5. Sequencing run
  • Introduction
  • 1. Software installation
  • 2. Hardware check
  • 3. Flow cell check
  • 4. Library preparation
  • 5. Sequencing run
  • 6. Results and metrics
  • 7. Next steps
56%
You are in the ‘Windows’ branch of Software installation.
Change branch:
5.2

Prepare your priming mix

Need help?

Make sure the contents of the Flow Cell Flush (FCF) and Flow Cell Tether (FCT) tubes are fully defrosted. Mix the tubes by pipetting up and down, and then spin down in a microfuge.

Important

For optimal sequencing performance and improved output on R10.4.1 flow cells (FLO-MIN114), we recommend adding Bovine Serum Albumin (BSA) to the flow cell priming mix at a final concentration of 0.2 mg/ml. We do not recommend using recombinant BSA.

Transfer 30 µl of Flow Cell Tether (FCT) directly into one of the Flow Cell Flush (FCF) tubes.

fct-fcf

Add 5 µl of Bovine Serum Albumin (BSA) at 50 mg/ml to the tube of Flow Cell Flush (FCF).

bsa-5ul

Mix the contents of the tube by flicking the tube with your finger and then spin down briefly in a microfuge.

Priming mix

The resulting buffer is now referred to as the Priming Mix (FCF tube). You will use this to flush out the storage buffer from the flow cell.

Prev 5.3 Prime your flow cell