- Materials
-
- Loading Beads (LB)
- Sequencing Buffer (SQB)
- Flow Cell Priming Kit XL (EXP-FLP002-XL)
- Consumables
-
- Eppendorf twin.tec® PCR plate 96 LoBind, semi-skirted (Eppendorf™, cat # 0030129504) with heat seals
- OR 0.2 ml thin-walled PCR tubes
- 1.5 ml Eppendorf DNA LoBind tubes
- 15 or 50 ml Falcon tubes
- Nuclease-free water (e.g. ThermoFisher, AM9937)
- Equipment
-
- MinION or GridION device
- MinION and GridION Flow Cell Light Shield
- Multichannel pipettes suitable for dispensing 2–20 μl and 20–200 μl, and tips
- P1000 pipette and tips
- P100 pipette and tips
- P20 pipette and tips
- P10 pipette and tips
-
Thaw the Sequencing Buffer (SQB), Loading Beads (LB), Flush Tether (FLT) and Flush Buffer (FB) at room temperature, before mixing the reagents by vortexing and spin down at room temperature.
-
Prepare the flow cell Priming Mix: in a suitable vial, prepare a mix of Flush Buffer (FB) and Flush Tether (FLT) for priming all flow cells to be loaded. Allow 30 µl of Flush Tether (FLT) and 1.17 ml of Flush Buffer (FB) for every flow cell; no excess is required. Once combined, mix well by briefly vortexing.
-
Open the MinION Mk1B lid and slide the flow cell under the clip.
Press down firmly on the flow cell to ensure correct thermal and electrical contact.
-
Optional actionComplete a flow cell check to assess the number of pores available before loading the library.
This step can be omitted if the flow cell has been checked previously.
See the flow cell check instructions in the MinKNOW protocol for more information.
-
Slide the priming port cover clockwise to open the priming port.
-
After opening the priming port, check for a small air bubble under the cover. Draw back a small volume to remove any bubbles:
- Set a P1000 pipette to 200 µl
- Insert the tip into the priming port
- Turn the wheel until the dial shows 220-230 µl, to draw back 20-30 µl, or until you can see a small volume of buffer entering the pipette tip
Note: Visually check that there is continuous buffer from the priming port across the sensor array.
-
Load 800 µl of the priming mix into the flow cell via the priming port, avoiding the introduction of air bubbles. Wait for five minutes. During this time, prepare the library for loading by following the steps below.
-
Meanwhile, thoroughly mix the contents of the thawed Loading Beads (LB) tube(s) by vortexing.
-
In a separate 0.2 ml PCR tube for each library, prepare for loading by adding the following reagents:
Reagent Volume Sequencing Buffer (SQB) 37.5 µl Loading Beads (LB), mixed immediately before use 25.5 µl DNA library 12 µl Total 75 µl -
Complete the flow cell priming:
- Gently lift the SpotON sample port cover to make the SpotON sample port accessible.
- Load 200 µl of the priming mix into the flow cell priming port (not the SpotON sample port), avoiding the introduction of air bubbles.
-
Mix the prepared libraries gently by pipetting up and down just prior to loading.
-
Add 75 μl of the prepared library to the flow cell via the SpotON sample port in a dropwise fashion. Ensure each drop flows into the port before adding the next.
-
Gently replace the SpotON sample port cover, making sure the bung enters the SpotON port and close the priming port.
-
Place the light shield onto the flow cell, as follows:
Carefully place the leading edge of the light shield against the clip.
Note: Do not force the light shield underneath the clip.Gently lower the light shield onto the flow cell. The light shield should sit around the SpotON cover, covering the entire top section of the flow cell.