-
Eject the bootable USB by going to the File Explorer and selecting the Eject icon next to the USB device.
The keyboard and mouse may take several minutes to be recognised by the device during the first time booting up.
-
Press Ctrl+Alt+T to open a terminal window.
-
If your hostname has not been maintained through the restore process (it should be visible on the command line), type:
ont-gridion-serial set GXBXXXX
where
GXBXXXX
is the serial ID as included on the label on the back of your device. -
Press Enter.
-
Type "bash", then press Enter to confirm that the hostname has been set.
-
Configure your network.
A script has been written to help you to configure your network.
Open a Terminal window and enter:ont-network-settings
The options for this script are:
--dhcp
(optional) - Use DCHP and removes need for all other features except for the interfaces, bond and overwrite
--address ip_address
(required) - IP address of the device
--subnet subnet_mask
(required) - Subnet mask for the device
--gateway gateway_address
(required) - Default gateway for the device
--dns DNS_server1,DNS_server2
(required) - DNS servers for the device
--search your_domain
(required) - Your domain suffix (e.g. yourcompany.local)
--overwrite
(optional)
--help
- Details the options aboveTo update the settings, run the suitable command from the above options with the
--overwrite
flag and reboot your device. -
Open MinKNOW and check for software updates, or check for updates in the terminal:
sudo apt update
sudo apt install ont-gridion-release
ont-system-test
(not available on some ISO images)
ont-validate-install
You should see the following message if the installation was successful:
All Packages installed and nvidia is happy, so resetting ont-restore-apt
If the installation was unsuccessful, run the following commands:
sudo apt update
sudo apt dist-upgrade
sudo ont-validate-install
sudo reboot
If the issue has not been resolved after a reboot, contact Technical Support at support@nanoporetech.com.