- Materials
-
- 300 ng end-prepped DNA in 30 µl
- PCR Adapter (PCA)
- Consumables
-
- Nuclease-free water (e.g. ThermoFisher, AM9937)
- NEB Blunt/TA Ligase Master Mix (NEB, cat # M0367)
- Agencourt AMPure XP beads (Beckman Coulter™ cat # A63881)
- Freshly prepared 80% ethanol in nuclease-free water
- 1.5 ml Eppendorf DNA LoBind tubes
- Equipment
-
- Magnetic rack, suitable for 1.5 ml Eppendorf tubes
- Hula mixer (gentle rotator mixer)
- Qubit fluorometer (or equivalent for QC check)
- Microfuge
- Vortex mixer
-
Add the reagents in the order given below, mixing by pipetting 10-20 times between each sequential addition:
Reagent Volume End-prepped DNA 30 µl PCR Adapters (PCA) 20 µl NEB Blunt/TA Ligase Master Mix 50 µl Total 100 µl -
Ensure the components are thoroughly mixed by pipetting, and spin down.
-
Incubate the reaction for 10 minutes at room temperature.
-
Resuspend the AMPure XP beads by vortexing.
-
Add 100 µl of resuspended AMPure XP beads to the end-prep reaction and mix by pipetting.
End-prep cleanup demo
-
Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
-
Prepare 500 μl of fresh 80% ethanol in nuclease-free water.
-
Spin down the sample and pellet on a magnet. Keep the tube on the magnet, and pipette off the supernatant when clear and colourless.
-
Keep the tube on the magnet and wash the beads with 200 µl of freshly prepared 80% ethanol without disturbing the pellet. Remove the ethanol using a pipette and discard.
-
Repeat the previous step.
-
Spin down and place the tube back on the magnet. Pipette off any residual ethanol. Allow to dry for ~30 seconds, but do not dry the pellet to the point of cracking.
-
Remove the tube from the magnetic rack and resuspend pellet in 49 µl nuclease-free water. Incubate for 2 minutes at room temperature.
-
Pellet the beads on a magnet until the eluate is clear and colourless.
-
Remove and retain 49 µl of eluate into a clean 1.5 ml Eppendorf DNA LoBind tube.
-
Quantify 1 µl of adapted DNA using a Qubit fluorometer.