Version for device:
Other available device versions: MinION Flongle
A Flow Cell Check is used to assess the number of nanopores that are available in the flow cell. It should be run when you are ready to use the flow cell, and within 3 months of purchase (except for Flongle flow cells, which currently have a shelf life of four weeks).
Oxford Nanopore Technologies will replace any flow cell that falls below the warranty number of active pores, provided the result is reported within two days of performing the Flow Cell Check and the storage recommendations have been followed.
Flow cell | Minimum number of active pores covered by warranty |
---|---|
Flongle Flow Cell (FLO-FLG001) | 50 |
Flongle Flow Cell (FLO-FLGOP1) | 30 |
MinION/GridION Flow Cell | 800 |
PromethION Flow Cell | 5000 |
We recommend to store flow cells unopened, as follows:
Flow cell | Storage instructions |
---|---|
Flongle Flow Cells | 2–8°C for four weeks |
MinION/GridION Flow Cells | room temperature for one month, or at 2–8°C for 12 weeks |
PromethION Flow Cells | 2–8°C for 12 weeks |
Flongle flow cell check
Flow cells are shipped with a QC DNA molecule present in the buffer. This molecule produces a distinctive nanopore signal. The MinKNOW™ software uses this signal to validate the integrity of the nanopore array before use and provides the user with an estimate of the number of simultaneously available channels for the experiment.
The adapter should sit evenly and flat on the MinION Mk1B or GridION platform. This ensures the flow cell assembly is flat during the next stage.
The flow cell should sit evenly and flat inside the adapter, to avoid any bubbles forming inside the fluidic compartments.
To log in, you must be connected to the internet.
From here, you can also change the language in the user interface in the bottom right corner.
You will be prompted to enter your email address and Nanopore password:
If you experience login issues, please visit the Community Support channel (https://community.nanoporetech.com/support) and use Continue as guest for temporary use.
Below is an example of a GridION device with a Flongle inserted into position X1 and MinION Flow Cells in positions X2 - X5.
Below is an example of a GridION device with a Flongle inserted into position X1.
The ID is case sensitive with no spaces.
The MinKNOW software will not allow you to proceed until the flow cell ID has been entered correctly.
The user will be notified in the bottom right corner of the GUI if any information entered is in the wrong format or missing.
The Flongle flow cell ID is shown in the blue box below:
A loading bar appears under the flow cell during the checks.
This example is a Flongle CTC (loading bar highlighted in red).
The quality of the flow cell will be shown as one of the three outcomes:
Yellow exclamation mark (Flongle flow cell) The number of sequencing pores is below warranty.
Green tick (MinION flow cell) The number of sequencing pores is above warranty.
Question mark (PromethION flow cell) A Flow Cell Check has not been run on the flow cell during this MinKNOW session.
Note: The indicator of quality (question mark, exclamation mark or tick) will only remain visible during a MinKNOW session. Once the MinKNOW service has ended, the status of the flow cell will be erased.
Flow cells are shipped with a QC DNA molecule present in the buffer. This molecule produces a distinctive nanopore signal. The MinKNOW™ software uses this signal to validate the integrity of the nanopore array before use and provides the user with an estimate of the number of simultaneously available channels for the experiment.
Power on the MinION Mk1C, as described in the MinION Mk1C protocol under Power.
Press down firmly to ensure good thermal and electrical contact.
Note: The above image uses a CTC for illustrative purposes. Please insert a flow cell.
To log in, you must be connected to the internet.
From here, you can also change the language in the user interface in the bottom right corner.
You will be prompted to enter your email address and Nanopore password:
If you experience login issues, please visit the Community Support channel (https://community.nanoporetech.com/support) and use Continue as guest for temporary use.
Run state can be viewed here and a progress bar is displayed at the base of the flow cell.
Number of pores available will be displayed here once check is completed.
The quality of the flow cell will be shown as one of the three outcomes on the Sequencing Overview page:
Yellow exclamation mark: The number of sequencing pores is below warranty.
Green tick: The number of sequencing pores is above warranty and ready for sequencing.
Question mark: A flow cell check has not been run on the flow cell during this MinKNOW session.
Note: the indicator of quality will only remain visible during the MinKNOW session when testing occurred. Once the sequencing device has been re-started, the status of the flow cell will be erased.
Flow cells are shipped with a QC DNA molecule present in the buffer. This molecule produces a distinctive nanopore signal. The MinKNOW™ software uses this signal to validate the integrity of the nanopore array before use and provides the user with an estimate of the number of simultaneously available channels for the experiment.
Insert the flow cell in the MinION Mk1B/GridION by sliding the flow cell under the clip. Firmly press down on the flow cell to ensure good thermal and electrical contact.
Figure SSFC: Assembled MinION Mk1B and MinION SpotON Flow Cell connected to through the USB 3.0 port to the host computer
To log in, you must be connected to the internet.
From here, you can also change the language in the user interface in the bottom right corner.
You will be prompted to enter your email address and Nanopore password:
If you experience login issues, please visit the Community Support channel (https://community.nanoporetech.com/support) and use Continue as guest for temporary use.
Below is an example of a GridION device with a Flongle inserted into position X1 and MinION Flow Cells in positions X2 - X5.
A loading bar appears under the flow cell during the checks.
This example is a Flongle CTC (loading bar highlighted in red).
The quality of the flow cell will be shown as one of the three outcomes:
Yellow exclamation mark (Flongle flow cell): The number of sequencing pores is below warranty. Take the flow cell out of the device, re-insert it and run a flow cell check again. If the flow cell is still below warranty, contact support@nanoporetech.com
Green tick (MinION flow cell): The number of sequencing pores is above warranty.
Question mark (PromethION flow cell): A flow cell check has not been run on the flow cell during this MinKNOW session.
Note: The indicator of quality will only remain visible during a MinKNOW session. Once the MinKNOW session has ended, the status of the flow cell will be erased.
PromethION flow cell check
Flow cells are shipped with a QC DNA molecule present in the buffer. This molecule produces a distinctive nanopore signal. The PromethION software uses this signal to validate the integrity of the nanopore array before use and provides the user with an estimate of the number of simultaneously available channels for the experiment. Active pores are reported in four groups, each of which may be used in turn when running long experiments e.g. a 64 hour sequencing run. Shorter experiments may not use all groups.
Insert the chip connector fully into the docking port. Take note to ensure the latch to hold the chip in place clicks in.
Line up the flow cell with the connector (horizontally and vertically) and smoothly insert it into position.
Press the flow cell down firmly into the loader, and ensure that the latch engages.
Insertion of the flow cell at the wrong angle can damage the pins on the PromethION and affect your sequencing results. If you find the pins on a PromethION position damaged, please contact support@nanoporetech.com for assistance.
To log in, you must be connected to the internet.
From here, you can also change the language in the user interface in the bottom right corner.
You will be prompted to enter your email address and Nanopore password:
If you experience login issues, please visit the Community Support channel (https://community.nanoporetech.com/support) and use Continue as guest for temporary use.
A loading bar appears under the flow cell during the checks.
This example is a Flongle CTC (loading bar highlighted in red).
The quality of the flow cell will be shown as one of the three outcomes:
Yellow exclamation mark (Flongle flow cell): The number of sequencing pores is below warranty. Take the flow cell out of the device, re-insert it and run a flow cell check again. If the flow cell is still below warranty, contact support@nanoporetech.com
Green tick (MinION flow cell): The number of sequencing pores is above warranty.
Question mark (PromethION flow cell): A flow cell check has not been run on the flow cell during this MinKNOW session.
Note: The indicator of quality will only remain visible during a MinKNOW session. Once the MinKNOW session has ended, the status of the flow cell will be erased.
Purchase a MinION Starter Pack from Avantor to get full community access and benefit from:
Already have a Nanopore Community account?
Log in hereRequest a call with our experts for detailed advice on implementing nanopore sequencing.
Request a callVisit our microbial sequencing spotlight page on vwr.com.
Microbial sequencingWe use cookies and similar technologies on our websites to help provide you with the best possible online experience.
By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.