Before starting
This protocol outlines how to carry out PCR amplification and native barcoding of influenza amplicons on a 96-well plate using the Native Barcoding Expansion 96 (EXP-NBD196) in conjunction with Ligation Sequencing Kit (SQK-LSK109).
When processing multiple samples at once, we recommend making master mixes with an additional 10% of the volume. We also recommend using a template-free pre-PCR hood for making up the master mixes, and a separate template pre-PCR hood for handling the samples. It is important to clean and/or UV irradiate these hoods between sample batches. Furthermore, to track and monitor cross-contamination events, it is important to run a negative control reaction at the reverse transcription stage using nuclease-free water instead of sample, and carrying this control through the rest of the prep.
All post-PCR procedures must be carried out in a separate area to the pre-PCR preparation, with dedicated equipment for liquid handling in each area.
-
Introduction to the influenza whole genome sequencing protocol
To enable support for the rapidly expanding user requests, the team at Oxford Nanopore Technologies have put together a workflow describing how to carry out PCR amplification and native barcoding of influenza amplicons using the Native Barcoding Expansion 96 (EXP-NBD196) in conjunction with Ligation Sequencing Kit (SQK-LSK109). There are 96 unique barcodes available, allowing the user to pool up to 96 different samples in one sequencing experiment.
While this protocol is available in the Nanopore Community, we kindly ask users to ensure they are citing the following references, that this protocol is based on.
Single-reaction genomic amplification accelerates sequencing and vaccine production for classical and Swine origin human influenza A viruses by Bin Zhou et al., 2009. and Universal influenza B virus genomic amplification facilitates sequencing, diagnostics, and reverse genetics by Bin Zhou et al., 2014.
Steps in the sequencing workflow:
Prepare for your experiment
You will need to:
- Ensure you have the sequencing kit, the correct equipment and third-party reagents
- Download the software for acquiring and analysing your data
- Check your flow cell to ensure it has enough pores for a good sequencing run
Prepare your library
You will need to:
- Perform RT-PCR amplification on influenza samples
- Prepare the DNA ends for adapter attachment
- Ligate native barcodes supplied in the kit to the DNA ends
- Ligate sequencing adapters supplied in the kit to the DNA ends
- Prime the flow cell, and load your DNA library into the flow cell
Sequencing
You will need to:
- Start a sequencing run using the MinKNOW software, which will collect raw data from the device and convert it into basecalled reads
- Demultiplex barcoded reads in MinKNOW or Guppy basecalling, choosing the EXP-NBD196 kit option
- Use the wf-flu workflow in EPI2ME Labs to analyse your data