Preparing for installation
Before you install the device, the following must be available:
- 1x Ethernet cable (1 Gbps)
- Ethernet port providing outbound-only access to the internet over port 443, and TCP access to AWS eu-west-1 IP ranges listed here for EPI2ME analysis
- Outbound-only access over HTTPS/port 443 to 52.17.110.146, 52.31.111.95, 79.125.100.3 or DNS rule for ping.oxfordnanoportal.com for telemetry feedback
- Outbound-only access over HTTPS/port 443 to 178.79.175.200 and 96.126.99.215 or DNS rule for cdn.oxfordnanoportal.com for software updates
- 1x monitor (HDMI/DisplayPort compatible)
- 1x appropriate monitor cable
- 1x USB connected keyboard
- 1x USB connected mouse
- Optional: UPS solution with C14 cable
-
Recommended laboratory location
- Place the device on a strong, clean laboratory bench
- Ensure Ethernet and power connectors are in close proximity
- Ensure power socket is easily accessible should a disconnection be required in case of emergency
- Sequencing should be carried out at a room temperature of +18°C to +25°C
- Allow 30 cm clearance to rear and sides of the device
- Do not cover any ventilation grilles
- WARNING - The rear of the device will heat up during operation
-
Unpack the device and place it on the bench where it is to be run.
-
Attach all the required cables to the device using the diagram below to assist.
- 1x Ethernet
- 1x HDMI or DisplayPort to monitor*
- 1x USB for keyboard
- 1x USB for mouse
- 1x C13 cable
*1x DisplayPort to HDMI adapter is provided
-
Connect the power cable to your mains/UPS solution. Use the mains cord supplied, or a suitable approved and rated power cord. Ensure the device is connected to a protective earth mains supply. Turn the power source on.
-
Double-click the MinKNOW icon located on the desktop to open the MinKNOW UI.
-
We recommend users to log into the MinKNOW software using their Community credentials.
To log in, you must be connected to the internet.
From here, you can also change the language in the user interface in the bottom right corner.
You will be prompted to enter your email address and Nanopore password:
If you experience login issues, please visit the Community Support channel (https://community.nanoporetech.com/support) and use Continue as guest for temporary use.
-
Select the sequencing device in the Connection Manager to connect to the computer.
-
You will be taken to the Sequencing Overview, displaying the connected flow cells or CTCs.