Flow cell check
Navigate to the Start homepage and select Flow Cell Check.
Choose the flow cell type from the drop-down menu. The flow cell ID will be automatically filled in. Click Start to begin the flow cell check.
The user will be automatically navigated to the System Overview page. A loading bar will be displayed under the flow cell during the checks.
The flow cell will cycle through the 2048 wells to see how many pores are active on the sensor array. It will do this in four groups of 512 at a time.
There may be a slight delay while the MinION Mk1B reaches its optimal sequencing temperature, as shown in the progress bar at the top of the GUI.
Please see this FAQ for more information.
Flow cell check completion will be indicated on the System Overview page with an icon on the flow cell with the number of pores available.
At the end of the flow cell check, MinKNOW will report the total number of nanopores in the flow cell.
The warranty for flow cells covers 800 pores or above. If you have fewer than 800 pores, please follow the instructions in this FAQ. If you still see under 800 pores, contact support@nanoporetech.com .
You have confirmed that your flow cell is working correctly and is ready for sequencing a biological sample, which you will prepare in the next section.