Power off the Data Acquisition Unit.
-
Power on the Data Acquisition Unit.
-
When the below screen appears, press F11 to enter the Boot Menu.
-
Choose the UEFI USB boot option using the cursor keys.
In the below example the USB drive is called SanDisk. If the system has not been prepared to boot from UEFI, this will not work. See the section Preparing the PromethION for UEFI boot for how to configure this.
-
Press Enter once the UEFI option is chosen, then highlight Customer Factory Reset, and press Enter. This option will detect your installation and try to keep network configurations, hostname and /data partition. We recommend backing up data prior to proceeding with this step wherever possible.
Make sure to connect your device to the network if configured to run online.
Option What it does When to use option Customer Factory Reset - Keeps network configuration
- Keeps hostname (i.e. PCA1XXX)
- Keeps data
- Does not clear the metadata or format all the disks- To reset the device to a MinKNOW version reflected on the image name (e.g. 22.10 ISO = MinKNOW 22.10)
- To reset the device to a known state whilst attempting to keep existing dataReset Install and keep /data - Keeps network configuration
- Keeps hostname (i.e. PCA1XXX)
- Keeps data
- Does not clear the metadata or format all the disksFunctionally identical to the Customer Factory Reset option Format Disks and lose /data - Does not keep network configuration
- Does not keep hostname (i.e. PCA1XXX)
- Does not keep data
- Clears the metadata and formats all the disks- Issue with data storage or a new unformatted drive (e.g. faulty drive or drive replaced)
- If the keep /data option does not resolve the issueSECURE WIPE - can take hours - Does not keep network configuration
- Does not keep hostname (i.e. PCA1XXX)
- Does not keep dataUse only if requested by a member of Technical Services - this option completely wipes the disks by writing all data to 0 -
You will see the screen below. Select Standard Installation. A confirmation prompt will appear; click Okay.
Note: The blue desktop may fill the entire screen. You can press
Esc
to view the menu in the default browser. -
The following screen will appear and the progress will be shown on the progress bar. The entire process may take ~1 hour to complete.
Note: It is normal for the system to restart multiple times. Do not turn off the Data Acquisition Unit while installation is in progress. It is also expected behaviour for the progress bars to stay temporarily stuck at specific stages.
-
Once the rebuild process has completed, the standard login screen will be displayed. After logging in, you will see the MinKNOW icon on the desktop. If the MinKNOW icon is not present, carry out one final power cycle of the Data Acquisition Unit.
Any previous changes to the device password will have defaulted back to
prom
. If you need to update the default password, follow the instructions here -
Once the MinKNOW icon is displayed on the desktop, power down the Data Acquisition Unit. To use the PromethION for sequencing, now power on the Sequencing Unit followed by the Data Acquisition Unit and check all flow cell positions are detected.