Boot to the "Platform Rescue and Recovery" screen.
-
Click the "Format disks and lose /data" option.
-
Once returned to the following screen, click the top "Customer Factory Reset im" option.
Note: this step may happen automatically and proceed immediately to step 4 without user input.
-
If the system takes more than 20 minutes between Step 2 and returning to the following screen, click "Standard Installation".
-
Allow the factory reset to continue.
Note: It is normal behaviour for the system to restart multiple times. Do not turn off the PromethION Data Acquisition Unit/GridION while the reset is in progress.
It is also expected behaviour for the progress bars to stay temporarily stuck at specific stages.
-
Once the reset is complete, you will be prompted to login.