- Materials
-
- Barcode Adapter (BCA)
- Consumables
-
- NEB Blunt/TA Ligase Master Mix (NEB, cat # M0367)
- Agencourt AMPure XP beads (Beckman Coulter™ cat # A63881)
- Nuclease-free water (e.g. ThermoFisher, AM9937)
- Freshly prepared 70% ethanol in nuclease-free water
- 10 mM Tris-HCl pH 8.5
- 0.2 ml 96-well PCR plate
- Equipment
-
- Microplate centrifuge, e.g. Fisherbrand™ Mini Plate Spinner Centrifuge (Fisher Scientific, 11766427)
- Hula mixer (gentle rotator mixer)
- Vortex mixer
- Magnetic rack suitable for 96-well PCR plates, e.g. DynaMag™-96 Side Skirted Magnet (Thermo Fisher, cat # 12027)
- Ice bucket with ice
- Multichannel pipette and tips
- Optional equipment
-
- Agilent Bioanalyzer (or equivalent)
-
Add the reagents to a fresh 96-well plate, in the order given below:
Between each addition, pipette mix 10-20 times.
Reagent Volume End-prepped DNA 30 µl Barcode Adapter 20 µl Blunt/TA Ligase Master Mix 50 µl Total 100 µl -
Ensure the components are thoroughly mixed by pipetting.
-
Seal the plate with adhesive film or PCR strip caps and briefly spin down in a plate spinner.
-
Incubate the reaction for 10 minutes at room temperature.
-
Resuspend the AMPure XP beads by vortexing.
-
Add 40 µl of resuspended AMPure XP beads to each sample and mix by pipetting up and down ten times.
-
Allow DNA to bind to beads for 5 minutes at room temperature.
-
Prepare sufficient fresh 70% ethanol in nuclease-free water.
-
Place on a magnetic rack, allow beads to pellet and pipette off supernatant.
-
Keep the tube on the magnet and wash the beads with 180 µl of freshly-prepared 70% ethanol without disturbing the pellet. Remove the ethanol using a pipette and discard.
-
Repeat the previous step.
-
Cover the plate with adhesive film and leave plate on magnet for 2 minutes to allow residual liquid to collect at the bottom. Remove the adhesive film, return the plate to the magnet and aspirate residual wash solution.
-
Briefly incubate the plate on a thermal cycler at 37° C with the lid open and the plate wells unsealed.
-
Remove the plate from the magnet and resuspend pellet in 25 µl nuclease-free water. Incubate for 2 minutes at room temperature.
-
Pellet the beads on a magnet until the eluate is clear and colourless.
-
Remove eluate once it is clear and colourless. Transfer each eluted sample to a new 96-well PCR plate.
-
Quantify 1 µl of end-prepped DNA using a Qubit fluorometer.
-
Dilute the library to a concentration of 10 ng/µl with nuclease-free water or 10 mM Tris-HCl pH 8.5.