Version for device:
Other available device versions: MinION
Device information
The Oxford Nanopore Technologies® MinION™ is an electronic device that provides the interface between the user’s computer and the nanopore sensor array. The MinION powers to the application-specific integrated circuit (ASIC), performs temperature control and transfers data to the PC through a single USB 3.0 port. The MinION Mk1B can be used with MinION flow cells and Flongle adapter and flow cells. The MinION device can be used for DNA and RNA sequencing. The MinION is operated using the MinKNOW™ software, which controls the device, experimental scripts and also performs basecalling.
Component | Specification |
---|---|
Size and weight | H 23 mm x W 33 mm x D 105 mm, 100 g |
Power | 5 W |
Ports | 1 x USB Type-B |
Environmental conditions | Designed to sequence at 18°C to 25°C |
MinION Mk1B IT requirements
MinION Mk1B technical specification
Safety and regulatory information
Our MinION Mk1B FAQs are located here.
What's in the box
The MinION Mk1B is shipped with:
Flongle adapters are shipped with a Flongle CTC.
LED lights
There are four LED lights on each side of the device. The LEDs on the left side of the device display the following patterns:
The LEDs on the right side of the device display the following patterns:
Lid
The lid can be flipped open to access the flow cell housing.
Clip
The clip holds the flow cell or Configuration Test Cell securely in place.
Configuration Test Cell (CTC)
The CTC is used during the hardware check to ensure that the communication between the device and the flow cell is working correctly.
USB 3.0 port
The USB 3.0 port is for connecting the MinION to the host computer for powering the device and data transfer.
Installing sequencing software
All Oxford Nanopore devices use MinKNOW™ as the primary software. The MinKNOW software carries out several core tasks: data acquisition, real-time analysis and feedback, basecalling, data streaming, controlling the device, and ensuring that the platform chemistry is performing correctly to run the samples. MinKNOW takes the raw data and converts it into reads by recognition of the distinctive change in current that occurs when a DNA strand enters and leaves the pore. MinKNOW then basecalls the reads, and writes out the data into FASTQ or BAM files.
Please follow the MinKNOW protocol for information on how to install the software on Windows, Mac OS X or Linux.
Setting up the MinION Mk1B
Lights will be visible inside the MinION Mk1B device and the fan may be audible.
Setting up the MinION Mk1B for Flongle
The adapter should sit evenly and flat on the MinION Mk1B. This ensures the flow cell assembly is flat during the next stage.
Lights will be visible inside the MinION Mk1B device and the fan may be audible.
A hardware check must be performed on all new devices or when software has been updated. This uses the Configuration Test Cell(s) (CTC), that come pre-inserted into your device in place of flow cells.
Note: If using a Flongle Flow Cell, we recommend regularly checking the Flongle adapter by inserting the empty adapter and completing a hardware check. For checking the device or flow cell position, use a MinION CTC, even if a Flongle Flow Cell will be used for sequencing.
Ensure the position selected is ready to start, as indicated on the GUI.
Using MinION flow cells:
Using Flongle flow cells:
Note: To check the Flongle adapter, insert the EMPTY adapter. To check device or position, insert only a MinION CTC.
A loading bar appears under the flow cell during the checks.
This example is a Flongle CTC (loading bar highlighted in red).
Hardware check pass is indicated by a green check icon.
An orange check icon is a fail.
If the hardware check fails, remove and reinsert the CTC, and run a hardware check again. If the check fails for a second time, please contact Technical Support via email (support@nanoporetech.com).
Flow cell check
We strongly recommend performing a flow cell check before loading a DNA or RNA library to assess the number of pores available.
Oxford Nanopore Technologies will replace any flow cell that falls below the warranty number of active pores within three months of purchase, provided that you report the results within two days of performing the flow cell check and you have followed the storage recommendations.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell (FLO-FLG001) | 50 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
Note: For Flongle, ensure to fill in the flow cell ID.
The ID is case sensitive with no spaces.
The MinKNOW software will not allow you to proceed until the flow cell ID has been entered correctly.
The user will be notified in the bottom right corner of the GUI if any information entered is in the wrong format or missing.
The Flongle flow cell ID is shown in the blue blox below:
A loading bar will be displayed under the flow cell during the check.
The quality of the flow cell will be shown as one of the three outcomes:
Yellow exclamation mark (Flongle flow cell): The number of sequencing pores is below warranty. Take the flow cell out of the device, re-insert it and run a flow cell check again. If the flow cell is still below warranty, contact support@nanoporetech.com
Green tick (MinION flow cell): The number of sequencing pores is above warranty.
Question mark (PromethION flow cell): A flow cell check has not been run on the flow cell during this MinKNOW session.
Note: The indicator of quality will only remain visible during a MinKNOW session. Once the MinKNOW session has ended, the status of the flow cell will be erased.
Sequencing and monitoring the experiment
To run a sequencing experiment, you will need your DNA or RNA sample purified and a sequencing library prepared. For more information about sample preparation, refer to the Prepare section of our Documentation.
You can access the MinKNOW protocol by following this link.
Availability of updates to the MinKNOW software are indicated via the MinKNOW host settings or as a pop-up when first opening the software. The user should follow the on-screen instructions to install the new versions.
The details of the update will be communicated in Nanopore Community announcements.
We strongly recommend users to update as soon as reasonably possible after the release has been made available.
Note: Users will not be able to update their device if connected remotely or if a run is in progress.
If you have created custom scripts in MinKNOW, care should be taken to store them securely so they are not overwritten during software updates.
Please also be aware that the script structure can change between MinKNOW versions; custom scripts written for one version may not be compatible with the next.
Select Get Update to update the device software automatically.
Updates may be skipped. However, we recommend to update the device as soon as updates are available. Some updates will be mandatory to use the device and are unable to be skipped.
Note: For MinION Mk1B, clicking Get Update will open the Software Downloads page on the connected computer for you to download the updated MinION software.
An update button will only appear when a new version of the software is available to download.
Note: For MinION Mk1B devices, clicking Get Update will open the Software Downloads page on the connected computer for you to select and download the updated MinION software.
In some cases (e.g. if there is a problem with the version of MinKNOW that you are using), it may be necessary to re-install the software. To do this, first do a full uninstall of the current version by following the instructions in the Support article: How to uninstall MinION Software for Mk1B and P2 Solo
sudo apt clean
sudo apt update
sudo apt install ont-standalone-minknow-release
The latest version of MinKNOW will be installed. Once these commands have completed successfully without errors, reboot the device:
sudo reboot
In some cases (e.g. if there is a problem with the version of MinKNOW that you are using), it may be necessary to re-install the software. To do this, first do a full uninstall of the current version by following the instructions in the Support article: How to uninstall MinION Software for Mk1B and P2 Solo
Export logs
Log files for each sequencing experiment can be found in:
Windows
C:\data\logs
Mac OS X
/private/var/log/MinKNOW
Linux
/var/log/minknow
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.