Overview

A protocol for full-length polyadenylated RNA

  • This protocol is for full-length polyadenylated tail RNA
  • Starting material can be total RNA or an enriched sample (e.g. poly(A)+ or ribodepleted)
  • Native RNA is sequenced
  • No fragmentation
  • 2 hr library prep

For Research Use Only

This is a Legacy product
This kit is soon to be discontinued and we recommend all customers to upgrade to the latest chemistry for their relevant kit which is available on the Store. If customers require further support for any ongoing critical experiments using a Legacy product, please contact Customer Support via email: support@nanoporetech.com.

Introduction to the protocol
Library preparation
Sequencing and data analysis
Troubleshooting
Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

Overview of the protocol

  1. Important This is a Legacy product

    This kit is soon to be discontinued and we recommend all customers to upgrade to the latest chemistry for their relevant kit which is available on the Store. If customers require further support for any ongoing critical experiments using a Legacy product, please contact Customer Support via email: support@nanoporetech.com. For further information on please see the product update page.

  2. Direct RNA Sequencing Kit features

    This kit is highly recommended for users who:

    • are exploring attributes of native RNA such as modified bases.
    • would like to remove RT or PCR bias.
    • have transcripts that are difficult to reverse transcribe.
  3. Introduction to the Direct RNA Sequencing protocol

    This protocol describes how to carry out sequencing of native RNA using the Direct RNA Sequencing Kit (SQK-RNA002).

    Steps in the sequencing workflow:

    Prepare for your experiment

    You will need to:

    • Extract your RNA, and check its length, quantity and purity. The quality checks performed during the protocol are essential in ensuring experimental success.
    • Ensure you have your sequencing kit, the correct equipment and third-party reagents
    • Download the software for acquiring and analysing your data
    • Check your flow cell(s) to ensure it has enough pores for a good sequencing run

    Library preparation

    You will need to:

    • Synthesise the complementary strand of the RNA
    • Attach sequencing adapters supplied in the kit to the ends of the RNA-cDNA hybrid
    • Prime the flow cell, and load your RNA library into the flow cell

    Direct RNA002 workflow

    Sequencing and analysis

    You will need to:

    • Start a sequencing run using the MinKNOW software, which will collect raw data from the device and convert it into basecalled reads
  4. Important Unlike DNA, RNA is translocated through the nanopore in the 3'-5' direction. However, the basecalling algorithms automatically flip the data, and the reads are displayed 5'-3'.
  5. Important Compatibility of this protocol

    This protocol should only be used in combination with:

    • Direct RNA Sequencing Kit (SQK-RNA002)
    • R9.4.1 (FLO-MIN106) flow cells
    • Flow Cell Wash Kit (EXP-WSH004)
Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

Overview of the protocol

  1. Important This is a Legacy product

    This kit is soon to be discontinued and we recommend all customers to upgrade to the latest chemistry for their relevant kit which is available on the Store. If customers require further support for any ongoing critical experiments using a Legacy product, please contact Customer Support via email: support@nanoporetech.com. For further information on please see the product update page.

  2. Direct RNA Sequencing Kit features

    This kit is highly recommended for users who:

    • are exploring attributes of native RNA such as modified bases.
    • would like to remove RT or PCR bias.
    • have transcripts that are difficult to reverse transcribe.
  3. Introduction to the Direct RNA Sequencing protocol

    This protocol describes how to carry out sequencing of native RNA using the Direct RNA Sequencing Kit (SQK-RNA002).

    Steps in the sequencing workflow:

    Prepare for your experiment

    You will need to:

    • Extract your RNA, and check its length, quantity and purity. The quality checks performed during the protocol are essential in ensuring experimental success.
    • Ensure you have your sequencing kit, the correct equipment and third-party reagents
    • Download the software for acquiring and analysing your data
    • Check your flow cell(s) to ensure it has enough pores for a good sequencing run

    Library preparation

    You will need to:

    • Synthesise the complementary strand of the RNA
    • Attach sequencing adapters supplied in the kit to the ends of the RNA-cDNA hybrid
    • Prime the flow cell, and load your RNA library into the flow cell

    Direct RNA002 workflow

    Sequencing and analysis

    You will need to:

    • Start a sequencing run using the MinKNOW software, which will collect raw data from the device and convert it into basecalled reads
  4. Important Unlike DNA, RNA is translocated through the nanopore in the 3'-5' direction. However, the basecalling algorithms automatically flip the data, and the reads are displayed 5'-3'.
  5. Important Compatibility of this protocol

    This protocol should only be used in combination with:

    • Direct RNA Sequencing Kit (SQK-RNA002)
    • R9.4.1 (FLO-PRO002) flow cells
    • Flow Cell Wash Kit (EXP-WSH004)
Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

Equipment and consumables

Materials
  • 50 ng of poly(A)-tailed RNA or 500 ng of total RNA in 9 µl
  • Direct RNA Sequencing Kit (SQK-RNA002)
  • Flow Cell Priming Kit (EXP-FLP002)
  • Consumables
  • 1.5 ml Eppendorf DNA LoBind tubes
  • 0.2 ml thin-walled PCR tubes
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • Freshly prepared 70% ethanol in nuclease-free water
  • SuperScript™ III Reverse Transcriptase (Thermo Fisher Scientific, cat # 18080044)
  • 10 mM dNTP solution (e.g. NEB, cat # N0447)
  • NEBNext® Quick Ligation Reaction Buffer (NEB, B6058)
  • T4 DNA Ligase 2M U/ml (NEB, cat # M0202M)
  • Agencourt RNAClean XP beads (Beckman Coulter™, cat # A63987)
  • Qubit RNA HS Assay Kit (ThermoFisher, cat # Q32852)
  • Qubit dsDNA HS Assay Kit (ThermoFisher, Q32851)
  • Equipment
  • Hula mixer (gentle rotator mixer)
  • Magnetic rack, suitable for 1.5 ml Eppendorf tubes
  • Microfuge
  • Vortex mixer
  • Ice bucket with ice
  • Timer
  • Thermal cycler
  • Qubit fluorometer (or equivalent for QC check)
  • P1000 pipette and tips
  • P200 pipette and tips
  • P100 pipette and tips
  • P20 pipette and tips
  • P10 pipette and tips
  • P2 pipette and tips
  • Optional equipment
  • Agilent Bioanalyzer (or equivalent)
  • Eppendorf 5424 centrifuge (or equivalent)
    1. For this protocol, you will need 50 ng of poly(A)-tailed RNA or 500 ng of total RNA in 9 µl.
    2. Direct RNA Sequencing Kit contents

      dRNA tubes (1)

      Name Acronym Cap colour No. of vials Fill volume per vial (μl)
      RT Adapter RTA Blue 1 10
      RNA Adapter Mix RMX Green 1 45
      RNA Calibrant Strand RCS Yellow 1 25
      Wash Buffer WSB Orange 2 1,200
      Elution Buffer ELB Black 1 300
      RNA Running Buffer RRB Red 1 600

      Note: The RNA CS (RCS) is the calibration strand and contains the Enolase II from YHR174W, extracted from the yeast saccharomyces cerevisiae. The reference FASTA files for the yeast is available here.

    3. Flow Cell Priming Kit contents (EXP-FLP002)

      FLP

      Name Acronym Cap colour No. of vials Fill volume per vial (μl)
      Flush Buffer FB Blue 6 1,170
      Flush Tether FLT Purple 1 200
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Equipment and consumables

    Materials
  • 50 ng of poly(A)-tailed RNA or 500 ng of total RNA in 9 µl
  • Direct RNA Sequencing Kit (SQK-RNA002)
  • Flow Cell Priming Kit (EXP-FLP002)
  • Consumables
  • 1.5 ml Eppendorf DNA LoBind tubes
  • 0.2 ml thin-walled PCR tubes
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • Freshly prepared 70% ethanol in nuclease-free water
  • SuperScript™ III Reverse Transcriptase (Thermo Fisher Scientific, cat # 18080044)
  • 10 mM dNTP solution (e.g. NEB N0447)
  • NEBNext® Quick Ligation Reaction Buffer (NEB, B6058)
  • T4 DNA Ligase 2M U/ml (NEB, cat # M0202M)
  • Agencourt RNAClean XP beads (Beckman Coulter™, cat # A63987)
  • Qubit RNA HS Assay Kit (ThermoFisher, cat # Q32852)
  • Qubit dsDNA HS Assay Kit (ThermoFisher, Q32851)
  • Equipment
  • Hula mixer (gentle rotator mixer)
  • Magnetic rack, suitable for 1.5 ml Eppendorf tubes
  • Microfuge
  • Vortex mixer
  • Thermal cycler
  • P1000 pipette and tips
  • P200 pipette and tips
  • P100 pipette and tips
  • P20 pipette and tips
  • P10 pipette and tips
  • P2 pipette and tips
  • Ice bucket with ice
  • Timer
  • Qubit fluorometer (or equivalent for QC check)
  • Optional equipment
  • Agilent Bioanalyzer (or equivalent)
  • Eppendorf 5424 centrifuge (or equivalent)
    1. For this protocol, you will need 50 ng of poly(A)-tailed RNA or 500 ng of total RNA in 9 µl.
    2. Direct RNA Sequencing Kit contents

      dRNA tubes (1)

      Name Acronym Cap colour No. of vials Fill volume per vial (μl)
      RT Adapter RTA Blue 1 10
      RNA Adapter Mix RMX Green 1 45
      RNA Calibrant Strand RCS Yellow 1 25
      Wash Buffer WSB Orange 2 1,200
      Elution Buffer ELB Black 1 300
      RNA Running Buffer RRB Red 1 600

      Note: The RNA CS (RCS) is the calibration strand and contains the Enolase II from YHR174W, extracted from the yeast saccharomyces cerevisiae. The reference FASTA files for the yeast is available here.

    3. Flow Cell Priming Kit (EXP-FLP002) contents

      FLP

      Name Acronym Cap colour No. of vial Fill volume per vial (μl)
      Flush Buffer FB Blue 6 1,170
      Flush Tether FLT Purple 1 200
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Computer requirements and software

    Computer requirements and software

    1. MinION Mk1B IT requirements

      Sequencing on a MinION Mk1B requires a high-spec computer or laptop to keep up with the rate of data acquisition. For more information, refer to the MinION Mk1B IT requirements document.

    2. MinION Mk1C IT requirements

      The MinION Mk1C contains fully-integrated compute and screen, removing the need for any accessories to generate and analyse nanopore data. For more information refer to the MinION Mk1C IT requirements document.

    3. MinION Mk1D IT requirements

      Sequencing on a MinION Mk1D requires a high-spec computer or laptop to keep up with the rate of data acquisition. For more information, refer to the MinION Mk1D IT requirements document.

    4. Software for nanopore sequencing

      MinKNOW

      The MinKNOW software controls the nanopore sequencing device, collects sequencing data and basecalls in real time. You will be using MinKNOW for every sequencing experiment to sequence, basecall and demultiplex if your samples were barcoded.

      For instructions on how to run the MinKNOW software, please refer to the MinKNOW protocol.

      EPI2ME (optional)

      The EPI2ME cloud-based platform performs further analysis of basecalled data, for example alignment to the Lambda genome, barcoding, or taxonomic classification. You will use the EPI2ME platform only if you would like further analysis of your data post-basecalling.

      For instructions on how to create an EPI2ME account and install the EPI2ME Desktop Agent, please refer to this link.

    5. Check your flow cell

      We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.

      Flow cell Minimum number of active pores covered by warranty
      Flongle Flow Cell 50
      MinION/GridION Flow Cell 800
      PromethION Flow Cell 5000
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    1. GridION IT requirements

      The GridION device contains all the hardware required to control up to five flow cells and acquire the data. The device is further enhanced with high performance GPU technology for real-time basecalling. Read more in the GridION IT requirements document.

    2. Software for nanopore sequencing

      MinKNOW

      The MinKNOW software controls the nanopore sequencing device, collects sequencing data and basecalls in real time. You will be using MinKNOW for every sequencing experiment to sequence, basecall and demultiplex if your samples were barcoded.

      For instructions on how to run the MinKNOW software, please refer to the MinKNOW protocol.

      EPI2ME (optional)

      The EPI2ME cloud-based platform performs further analysis of basecalled data, for example alignment to the Lambda genome, barcoding, or taxonomic classification. You will use the EPI2ME platform only if you would like further analysis of your data post-basecalling.

      For instructions on how to create an EPI2ME account and install the EPI2ME Desktop Agent, please refer to this link.

    3. Check your flow cell

      We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.

      Flow cell Minimum number of active pores covered by warranty
      Flongle Flow Cell 50
      MinION/GridION Flow Cell 800
      PromethION Flow Cell 5000
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    1. PromethION 24/48 IT requirements

      The PromethION device contains all the hardware required to control up to 24 (for the P24 model) or 48 (for the P48 model) sequencing experiments and acquire the data. The device is further enhanced with high performance GPU technology for real-time basecalling. Read more in the PromethION IT requirements document.

    2. PromethION 2 Solo IT requirements

      The PromethION 2 (P2) Solo is a device which directly connects into a GridION Mk1 or a stand-alone computer that meets the miminum specifications for real-time data streaming and analysis. Up to two PromethION flow cells can be can be run and each is independently addressable, meaning experiments can be run concurrently or individually. For information on the computer IT requirements, please see the PromethION 2 Solo IT requirements document.

    3. Software for nanopore sequencing

      MinKNOW

      The MinKNOW software controls the nanopore sequencing device, collects sequencing data and basecalls in real time. You will be using MinKNOW for every sequencing experiment to sequence, basecall and demultiplex if your samples were barcoded.

      For instructions on how to run the MinKNOW software, please refer to the MinKNOW protocol.

      EPI2ME (optional)

      The EPI2ME cloud-based platform performs further analysis of basecalled data, for example alignment to the Lambda genome, barcoding, or taxonomic classification. You will use the EPI2ME platform only if you would like further analysis of your data post-basecalling.

      For instructions on how to create an EPI2ME account and install the EPI2ME Desktop Agent, please refer to this link.

    4. Check your flow cell

      We highly recommend that you check the number of pores in your flow cell prior to starting a sequencing experiment. This should be done within 12 weeks of purchasing for MinION/GridION/PromethION or within four weeks of purchasing Flongle Flow Cells. Oxford Nanopore Technologies will replace any flow cell with fewer than the number of pores in the table below, when the result is reported within two days of performing the flow cell check, and when the storage recommendations have been followed. To do the flow cell check, please follow the instructions in the Flow Cell Check document.

      Flow cell Minimum number of active pores covered by warranty
      Flongle Flow Cell 50
      MinION/GridION Flow Cell 800
      PromethION Flow Cell 5000
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Library preparation

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Materials
  • 50 ng of poly(A)-tailed RNA or 500 ng of total RNA in 9 µl
  • RT Adapter (RTA)
  • RNA CS (RCS)
  • RNA Adapter (RMX)
  • Wash Buffer (WSB)
  • Elution Buffer (ELB)
  • Consumables
  • NEBNext® Quick Ligation Reaction Buffer (NEB, B6058)
  • T4 DNA Ligase 2M U/ml (NEB, cat # M0202M)
  • 0.2 ml thin-walled PCR tubes
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • Agencourt RNAClean XP beads (Beckman Coulter™, cat # A63987)
  • Freshly prepared 70% ethanol in nuclease-free water
  • 1.5 ml Eppendorf DNA LoBind tubes
  • SuperScript™ III Reverse Transcriptase (Thermo Fisher Scientific, cat # 18080044)
  • 10 mM dNTP solution (e.g. NEB, cat # N0447)
  • Qubit dsDNA HS Assay Kit (ThermoFisher, Q32851)
  • Equipment
  • Magnetic rack, suitable for 1.5 ml Eppendorf tubes
  • Hula mixer (gentle rotator mixer)
  • Thermal cycler
  • Optional equipment
  • Qubit fluorometer (or equivalent for QC check)
    1. Prepare the RNA in nuclease-free water.

      • Transfer 50 ng of poly(A)-tailed RNA or 500 ng of total RNA into a 1.5 ml Eppendorf DNA LoBind tube
      • Adjust the volume to 9 μl with nuclease-free water
      • Mix thoroughly by flicking the tube to avoid unwanted shearing
      • Spin down briefly in a microfuge

    2. In a 0.2 ml thin-walled PCR tube, mix the reagents in the following order:

      Reagent Volume
      NEBNext Quick Ligation Reaction Buffer 3.0 µl
      RNA 9.0 µl
      RNA CS (RCS), 110 nM 0.5 µl
      RT Adapter (RTA) 1.0 µl
      T4 DNA Ligase 1.5 µl
      Total 15 µl

    3. Mix by pipetting and spin down.
    4. Incubate the reaction for 10 minutes at room temperature.
    5. Mix the following reagents together to make the reverse transcription master mix:

      Reagent Volume
      Nuclease-free water 9.0 µl
      10 mM dNTPs 2.0 µl
      5x first-strand buffer 8.0 µl
      0.1 M DTT 4.0 µl
      Total 23.0 µl

    6. Add the master mix to the 0.2 ml PCR tube containing the RT adapter-ligated RNA from the "RT Adapter ligation" step. Mix by pipetting.
    7. Add 2 µl of SuperScript III Reverse Transcriptase to the reaction and mix by pipetting.
    8. Transfer the sample to a clean 1.5 ml Eppendorf DNA LoBind tube.
    9. Resuspend the stock of Agencourt RNAClean XP beads by vortexing.
    10. Add 72 µl of resuspended Agencourt RNAClean XP beads to the reverse transcription reaction and mix by pipetting.
    11. Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
    12. Prepare 200 μl of fresh 70% ethanol in nuclease-free water.
    13. Spin down the sample and pellet on a magnet. Keep the tube on the magnet, and pipette off the supernatant when clear and colourless.
    14. Keep the tube on magnet, and wash the beads with 150 µl of freshly prepared 70% ethanol without disturbing the pellet as described below.

      1. Keeping the magnetic rack on the benchtop, rotate the bead-containing tube by 180°. Wait for the beads to migrate towards the magnet and form a pellet.
      2. Rotate the tube 180° again (back to the starting position), and wait for the beads to pellet.

    15. Remove the 70% ethanol using a pipette and discard.
    16. Spin down and place the tube back on the magnet until the eluate is clear and colourless. Keep the tube(s) on the magnet and pipette off any residual ethanol.
    17. Remove the tube from the magnetic rack and resuspend the pellet in 20 µl nuclease-free water. Incubate for 5 minutes at room temperature.
    18. Pellet the beads on a magnet until the eluate is clear and colourless.
    19. Remove and retain 20 µl of eluate into a clean 1.5 ml Eppendorf DNA LoBind tube.
    20. In the same 1.5 ml Eppendorf DNA LoBind tube, mix the reagents in the following order:

      Reagent Volume
      NEBNext Quick Ligation Reaction Buffer 8.0 µl
      RNA Adapter (RMX) 6.0 µl
      Nuclease-free water 3.0 µl
      T4 DNA Ligase 3.0 µl
      Total (including all reagents) 40 µl

    21. Mix by pipetting.
    22. Incubate the reaction for 10 minutes at room temperature.
    23. Resuspend the stock of Agencourt RNAClean XP beads by vortexing.
    24. Add 16 µl of resuspended Agencourt RNAClean XP beads to the reaction and mix by pipetting.
    25. Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
    26. Spin down the sample and pellet on a magnet. Keep the tube on the magnet, and pipette off the supernatant when clear and colourless.
    27. Add 150 μl of the Wash Buffer (WSB) to the beads. Close the tube lid and resuspend the beads by flicking the tube. Return the tube to the magnetic rack, allow the beads to pellet for 5 minutes and pipette off the supernatant when clear and colourless.
    28. Important Agitating the beads results in a more efficient removal of free adapter, compared to adding the wash buffer and immediately aspirating.
    29. Remove the tube from the magnetic rack and resuspend pellet in 21 µl Elution Buffer by the gently flicking the tube. Incubate for 10 minutes at room temperature.
    30. Pellet the beads on a magnet until the eluate is clear and colourless.
    31. Remove and retain 21 µl of eluate into a clean 1.5 ml Eppendorf DNA LoBind tube.
    32. Quantify 1 µl of reverse-transcribed and adapted RNA using the Qubit fluorometer DNA HS assay - recovery aim ~20 ng.
    33. End of step The reverse-transcribed and adapted RNA is now ready for loading into the flow cell.
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Library preparation

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Materials
  • 50 ng of poly(A)-tailed RNA or 500 ng of total RNA in 9 µl
  • RT Adapter (RTA)
  • RNA CS (RCS)
  • RNA Adapter (RMX)
  • Wash Buffer (WSB)
  • Elution Buffer (ELB)
  • Consumables
  • NEBNext® Quick Ligation Reaction Buffer (NEB, B6058)
  • T4 DNA Ligase 2M U/ml (NEB, cat # M0202M)
  • 0.2 ml thin-walled PCR tubes
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • Agencourt RNAClean XP beads (Beckman Coulter™, cat # A63987)
  • Freshly prepared 70% ethanol in nuclease-free water
  • 1.5 ml Eppendorf DNA LoBind tubes
  • SuperScript™ III Reverse Transcriptase (Thermo Fisher Scientific, cat # 18080044)
  • 10 mM dNTP solution (e.g. NEB, cat # N0447)
  • Qubit dsDNA HS Assay Kit (ThermoFisher, Q32851)
  • Equipment
  • Magnetic rack, suitable for 1.5 ml Eppendorf tubes
  • Hula mixer (gentle rotator mixer)
  • Thermal cycler
  • Optional equipment
  • Qubit fluorometer (or equivalent for QC check)
    1. Prepare the RNA in nuclease-free water.

      • Transfer 50 ng of poly(A)-tailed RNA or 500 ng of total RNA into a 1.5 ml Eppendorf DNA LoBind tube
      • Adjust the volume to 9 μl with nuclease-free water
      • Mix thoroughly by flicking the tube to avoid unwanted shearing
      • Spin down briefly in a microfuge

    2. In a 0.2 ml thin-walled PCR tube, mix the reagents in the following order:

      Reagent Volume
      NEBNext Quick Ligation Reaction Buffer 3.0 µl
      RNA 9.0 µl
      RNA CS (RCS), 110 nM 0.5 µl
      RT Adapter (RTA) 1.0 µl
      T4 DNA Ligase 1.5 µl
      Total 15 µl

    3. Mix by pipetting and spin down.
    4. Incubate the reaction for 10 minutes at room temperature.
    5. Mix the following reagents together to make the reverse transcription master mix:

      Reagent Volume
      Nuclease-free water 9.0 µl
      10 mM dNTPs 2.0 µl
      5x first-strand buffer 8.0 µl
      0.1 M DTT 4.0 µl
      Total 23.0 µl

    6. Add the master mix to the 0.2 ml PCR tube containing the RT adapter-ligated RNA from the "RT Adapter ligation" step. Mix by pipetting.
    7. Add 2 µl of SuperScript III Reverse Transcriptase to the reaction and mix by pipetting.
    8. Transfer the sample to a clean 1.5 ml Eppendorf DNA LoBind tube.
    9. Resuspend the stock of Agencourt RNAClean XP beads by vortexing.
    10. Add 72 µl of resuspended Agencourt RNAClean XP beads to the reverse transcription reaction and mix by pipetting.
    11. Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
    12. Prepare 200 μl of fresh 70% ethanol in nuclease-free water.
    13. Spin down the sample and pellet on a magnet. Keep the tube on the magnet, and pipette off the supernatant when clear and colourless.
    14. Keep the tube on magnet, and wash the beads with 150 µl of freshly prepared 70% ethanol without disturbing the pellet as described below.

      1. Keeping the magnetic rack on the benchtop, rotate the bead-containing tube by 180°. Wait for the beads to migrate towards the magnet and form a pellet.
      2. Rotate the tube 180° again (back to the starting position), and wait for the beads to pellet.

    15. Remove the 70% ethanol using a pipette and discard.
    16. Spin down and place the tube back on the magnet until the eluate is clear and colourless. Keep the tube(s) on the magnet and pipette off any residual ethanol.
    17. Remove the tube from the magnetic rack and resuspend the pellet in 20 µl nuclease-free water. Incubate for 5 minutes at room temperature.
    18. Pellet the beads on a magnet until the eluate is clear and colourless.
    19. Remove and retain 20 µl of eluate into a clean 1.5 ml Eppendorf DNA LoBind tube.
    20. In the same 1.5 ml Eppendorf DNA LoBind tube, mix the reagents in the following order:

      Reagent Volume
      NEBNext Quick Ligation Reaction Buffer 8.0 µl
      RNA Adapter (RMX) 6.0 µl
      Nuclease-free water 3.0 µl
      T4 DNA Ligase 3.0 µl
      Total (including all reagents) 40 µl

    21. Mix by pipetting.
    22. Incubate the reaction for 10 minutes at room temperature.
    23. Resuspend the stock of Agencourt RNAClean XP beads by vortexing.
    24. Add 16 µl of resuspended Agencourt RNAClean XP beads to the reaction and mix by pipetting.
    25. Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
    26. Spin down the sample and pellet on a magnet. Keep the tube on the magnet, and pipette off the supernatant when clear and colourless.
    27. Add 150 μl of the Wash Buffer (WSB) to the beads. Close the tube lid and resuspend the beads by flicking the tube. Return the tube to the magnetic rack, allow the beads to pellet for 5 minutes and pipette off the supernatant when clear and colourless.
    28. Important Agitating the beads results in a more efficient removal of free adapter, compared to adding the wash buffer and immediately aspirating.
    29. Remove the tube from the magnetic rack and resuspend pellet in 41 µl Elution Buffer. Incubate for 10 minutes at room temperature.
    30. Pellet the beads on a magnet until the eluate is clear and colourless.
    31. Remove and retain 41 µl of eluate into a clean 1.5 ml Eppendorf DNA LoBind tube.
    32. Quantify 1 µl of reverse-transcribed and adapted RNA using the Qubit fluorometer DNA HS assay - recovery aim ~20 ng.
    33. End of step The reverse-transcribed and adapted RNA is now ready for loading into the flow cell.
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Priming and loading the SpotON flow cell

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Materials
  • Prepared RNA library
  • RNA Running Buffer (RRB)
  • Flow Cell Priming Kit (EXP-FLP002)
  • Consumables
  • MinION Mk1B
  • MinION and GridION Flow Cell
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • 1.5 ml Eppendorf DNA LoBind tubes
    1. Thaw the RNA Running Buffer (RRB), Flush Tether (FLT) and one tube of Flush Buffer (FB) at room temperature.
    2. Mix the RNA Running Buffer (RRB), Flush Buffer (FB) and Flush Tether (FLT) tubes thoroughly by vortexing and spin down at room temperature.
    3. To prepare the flow cell priming mix, add 30 µl of thawed and mixed Flush Tether (FLT) directly to the tube of thawed and mixed Flush Buffer (FB), and mix by vortexing at room temperature.
    4. Open the MinION device lid and slide the flow cell under the clip.

      Press down firmly on the flow cell to ensure correct thermal and electrical contact.

      Flow Cell Loading Diagrams Step 1a

      Flow Cell Loading Diagrams Step 1b

    5. Optional action
      Complete a flow cell check to assess the number of pores available before loading the library.

      This step can be omitted if the flow cell has been checked previously.

      See the flow cell check instructions in the MinKNOW protocol for more information.

    6. Slide the priming port cover clockwise to open the priming port.

      Flow Cell Loading Diagrams Step 2

    7. Important Take care when drawing back buffer from the flow cell. Do not remove more than 20-30 µl, and make sure that the array of pores are covered by buffer at all times. Introducing air bubbles into the array can irreversibly damage pores.
    8. After opening the priming port, check for a small air bubble under the cover. Draw back a small volume to remove any bubbles:

      1. Set a P1000 pipette to 200 µl
      2. Insert the tip into the priming port
      3. Turn the wheel until the dial shows 220-230 µl, to draw back 20-30 µl, or until you can see a small volume of buffer entering the pipette tip

      Note: Visually check that there is continuous buffer from the priming port across the sensor array.

      Flow Cell Loading Diagrams Step 03 V5

    9. Load 800 µl of the priming mix into the flow cell via the priming port, avoiding the introduction of air bubbles. Wait for five minutes. During this time, prepare the library for loading by following the steps below.

      Flow Cell Loading Diagrams Step 04 V5

    10. Important Thoroughly mix the contents of the RRB tube by vortexing or pipetting, and spin down briefly.
    11. Take 20 µl of the prepared RNA library and mix it with 17.5 µl of nuclease-free water.
    12. In a new tube, prepare the library for loading as follows:

      Reagent Volume per flow cell
      RNA Running Buffer (RRB) 37.5 µl
      RNA library in nuclease-free water 37.5 µl
      Total 75 µl

    13. Complete the flow cell priming:

      1. Gently lift the SpotON sample port cover to make the SpotON sample port accessible.
      2. Load 200 µl of the priming mix into the flow cell priming port (not the SpotON sample port), avoiding the introduction of air bubbles.

      Flow Cell Loading Diagrams Step 5

      Flow Cell Loading Diagrams Step 06 V5

    14. Mix the prepared library gently by pipetting up and down just prior to loading.
    15. Add 75 μl of sample to the Flow Cell via the SpotON sample port in a dropwise fashion. Ensure each drop flows into the port before adding the next.

      Flow Cell Loading Diagrams Step 07 V5

    16. Gently replace the SpotON sample port cover, making sure the bung enters the SpotON port and close the priming port.

      Step 8 update

      Flow Cell Loading Diagrams Step 9

    17. Important Install the light shield on your flow cell as soon as library has been loaded for optimal sequencing output.

      We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.

    18. Place the light shield onto the flow cell, as follows:

      1. Carefully place the leading edge of the light shield against the clip.
        Note: Do not force the light shield underneath the clip.

      2. Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.

      J2264 - Light shield animation Flow Cell FAW optimised

    19. Caution The MinION Flow Cell Light Shield is not secured to the flow cell and careful handling is required after installation.
    20. End of step Close the device lid and set up a sequencing run on MinKNOW.
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Priming and loading the SpotON flow cell for GridION

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Materials
  • Prepared RNA library
  • RNA Running Buffer (RRB)
  • Flow Cell Priming Kit (EXP-FLP002)
  • Consumables
  • GridION device
  • MinION and GridION Flow Cell
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • 1.5 ml Eppendorf DNA LoBind tubes
    1. Thaw the RNA Running Buffer (RRB), Flush Tether (FLT) and one tube of Flush Buffer (FB) at room temperature.
    2. Mix the RNA Running Buffer (RRB), Flush Buffer (FB) and Flush Tether (FLT) tubes thoroughly by vortexing and spin down at room temperature.
    3. To prepare the flow cell priming mix, add 30 µl of thawed and mixed Flush Tether (FLT) directly to the tube of thawed and mixed Flush Buffer (FB), and mix by vortexing at room temperature.
    4. Slide open the GridION lid and insert the flow cell.

      Press down firmly on the flow cell to ensure correct thermal and electrical contact.
      Flow Cell Loading Diagrams Step 1a

      Flow Cell Loading Diagrams Step 1b

    5. Optional action
      Complete a flow cell check to assess the number of pores available before loading the library.

      This step can be omitted if the flow cell has been checked previously.

      See the flow cell check instructions in the MinKNOW protocol for more information.

    6. Slide the priming port cover clockwise to open the priming port.

      Flow Cell Loading Diagrams Step 2

    7. Important Take care when drawing back buffer from the flow cell. Do not remove more than 20-30 µl, and make sure that the array of pores are covered by buffer at all times. Introducing air bubbles into the array can irreversibly damage pores.
    8. After opening the priming port, check for a small air bubble under the cover. Draw back a small volume to remove any bubbles:

      1. Set a P1000 pipette to 200 µl
      2. Insert the tip into the priming port
      3. Turn the wheel until the dial shows 220-230 µl, to draw back 20-30 µl, or until you can see a small volume of buffer entering the pipette tip

      Note: Visually check that there is continuous buffer from the priming port across the sensor array.

      Flow Cell Loading Diagrams Step 03 V5

    9. Load 800 µl of the priming mix into the flow cell via the priming port, avoiding the introduction of air bubbles. Wait for five minutes. During this time, prepare the library for loading by following the steps below.

      Flow Cell Loading Diagrams Step 04 V5

    10. Important Thoroughly mix the contents of the RRB tube by vortexing or pipetting, and spin down briefly.
    11. Take 20 µl of the prepared RNA library and mix it with 17.5 µl of nuclease-free water.
    12. In a new tube, prepare the library for loading as follows:

      Reagent Volume per flow cell
      RNA Running Buffer (RRB) 37.5 µl
      RNA library in nuclease-free water 37.5 µl
      Total 75 µl

    13. Complete the flow cell priming:

      1. Gently lift the SpotON sample port cover to make the SpotON sample port accessible.
      2. Load 200 µl of the priming mix into the flow cell priming port (not the SpotON sample port), avoiding the introduction of air bubbles.

      Flow Cell Loading Diagrams Step 5

      Flow Cell Loading Diagrams Step 06 V5

    14. Mix the prepared library gently by pipetting up and down just prior to loading.
    15. Add 75 μl of sample to the Flow Cell via the SpotON sample port in a dropwise fashion. Ensure each drop flows into the port before adding the next.

      Flow Cell Loading Diagrams Step 07 V5

    16. Gently replace the SpotON sample port cover, making sure the bung enters the SpotON port and close the priming port.

      Step 8 update

      Flow Cell Loading Diagrams Step 9

    17. Important Install the light shield on your flow cell as soon as library has been loaded for optimal sequencing output.

      We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.

    18. Place the light shield onto the flow cell, as follows:

      1. Carefully place the leading edge of the light shield against the clip.
        Note: Do not force the light shield underneath the clip.

      2. Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.

      J2264 - Light shield animation Flow Cell FAW optimised

    19. Caution The MinION Flow Cell Light Shield is not secured to the flow cell and careful handling is required after installation.
    20. End of step Close the device lid and set up a sequencing run on MinKNOW.
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Priming and loading the flow cell

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Materials
  • Prepared RNA library
  • RNA Running Buffer (RRB)
  • Flow Cell Priming Kit (EXP-FLP002)
  • Consumables
  • PromethION Flow Cell
  • 1.5 ml Eppendorf DNA LoBind tubes
  • Nuclease-free water (e.g. ThermoFisher, AM9937)
  • Equipment
  • PromethION sequencing device
  • PromethION 2 Solo device
  • PromethION Flow Cell Light Shield
  • P1000 pipette and tips
  • P200 pipette and tips
  • P20 pipette and tips
    1. Thaw the RNA Running Buffer (RRB), Flush Tether (FLT) and one tube of Flush Buffer (FB) at room temperature.
    2. Mix the RNA Running Buffer (RRB), Flush Buffer (FB) and Flush Tether (FLT) tubes thoroughly by vortexing and spin down at room temperature.
    3. To prepare the flow cell priming mix, add 30 µl of thawed and mixed Flush Tether (FLT) directly to the tube of thawed and mixed Flush Buffer (FB), and mix by vortexing.
    4. Important After taking flow cells out of the fridge, wait 20 minutes before inserting the flow cell into the PromethION for the flow cell to come to room temperature. Condensation can form on the flow cell in humid environments. Inspect the gold connector pins on the top and underside of the flow cell for condensation and wipe off with a lint-free wipe if any is observed. Ensure the heat pad (black pad) is present on the underside of the flow cell.
    5. For PromethION 2 Solo, load the flow cell(s) as follows:

      1. Place the flow cell flat on the metal plate.

      2. Slide the flow cell into the docking port until the gold pins or green board cannot be seen.

      J2068 FC-into-P2-animation V5

    6. For the PromethION 24/48, load the flow cell(s) into the docking ports:

      1. Line up the flow cell with the connector horizontally and vertically before smoothly inserting into position.
      2. Press down firmly onto the flow cell and ensure the latch engages and clicks into place.

      Step 1a V3

      Step 1B

    7. Important Insertion of the flow cells at the wrong angle can cause damage to the pins on the PromethION and affect your sequencing results. If you find the pins on a PromethION position are damaged, please contact support@nanoporetech.com for assistance.

      Screenshot 2021-04-08 at 12.08.37

    8. Slide the inlet port cover clockwise to open.

      Prom loading 2

    9. Important Take care when drawing back buffer from the flow cell. Do not remove more than 20-30 µl, and make sure that the array of pores are covered by buffer at all times. Introducing air bubbles into the array can irreversibly damage pores.
    10. After opening the inlet port, draw back a small volume to remove any air bubbles:

      1. Set a P1000 pipette tip to 200 µl.
      2. Insert the tip into the inlet port.
      3. Turn the wheel until the dial shows 220-230 µl, or until you see a small volume of buffer entering the pipette tip.

      Step 3 v1

    11. Load 500 µl of the priming mix into the flow cell via the inlet port, avoiding the introduction of air bubbles. Wait five minutes.

      Step 4 v1

    12. Important Thoroughly mix the contents of the RRB tube by vortexing or pipetting, and spin down briefly.
    13. Prepare the library for loading in a single 1.5 ml Eppendorf DNA LoBind as follows:

      Reagent Volume
      RNA library 40 µl
      Nuclease-free water 35 µl
      RRB 75 µl
      Total 150 µl

    14. Complete the flow cell priming by slowly loading 500 µl of the priming mix into the inlet port.

      Step 5 v1

    15. Mix the prepared library gently by pipetting up and down just prior to loading.
    16. Using a P1000, insert the pipette tip into the inlet port and add 150 µl of library.

      Step 6 v2

    17. Close the valve to seal the inlet port.

      Step 7 V2

    18. Important Install the light shield on your flow cell as soon as library has been loaded for optimal sequencing output.

      We recommend leaving the light shield on the flow cell when library is loaded, including during any washing and reloading steps. The shield can be removed when the library has been removed from the flow cell.

    19. If the light shield has been removed from the flow cell, install the light shield as follows:

      1. Align the inlet port cut out of the light shield with the inlet port cover on the flow cell. The leading edge of the light shield should sit above the flow cell ID.
      2. Firmly press the light shield around the inlet port cover. The inlet port clip will click into place underneath the inlet port cover.

      J2264 - Light shield animation PromethION Flow Cell 8a FAW

      J2264 - Light shield animation PromethION Flow Cell 8b FAW

    20. End of step Close the PromethION lid when ready to start a sequencing run on MinKNOW.

      Wait a minimum of 10 minutes after loading the flow cells onto the PromethION before initiating any experiments. This will help to increase the sequencing output.

    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Data acquisition and basecalling

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    1. How to start sequencing

      Once you have loaded your flow cell, the sequencing run can be started on MinKNOW, our sequencing software that controls the device, data acquisition and real-time basecalling. For more detailed information on setting up and using MinKNOW, please see the MinKNOW protocol.

      MinKNOW can be used and set up to sequence in multiple ways:

      • On a computer either directly or remotely connected to a sequencing device.
      • Directly on a GridION or PromethION 24/48 sequencing device.

      For more information on using MinKNOW on a sequencing device, please see the device user manuals:


      To start a sequencing run on MinKNOW:

      1. Navigate to the start page and click Start sequencing.

      2. Fill in your experiment details, such as name and flow cell position and sample ID.

      3. Select the sequencing kit used in the library preparation on the Kit page.

      4. Configure the sequencing and output parameters for your sequencing run or keep to the default settings on the Run configuration tab.

      Note: If basecalling was turned off when a sequencing run was set up, basecalling can be performed post-run on MinKNOW. For more information, please see the MinKNOW protocol.

      5. Click Start to initiate the sequencing run.

    2. Data analysis after sequencing

      After sequencing has completed on MinKNOW, the flow cell can be reused or returned, as outlined in the Flow cell reuse and returns section.

      After sequencing and basecalling, the data can be analysed. For further information about options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.

      In the Downstream analysis section, we outline further options for analysing your data.

    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Data acquisition and basecalling

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    Data acquisition and basecalling

    1. How to start sequencing

      Once you have loaded your flow cell, the sequencing run can be started on MinKNOW, our sequencing software that controls the device, data acquisition and real-time basecalling. For more detailed information on setting up and using MinKNOW, please see the MinKNOW protocol.

      MinKNOW can be used and set up to sequence in multiple ways:

      • On a computer either direcly or remotely connected to a sequencing device.
      • Directly on a GridION or PromethION 24/48 sequencing device.

      For more information on using MinKNOW on a sequencing device, please see the device user manuals:


      To start a sequencing run on MinKNOW:

      1. Navigate to the start page and click Start sequencing.

      2. Fill in your experiment details, such as name and flow cell position and sample ID.

      3. Select the sequencing kit used in the library preparation on the Kit page.

      4. Configure the sequencing and output parameters for your sequencing run or keep to the default settings on the Run configuration tab.

      Note: If basecalling was turned off when a sequencing run was set up, basecalling can be performed post-run on MinKNOW. For more information, please see the MinKNOW protocol.

      5. Click Start to initiate the sequencing run.

    2. Data analysis after sequencing

      After sequencing has completed on MinKNOW, the flow cell can be reused or returned, as outlined in the Flow cell reuse and returns section.

      After sequencing and basecalling, the data can be analysed. For further information about options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.

      In the Downstream analysis section, we outline further options for analysing your data.

    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Data acquisition and basecalling

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    Data acquisition and basecalling

    Data acquisition and basecalling

    1. How to start sequencing

      Once you have loaded your flow cell, the sequencing run can be started on MinKNOW, our sequencing software that controls the device, data acquisition and real-time basecalling. For more detailed information on setting up and using MinKNOW, please see the MinKNOW protocol.

      MinKNOW can be used and set up to sequence in multiple ways:

      • On a computer either direcly or remotely connected to a sequencing device.
      • Directly on a GridION or PromethION 24/48 sequencing device.

      For more information on using MinKNOW on a sequencing device, please see the device user manuals:


      To start a sequencing run on MinKNOW:

      1. Navigate to the start page and click Start sequencing.
      start

      2. Fill in your experiment details, such as name and flow cell position and sample ID.
      Grid start seq

      3. Select the sequencing kit used in the library preparation on the Kit page.
      Kit selection non specific

      4. Configure the sequencing parameters for your sequencing run or keep to the default settings on the Run options and Analysis tabs.

      Note: If basecalling was turned off when a sequencing run was set up, basecalling can be performed post-run on MinKNOW. For more information, please see the MinKNOW protocol.

      5. On the Output page, set up the output parameters or keep to the default settings.
      step5c

      6. Click Start on the Review page to start the sequencing run.
      Step6

    2. Data analysis after sequencing

      After sequencing and basecalling, the data can be analysed. For further information about options for basecalling and post-basecalling analysis, please refer to the Data Analysis document.

      In the Downstream analysis section, we outline further options for analysing your data.

    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Flow cell reuse and returns

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    Data acquisition and basecalling

    Data acquisition and basecalling

    Flow cell reuse and returns

    Materials
  • Flow Cell Wash Kit (EXP-WSH004)
    1. After your sequencing experiment is complete, if you would like to reuse the flow cell, please follow the Flow Cell Wash Kit protocol and store the washed flow cell at +2°C to +8°C.

      The Flow Cell Wash Kit protocol is available on the Nanopore Community.

    2. Alternatively, follow the returns procedure to send the flow cell back to Oxford Nanopore.

      Instructions for returning flow cells can be found here.

    3. Important If you encounter issues or have questions about your sequencing experiment, please refer to the Troubleshooting Guide that can be found in the online version of this protocol.
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Downstream analysis

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    Data acquisition and basecalling

    Data acquisition and basecalling

    Flow cell reuse and returns

    Downstream analysis

    1. Post-basecalling analysis

      There are several options for further analysing your basecalled data:

      1. EPI2ME workflows

      For in-depth data analysis, Oxford Nanopore Technologies offers a range of bioinformatics tutorials and workflows available in EPI2ME. The platform provides a vehicle where workflows deposited in GitHub by our Research and Applications teams can be showcased with descriptive texts, functional bioinformatics code and example data.

      2. Research analysis tools

      Oxford Nanopore Technologies' Research division has created a number of analysis tools, which are available in the Oxford Nanopore GitHub repository. The tools are aimed at advanced users, and contain instructions for how to install and run the software. They are provided as-is, with minimal support.

      3. Community-developed analysis tools

      If a data analysis method for your research question is not provided in any of the resources above, please refer to the resource centre and search for bioinformatics tools for your application. Numerous members of the Nanopore Community have developed their own tools and pipelines for analysing nanopore sequencing data, most of which are available on GitHub. Please be aware that these tools are not supported by Oxford Nanopore Technologies, and are not guaranteed to be compatible with the latest chemistry/software configuration.

    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Issues during DNA/RNA extraction and library preparation

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    Data acquisition and basecalling

    Data acquisition and basecalling

    Flow cell reuse and returns

    Downstream analysis

    Issues during DNA/RNA extraction and library preparation

    1. Below is a list of the most commonly encountered issues, with some suggested causes and solutions.

      We also have an FAQ section available on the Nanopore Community Support section.

      If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.

    2. Low sample quality
      Observation Possible cause Comments and actions
      Low DNA purity (Nanodrop reading for DNA OD 260/280 is <1.8 and OD 260/230 is <2.0–2.2) The DNA extraction method does not provide the required purity The effects of contaminants are shown in the Contaminants document. Please try an alternative extraction method that does not result in contaminant carryover.

      Consider performing an additional SPRI clean-up step.
      Low RNA integrity (RNA integrity number <9.5 RIN, or the rRNA band is shown as a smear on the gel) The RNA degraded during extraction Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page.
      RNA has a shorter than expected fragment length The RNA degraded during extraction Try a different RNA extraction method. For more info on RIN, please see the RNA Integrity Number document. Further information can be found in the DNA/RNA Handling page.

      We recommend working in an RNase-free environment, and to keep your lab equipment RNase-free when working with RNA.
    3. Low DNA recovery after AMPure bead clean-up
      Observation Possible cause Comments and actions
      Low recovery DNA loss due to a lower than intended AMPure beads-to-sample ratio 1. AMPure beads settle quickly, so ensure they are well resuspended before adding them to the sample.

      2. When the AMPure beads-to-sample ratio is lower than 0.4:1, DNA fragments of any size will be lost during the clean-up.
      Low recovery DNA fragments are shorter than expected The lower the AMPure beads-to-sample ratio, the more stringent the selection against short fragments. Please always determine the input DNA length on an agarose gel (or other gel electrophoresis methods) and then calculate the appropriate amount of AMPure beads to use. SPRI cleanup
      Low recovery after end-prep The wash step used ethanol <70% DNA will be eluted from the beads when using ethanol <70%. Make sure to use the correct percentage.
    Documentation Prepare Library Prep Protocols Direct RNA sequencing (SQK-RNA002)

    Issues during the sequencing run

    Computer requirements and software

    Computer requirements and software

    Computer requirements and software

    Data acquisition and basecalling

    Data acquisition and basecalling

    Data acquisition and basecalling

    Flow cell reuse and returns

    Downstream analysis

    Issues during DNA/RNA extraction and library preparation

    Issues during the sequencing run

    1. Below is a list of the most commonly encountered issues, with some suggested causes and solutions.

      We also have an FAQ section available on the Nanopore Community Support section.

      If you have tried our suggested solutions and the issue still persists, please contact Technical Support via email (support@nanoporetech.com) or via LiveChat in the Nanopore Community.

    2. Fewer pores at the start of sequencing than after Flow Cell Check
      Observation Possible cause Comments and actions
      MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check An air bubble was introduced into the nanopore array After the Flow Cell Check it is essential to remove any air bubbles near the priming port before priming the flow cell. If not removed, the air bubble can travel to the nanopore array and irreversibly damage the nanopores that have been exposed to air. The best practice to prevent this from happening is demonstrated in this video.
      MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check The flow cell is not correctly inserted into the device Stop the sequencing run, remove the flow cell from the sequencing device and insert it again, checking that the flow cell is firmly seated in the device and that it has reached the target temperature. If applicable, try a different position on the device (GridION/PromethION).
      MinKNOW reported a lower number of pores at the start of sequencing than the number reported by the Flow Cell Check Contaminations in the library damaged or blocked the pores The pore count during the Flow Cell Check is performed using the QC DNA molecules present in the flow cell storage buffer. At the start of sequencing, the library itself is used to estimate the number of active pores. Because of this, variability of about 10% in the number of pores is expected. A significantly lower pore count reported at the start of sequencing can be due to contaminants in the library that have damaged the membranes or blocked the pores. Alternative DNA/RNA extraction or purification methods may be needed to improve the purity of the input material. The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover.
    3. MinKNOW script failed
      Observation Possible cause Comments and actions
      MinKNOW shows "Script failed"
      Restart the computer and then restart MinKNOW. If the issue persists, please collect the MinKNOW log files and contact Technical Support. If you do not have another sequencing device available, we recommend storing the flow cell and the loaded library at 4°C and contact Technical Support for further storage guidance.
    4. Pore occupancy below 40%
      Observation Possible cause Comments and actions
      Pore occupancy <40% Not enough library was loaded on the flow cell Ensure you load the recommended amount of good quality library in the relevant library prep protocol onto your flow cell. Please quantify the library before loading and calculate mols using tools like the Promega Biomath Calculator, choosing "dsDNA: µg to pmol"
      Pore occupancy close to 0 The Ligation Sequencing Kit was used, and sequencing adapters did not ligate to the DNA Make sure to use the NEBNext Quick Ligation Module (E6056) and Oxford Nanopore Technologies Ligation Buffer (LNB, provided in the sequencing kit) at the sequencing adapter ligation step, and use the correct amount of each reagent. A Lambda control library can be prepared to test the integrity of the third-party reagents.
      Pore occupancy close to 0 The Ligation Sequencing Kit was used, and ethanol was used instead of LFB or SFB at the wash step after sequencing adapter ligation Ethanol can denature the motor protein on the sequencing adapters. Make sure the LFB or SFB buffer was used after ligation of sequencing adapters.
      Pore occupancy close to 0 No tether on the flow cell Tethers are adding during flow cell priming (FLT/FCT tube). Make sure FLT/FCT was added to FB/FCF before priming.
    5. Shorter than expected read length
      Observation Possible cause Comments and actions
      Shorter than expected read length Unwanted fragmentation of DNA sample Read length reflects input DNA fragment length. Input DNA can be fragmented during extraction and library prep.

      1. Please review the Extraction Methods in the Nanopore Community for best practice for extraction.

      2. Visualise the input DNA fragment length distribution on an agarose gel before proceeding to the library prep. DNA gel2 In the image above, Sample 1 is of high molecular weight, whereas Sample 2 has been fragmented.

      3. During library prep, avoid pipetting and vortexing when mixing reagents. Flicking or inverting the tube is sufficient.
    6. Large proportion of unavailable pores
      Observation Possible cause Comments and actions
      Large proportion of unavailable pores (shown as blue in the channels panel and pore activity plot)

      image2022-3-25 10-43-25 The pore activity plot above shows an increasing proportion of "unavailable" pores over time.
      Contaminants are present in the sample Some contaminants can be cleared from the pores by the unblocking function built into MinKNOW. If this is successful, the pore status will change to "sequencing pore". If the portion of unavailable pores stays large or increases:

      1. A nuclease flush using the Flow Cell Wash Kit (EXP-WSH004) can be performed, or
      2. Run several cycles of PCR to try and dilute any contaminants that may be causing problems.
    7. Large proportion of inactive pores
      Observation Possible cause Comments and actions
      Large proportion of inactive/unavailable pores (shown as light blue in the channels panel and pore activity plot. Pores or membranes are irreversibly damaged) Air bubbles have been introduced into the flow cell Air bubbles introduced through flow cell priming and library loading can irreversibly damage the pores. Watch the Priming and loading your flow cell video for best practice
      Large proportion of inactive/unavailable pores Certain compounds co-purified with DNA Known compounds, include polysaccharides, typically associate with plant genomic DNA.

      1. Please refer to the Plant leaf DNA extraction method.
      2. Clean-up using the QIAGEN PowerClean Pro kit.
      3. Perform a whole genome amplification with the original gDNA sample using the QIAGEN REPLI-g kit.
      Large proportion of inactive/unavailable pores Contaminants are present in the sample The effects of contaminants are shown in the Contaminants Know-how piece. Please try an alternative extraction method that does not result in contaminant carryover.
    8. Reduction in sequencing speed and q-score later into the run
      Observation Possible cause Comments and actions
      Reduction in sequencing speed and q-score later into the run For Kit 9 chemistry (e.g. SQK-LSK109), fast fuel consumption is typically seen when the flow cell is overloaded with library (please see the appropriate protocol for your DNA library to see the recommendation). Add more fuel to the flow cell by following the instructions in the MinKNOW protocol. In future experiments, load lower amounts of library to the flow cell.
    9. Temperature fluctuation
      Observation Possible cause Comments and actions
      Temperature fluctuation The flow cell has lost contact with the device Check that there is a heat pad covering the metal plate on the back of the flow cell. Re-insert the flow cell and press it down to make sure the connector pins are firmly in contact with the device. If the problem persists, please contact Technical Services.
    10. Failed to reach target temperature
      Observation Possible cause Comments and actions
      MinKNOW shows "Failed to reach target temperature" The instrument was placed in a location that is colder than normal room temperature, or a location with poor ventilation (which leads to the flow cells overheating) MinKNOW has a default timeframe for the flow cell to reach the target temperature. Once the timeframe is exceeded, an error message will appear and the sequencing experiment will continue. However, sequencing at an incorrect temperature may lead to a decrease in throughput and lower q-scores. Please adjust the location of the sequencing device to ensure that it is placed at room temperature with good ventilation, then re-start the process in MinKNOW. Please refer to this link for more information on MinION temperature control.
    11. Guppy – no input .fast5 was found or basecalled
      Observation Possible cause Comments and actions
      No input .fast5 was found or basecalled input_path did not point to the .fast5 file location The --input_path has to be followed by the full file path to the .fast5 files to be basecalled, and the location has to be accessible either locally or remotely through SSH.
      No input .fast5 was found or basecalled The .fast5 files were in a subfolder at the input_path location To allow Guppy to look into subfolders, add the --recursive flag to the command
    12. Guppy – no Pass or Fail folders were generated after basecalling
      Observation Possible cause Comments and actions
      No Pass or Fail folders were generated after basecalling The --qscore_filtering flag was not included in the command The --qscore_filtering flag enables filtering of reads into Pass and Fail folders inside the output folder, based on their strand q-score. When performing live basecalling in MinKNOW, a q-score of 7 (corresponding to a basecall accuracy of ~80%) is used to separate reads into Pass and Fail folders.
    13. Guppy – unusually slow processing on a GPU computer
      Observation Possible cause Comments and actions
      Unusually slow processing on a GPU computer The --device flag wasn't included in the command The --device flag specifies a GPU device to use for accelerate basecalling. If not included in the command, GPU will not be used. GPUs are counted from zero. An example is --device cuda:0 cuda:1, when 2 GPUs are specified to use by the Guppy command.