FFPE-repair the DNA, and prepare the DNA ends for adapter ligation

Need help?

Transfer 20 µl of Lambda DNA (LMD) to a clean 0.2 ml PCR tube (Tube 1).

ligation-v2 0000 Layer-15

Add 27 µl nuclease-free water to the tube.

ligation-v2 0001 Layer-14

Add 1 µl DNA CS (DCS) to the tube.

ligation-v2 0002 Layer-13

Add 3.5 µl of the NEBNext FFPE DNA Repair Buffer to the tube.

ligation-v2 0003 Layer-12

Add 2 µl of the NEBNext FFPE DNA Repair Mix to the tube.

ligation-v2 0004 Layer-11

Add 3.5 µl of the Ultra II End-prep reaction buffer to the tube.

ligation-v2 0005 Layer-10

Add 3 µl of the Ultra II End-prep enzyme mix to the tube.

ligation-v2 0006 Layer-9

Mix the contents of the tube by flicking the tube with your finger. Spin down briefly in a microfuge.

  • Incubate in heat blocks or a thermal cycler at 20°C for 5 minutes, then at 65°C for 5 minutes
  • Take the tube out from the thermal cycler or heat block, and spin down briefly in a microfuge

4-3-5

Transfer the Lambda DNA sample (Tube 1) to a clean 1.5 ml Eppendorf DNA LoBind tube (Tube 2).

4-3-6

Add 60 µl of resuspended AXP to Tube 2 with the Lambda DNA sample, and mix by pipetting up and down.

axp-60

Put the tube in a Hula mixer, and leave to incubate for 5 minutes.

While the DNA sample is incubating, transfer 350 µl of pure ethanol to a clean 1.5 ml DNA LoBind tube.

4-3-8

Add 150 µl nuclease-free water to the ethanol.

4-3-9

Mix the contents of the tube by flicking the tube with your finger. Spin down briefly in a microfuge.

Take Tube 2 with the Lambda DNA sample off the Hula mixer, and spin down in a microfuge.

Place Tube 2 in a magnetic rack, and wait for the beads to collect in a pellet near the magnet, and the solution to become clear.

4-3-9-magnetic-rack

Keep the tube on the magnet to pipette off and discard the supernatant. Take care to not disturb the pellet.

4-3-9-discard-supernatant

Add 200 µl of the freshly-prepared ethanol to Tube 2, without taking the tube off the magnetic rack or disturbing the pellet.

Ensure the pellet is covered with ethanol, then pipette off and discard the ethanol.

4-3-9-add-ethanol

Repeat the previous step:
Add 200 µl of the freshly-prepared ethanol to the tube, without taking the tube off the magnetic rack or disturbing the pellet.

Ensure the pellet is covered with ethanol, then pipette off and discard the ethanol.

Take the tube off the magnetic rack and spin down briefly in a microfuge.

Place Tube 2 back in the magnetic rack, and wait for the beads to collect in a pellet near the magnet. There may be a small amount of residual liquid at the bottom of the tube. Pipette off and discard this liquid.

Remove the tube from the magnetic rack, add 60 µl nuclease-free water to the tube, and resuspend the beads by flicking the tube. Leave the tube to incubate for 2 minutes at room temperature.

ligation-v2 0007 Layer-8

Place the tube back in the magnetic rack, and wait for the beads to collect in a pellet near the magnet.

Pipette off the 60 µl of eluate and transfer it to a clean 1.5 ml Eppendorf DNA LoBind tube (Tube 3). Discard Tube 2 with the pellet.

ligation-v2 0008 Layer-7

FFPE repair and end-prep complete

You now have 60 µl of purified Lambda DNA (Tube 3), and are ready to ligate sequencing adapters to the DNA ends.

Prev 4.3 Ligate sequencing adapters to the DNA fragment ends

help

FAQ

Searching...



Can't find what you're looking for?

Ask the Community Contact Support

Cookies Notice

We use cookies and similar technologies on our websites to help provide you with the best possible online experience.

By using our sites and apps, you agree that we may store and access cookies and similar technologies on your device.