- Materials
-
- Primer Mix (PRM)
- Consumables
-
- LongAmp Taq 2X Master Mix (e.g. NEB, cat # M0287)
- Agencourt AMPure XP beads (Beckman Coulter™ cat # A63881)
- Nuclease-free water (e.g. ThermoFisher, AM9937)
- Freshly prepared 80% ethanol in nuclease-free water
- 1.5 ml Eppendorf DNA LoBind tubes
- 0.2 ml thin-walled PCR tubes
- Equipment
-
- Thermal cycler
- Ice bucket with ice
- Magnetic rack, suitable for 1.5 ml Eppendorf tubes
- Hula mixer (gentle rotator mixer)
- Vortex mixer
- Qubit fluorometer (or equivalent for QC check)
-
In a 0.2 ml thin-walled PCR tube mix the following:
Reagent Volume LongAmp Taq 2X Master Mix 50 µl PRM Adapters (10 μM) 2 µl Template DNA 48 µl Total 100 µl -
Amplify using the following cycling conditions:
Cycle step Temperature Time No. of cycles Initial denaturation 95 °C 3 mins 1 Denaturation 98 °C 20 secs 6 (b) Annealing 62 °C (a) 15 secs (a) 6 (b) Extension 65 °C (c) 3 mins 6 (b) Final extension 65 °C 3 mins 1 Hold 4 °C ∞ a. This is specific to the Oxford Nanopore primer and should be maintained
b. Adjust accordingly if input quantities are altered
c. This temperature is determined by the type of polymerase that is being used (given here for LongAmp Taq polymerase)
-
Resuspend the AMPure XP beads by vortexing.
-
Add 100 μl of the resuspended AMPure XP beads to the sample, and mix by flicking the tube.
-
Incubate on a Hula mixer (rotator mixer) for 5 minutes at room temperature.
-
Prepare 500 μl of fresh 80% ethanol in nuclease-free water.
-
Spin down the sample and pellet on a magnet. Keep the tube on the magnet, and pipette off the supernatant when clear and colourless.
-
Keep the tube on the magnet and wash the beads with 200 µl of freshly prepared 80% ethanol without disturbing the pellet. Remove the ethanol using a pipette and discard.
-
Repeat the previous step.
-
Spin down and place the tube back on the magnet. Pipette off any residual ethanol. Allow to dry for ~30 seconds, but do not dry the pellet to the point of cracking.
-
Remove the tube from the magnetic rack and resuspend the pellet in 35 µl nuclease-free water. Incubate for 2 minutes at room temperature.
-
Pellet the beads on a magnet until the eluate is clear and colourless.
-
Remove and retain 35 µl of eluate into a clean 1.5 ml Eppendorf DNA LoBind tube.
-
Quantify 2 µl of amplified DNA using a Qubit fluorometer.