connected computer for you to select and download the updated MinION software.
A change log for the update is available in this dialogue box.
(Advanced) Updating the software via SSH
Always refer to the release notes for any version-specific commands that may be supplied.
sudo apt clean
sudo apt update
sudo apt install ont-mk1c-release
These commands should complete successfully, without any errors or warnings. If you do see an error or warning message, check your MinION Mk1C is able to connect to the Internet. Run the following command to diagnose network connection problems:
sudo /opt/ont/mooneye/bin/ont-mooneye-check-network --verbose --diagnose
Once the commands have completed successfully without errors, please reboot the device using the command:
sudo reboot
export DEBIAN_FRONTEND=noninteractive
sudo apt upgrade
export DEBIAN_FRONTEND=noninteractive
sudo apt install ont-mk1c-release
OS update/device recovery using microSD cards
If you meet the requirements below, you can update your device before the microSD card arrives.
Note: there is no guarantee this process will work, as not all vendors of microSD cards have been tested internally at Oxford Nanopore.
If your MinION Mk1C serial number is greater than MC-112454, your device shipped with Ubuntu 18.04 (Bionic).
Alternatively, on the user interface navigate to Settings and then Software. Click on the ‘I’ icon next to the MinKNOW version: you should see either ubuntu-bionic
or ubuntu-xenial
.
If your Mk1C is currently on Ubuntu Xenial, download the following image (with MinKNOW 21.11 included):
https://nanoporetech.box.com/s/bsuzar12lvjp15qawz421t9zghk59jw0
Please visit our Help website for information on how to update MinKNOW on the MinION Mk1C after upgrading to Bionic (https://help.nanoporetech.com).
The recovery SD card for devices on Bionic is available here (latest image has MinKNOW 22.12.5 preinstalled):
https://nanoporetech.box.com/s/yftsu9fr9ukqilairdjx8xadtr1u5ygl
If your MinION Mk1C is already on Bionic, you will need to update to the more stable version if you are unable to update to MinKNOW 22.03.
Power down the Mk1C. Insert the microSD card and power on your Mk1C. After 30 seconds you should see a System Recovery pop-up screen:
Select Factory Reset to re-image the device with the software on the microSD card.
Follow the on-screen instructions to complete the OS update.
If no screen is observed, it is likely you have downloaded the wrong microSD card OS image, so check your OS version again. Alternatively, your device needs to perform some pre-requisite actions, see the Troubleshooting section ("If your Mk1C does not boot from microSD card") for more information. If you have issues, contact support@nanoporetech.com.
Please visit our Oxford Nanopore FAQ and Help Centre if you are having difficulties updating MinKNOW on the MinION Mk1C.
All sequencing data stored in the /data folder is retained after the OS upgrade. However, please note that you may need to configure some device settings after the update.
If you use a mounted external storage to the Mk1C, you will need to remount the external storage. Follow the instructions in the Mount network drive section in this user manual.
If you cannot connect to the internet (you may see the error "Unknown connection 'static'"), follow the instructions in the (Advanced) Network configuration - static vs dynamic IP section in this user manual.
Updating the signing key to update MinKNOW (IMPORTANT)
For details on how to connect via Secure Shell for various operating systems, see the (Advanced) Connecting by the command-line section of this user manual.
wget -O- https://cdn.oxfordnanoportal.com/apt/ont-repo.pub.2022 | sudo apt-key add -
This command can also be used for MinKNOW for MinION Mk1B Linux installations.
If you are unable to update MinKNOW or check for updates, repeat the steps above again. If the issue is not resolved, contact Customer Support at support@nanoporetech.com.
Export logs
The logs will be downloaded as a TGZ file in the logs directory.
If your Mk1C does not boot from microSD card
https://nanoporetech.box.com/s/2vsqb3d6n5mbbtyrbpu1kyai2p3ohknj
Use the following command at your PC/Mac command line/terminal, respectively, to copy the attached file to your Mk1C:
scp ./u-boot.img minit@<your-device>:/home/minit
Note: if you encounter issues, make sure that your PC/Mac is connected via Ethernet to the same network as your MinION Mk1C.
ssh minit@<your-device>
sudo dd if=~/u-boot.img of=/dev/disk/by-partlabel/kernel
then run:
sudo dd if=~/u-boot.img of=/dev/disk/by-partlabel/kernel_b
sudo reboot now
Leave the device to boot to completion, and do not turn it off mid-reboot.
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.