4. Library preparation
  • Introduction
  • 1. Software installation
  • 2. Hardware check
  • 3. Flow cell check
  • 4. Library preparation
  • 5. Sequencing run
  • 6. Results and metrics
  • 7. Next steps
40%
4.1

Prepare your reagents

Need help?

Keep these on ice:

  • RT Adapter (RTA)
  • RNA Calibration Strand (RCS)
  • RNA Ligation Adapter (RLA)
  • 10 mM dNTP solution

Defrost these at room temperature:

  • NEBNext Quick Ligation Reaction Buffer
  • T4 DNA Ligase

Once the contents of each tube have thawed, spin down the tubes in a microfuge and mix the contents by pipetting up and down for 10 full-volume pipette mixes.

Defrost these at room temperature:

  • Wash Buffer (WSB)
  • RNA Elution Buffer (REB)

Vortex both tubes, then spin them down and place on ice.

Keep the reagent tubes on ice when they are not in use.

Set up a thermal cycler with the program 50°C for 50 minutes, then 70°C for 10 minutes, then hold at 4°C. Alternatively, set two heat blocks at 50°C and 70°C.

Prev 4.2 Reverse-transcribe your RNA